Author Topic: SRS PS310: ROM code supports GPIB option Needed  (Read 12747 times)
Offline z80twTopic starter
  • Country: tw
Re: SRS PS310: ROM code supports GPIB option Needed
« Reply #25 on: February 02, 2017, 03:07:09 pm »
I just replaced the ERPOM IC of PS310 with the one with ROM code of PS350 in it .
When turning it on the display showed the serial number "4243" and model no. "350" (1~2 seconds,pic 1),
after that the PS310 entered reset status(pic 2).I pressed the GPIB key and got gpib address "GP14" on display(pic 3).
I haven't tested the actual output voltage and GPIB communication yet.
Jack
« Last Edit: February 02, 2017, 03:10:31 pm by z80tw »
  • Country: tw
« Reply #26 on: February 03, 2017, 01:33:33 pm »
There is code at the beginning of the fw which read the keys.
I would guess it is the CLR key for SRAM reset.
And that piece of code indeeds writes default values into the SRAM (in 4000h).
But it then goes into a separate loop, not the main loop.
Today I did a little modification in  the ROM code of PS350:
Model No. 350------>310 (0157~0159: 33 35 30---->33 31 30)
Serial No. 4243----->7397(1FFC~ 1FFF: 34 32 34 33----->37 33 39 37)
After that I tested the PS310 with the modified ROM code of PS350:
1.Actual output voltage
SET/Display         Actual output
50V                    12.01V
100V                   24.46v
200v                    49.679v
500v                    124.498v
1000v                  250.498v
(PS310 Max output:1250V, PS350 Max output:5000V, 5000/1250=4)
2.GPIB communication:
PC---->PS310:*IDN?
PS310 response:StanfordResearchSystems,PS310,7397,1.40
PC---->PS310:VSET1.0E3 (set output=1000V)
PC---->PS310:VSET?
PS310 response:1.000E3 (=1000V)
PC---->PS310:HVON(output turned on)
PC---->PS310:HVOF(output turned off)
The GPIB works!
Jack
« Last Edit: February 03, 2017, 01:51:38 pm by z80tw »
Offline anotherlin
« Reply #27 on: February 03, 2017, 02:12:19 pm »
Today I did a little modification on  ROM code of PS350:
0157~0159      310 (Model No,PS310)
1FFC~1FFF       7397(Serial No,last 4 bytes, total 8K bytes)
Display               Actual voltage
50V                    12.01V
100V                   24.46v
200v                    49.679v
500v                    124.498v
1000v                  250.498v
Seems like "350" at 156h is for display purpose.
PS350 is 5kV vs PS310 is 1.25kV. Values are divided by 4, seems consistent.
I've found some code right at the beginning that check the keyboard (see IN A, (30h) at address 2Eh) and jumps to a subroutine instead of the main loop.
If you're not afraid, can you try to boot with key "ENT" and/or "7" pressed and see what happens?
"Lots of people have made $100K or more mistakes and didn't get the boot. It's called training, why fire them after such an expensive lesson?" -- EEVblog Electronics Community Forum
  • Country: tw
« Reply #28 on: February 03, 2017, 04:10:25 pm »
I did as you asked,this is what happened after booting with key ENT pressed,
the top segment(a) of left most seven segment LED light turned on(pic1).now the PS310 response only to two keys,
up-arrow and down-arrow  keys,
if you press up-arrow key the (a) segment turns off and (b) segment turns on,
push up -arrow key again (c) segment is on,(b) segment off.
if you continue to push the up-arrow key again and again the display will continue from segment (a) to (h)-(dot)(pic2 & pic3)
then repeat it again on next seven segment LED,one segment turns on at one time.
after two 7 segments displayed in this manner all 7 segment LEDs turns on(pic4),after that
is the indicator LED "trip" turns on then "AUTO"--->"MAN"--->"REAR"---->"SET"--->"LIMIT(voltage)"--->"LIMIT(ma)"--->"TRIP"---->"LIMIT(status)"
---->"REM"---->"SRQ",one by one at one time.
after that you push one key you will see a code on display(pic 5):
key press        display
Man                01
GPIB               02
left-arrow       03
Right arrow    04
up-arrow        05 (pic 5)
down-arrow   06
Enter              07
Select             08
STO                 09
RCL                 10
CLR                 11
.                      12
0                      13
7                       14
4                      15
1                        16
8                        17
5                         18
2                         19
9                        20
6                       21
3                       22
It looks like it's a key board and display test program.
