Author Topic: Is readout compatible between scope manufacturers?  (Read 1644 times)
Offline alank2Topic starter
« on: October 20, 2014, 11:49:06 pm »
I see a lot of probes with a "readout" pin on them to set the 10X, etc.  Did they just choose a resistance that was 10X for example?
Is this compatible between scopes (tek, hp, rigol, etc.)?
Offline alex.forencich
    • Alex Forencich
Re: Is readout compatible between scope manufacturers?
« Reply #1 on: October 24, 2014, 05:06:56 am »
Yeah, there are standard values for the readout pin resistor.  All 10X probes should have the same resistor value, for example.  I think it is 10 or 11k.
Python-based instrument control: Python IVI, Python VXI-11, Python USBTMC
Offline David Hess
  • DavidH
« Reply #2 on: October 25, 2014, 03:03:12 pm »
At least for the standard x10, x100, and x1000 values I think they are all the same.  Some of the later values which were introduced might have varied between manufacturers.
As far as I know, the original standard came from Barry Gilbert when he designed the readout for the Tektronix 7000 series where the probe readout resistance directly alters the 1 of 10 encoded current into the analog readout generator.  It is possible to go though his design to calculate what the resistor values should be.