Jack
« Last Edit: February 03, 2017, 04:13:19 pm by z80tw »
Offline anotherlin
« Reply #29 on: February 06, 2017, 08:11:50 am »
Great a (not so useful) undocumented feature :)
  • Country: tw
« Reply #30 on: February 06, 2017, 02:42:39 pm »
I set the GPIB bus address to 09 and saved it,then turned the power off and on again
,after power on I pressed the GPIB key this time I saw "GP09"(see pic) on the display,
I took out the EPROM(PS350) and verified it with the ROM file with my EPROM writer,
they are the same ,no change at all,thus the new GPIB address must be saved in SRAM.
I downloaded a free z80 disassembler and disassemled the binary files(PS310 & PS350),
Following information came from my brief analysis of PS350 ROM files and schematics:
1.Memory address of EPROM: 0000~1FFF (program space:8KB,ST M2764A)
2.SRAM: 4000~47FF (data space:2KB,Hitachi HM6116)
3.I/O address:
20h:Status1
28h:Status2
30h:-KBD
38h:-GPIB
48h:-Sets
50h:-DAC
58h:-LED
60h:-SEGB
68h:-SEGA
70h:-STROBE
78h:
bit0:-SHUTDOWN
bit1:UPOK
BIT2:-FLAG RESET
BIT3:-TIMER RESET
BIT4:(spare)
BIT5:-POS
bit6:-NEG
bit7:FILTER
Key code of "GPIB" :20h
I found only two read key code commands ("in a,(30h)") in the ROM file of PS350
1.0000~0044h:
002c db30      in      a,(30h) ----read key code
002e fe80      cp      80h ------check if "ENTER" / "7" key pressed or not
0030 ca1714    jp      z,1417h ------if yes then jump to 1417h,---the keyboard/display test program your asked me to check.
0033 c36800    jp      0068h-----if neither "ENTER" nor "7" pressed then jump to 0068h-----proceed regular power on procedures?
2.Subroutine 0FC2~0FD5h: 0FD2h  db30      in      a,(30h)
0fc2 3a1440    ld      a,(4014h)
0fc5 fe02      cp      02h
0fc7 d8        ret     c
0fc8 fe05      cp      05h
0fca d0        ret     nc
0fcb 218740    ld      hl,4087h
0fce 5f        ld      e,a
0fcf 1600      ld      d,00h
0fd1 19        add     hl,de
0fd2 db30      in      a,(30h)
0fd4 77        ld      (hl),a
0fd5 c9        ret
I will continue to analyze the ROM files.
Let me know if you find any thing wrong about my analysis.
Jack
« Last Edit: February 09, 2017, 01:31:44 pm by z80tw »
  • Country: tw
« Reply #31 on: February 08, 2017, 02:55:59 pm »
After I did couple modifications on the ROM code of PS310 the GPIB function now is enabled! ;D
but I still can't change the GPIB bus address by pressing the "GPIB" key,
when I press the "GPIB" key I still see "Err5" on the display.
The codes I have modified:
1.(pic 1)
00b3h     cd1710    call  0117h    ------> cd301f     call 1f30h
2.(pic 2)
1f30h~1f32h: ff ff ff ----->cd 58 12   call  1258h  (basically I inserted only one command in the program:call 1258h,to initialize GPIB controller )
1f33h~1f35h: ff ff ff ----->cd 17 01   call  0117h
1f36h:ff   ---->c9   RET
After installing new rom code in PS310 I turned on the power.
PS310  powered on and displayed normally,
then I pressed the "GPIB" key but saw "Err5" on display again.
next I tested the GPIB communication.it works!(pic 3)
I issued coulple commands and got correct response from PS310:
*IDN?
VSET=1.0E2(Vset=100V)
VSET?
PS310 response:1.00E2(=100V)
HVON (output turned on)
HVOF (output turned off)
Now I am going to figure out how to make the setting of GPIB bus address works.
PS.current GPIB bus address of PS310 is: 9 (saved in SRAM since I changed it yesterday)
Jack
« Last Edit: February 09, 2017, 01:44:21 pm by z80tw »
  • Country: tw
« Reply #32 on: February 11, 2017, 03:08:20 pm »
Today I found out why I couldn't set GPIB address by keyboard of front panel,because the CPU jumped to "Err5" subroutine
every time I pressed "GPIB" key! in the ROM file of PS350 there is no "Err5" subroutine,instead of "Err5" subroutine there is
a subroutine which will enable GPIB interface and address setting by keyboard of front panel .
Thus I have to replace the "Err5" sub with the GPIB option sub(copied from PS350).The software version modified by me
I called it Version 1.41(original Ver 1.40)(pic 1)
Now GPIB interface of my PS310 works, the GPIB address can be set by keyboard of front panel and saved in SRAM(pic 2,no more "Err5")
How I replaced the "Err5" sub with GPIB option sub from ROM file of PS350?
1.I changed the code of 028fh~0292h:(pic3)
028f~0291   cd 40 1f     call  1f40h
0292             c9             RET
2. Copied GPIB option subroutine of PS350 (029b~02c0)  and pasted it
to address 1f40~1f65h:(pic4)
1f40 af        xor     a
1f41 321740    ld      (4017h),a
1f44 328240    ld      (4082h),a
1f47 328340    ld      (4083h),a
1f4a 3e04      ld      a,04h
1f4c 322440    ld      (4024h),a
1f4f 3e01      ld      a,01h
1f51 321940    ld      (4019h),a
1f54 211240    ld      hl,4012h
1f57 cbee      set     5,(hl)
1f59 3a4c41    ld      a,(414ch)
1f5c e67f      and     7fh
1f5e 322240    ld      (4022h),a
1f61 af        xor     a
1f62 322340    ld      (4023h),a
1f65 c9        ret
Now my problem solved! :-DD
Thanks to all of you who has helped me by providing schematics,ROM code of PS350,hardware and software analysis,
without your helps I could not have fixed this problem.
Jack
« Last Edit: February 13, 2017, 04:19:43 pm by z80tw »
The following users thanked this post: dr.diesel, TheSteve
Offline picburner
« Reply #33 on: February 11, 2017, 04:06:58 pm »
Congratulations, you've done a really good job!
Offline bingo600
« Reply #34 on: February 12, 2017, 07:49:39 am »
Cool detective work  :-+
If you're nice , you'd upload the fixed eprom contents here.
/Bingo
  • Country: tw
« Reply #35 on: February 12, 2017, 10:44:32 am »
/Bingo
Here attached is the modified ROM file,Version 1.41,Model No.PS310,Serial No.2017(<---users have to modify it  according to the serial no of their PS310.)hope it will help other owners of SRS PS-310 / PS-325 who has same problem I faced.
Jack
The following users thanked this post: bingo600
Offline anotherlin
« Reply #36 on: February 13, 2017, 08:43:58 am »
This is not problem solving, but a nice firmware hack! :)
Congrat!
  • Country: tw
« Reply #37 on: February 13, 2017, 02:44:13 pm »
Congrat!
Firmware hack! but without your inspiration I couldn't have done it! >:D
thanks for all efforts you have done for me.
Jack