« on: February 11, 2017, 02:18:40 pm »
My HP6612C, when disconnected from anything shows a current level of about -0.7mA, is this because it needs to be calibrated? The volts hover around 9.5mV (with the output switched off).

I bought it from someone in Korea on eBay and although it does have a relatively recent calibration sticker on it, perhaps the environmental difference between the UK and Korea is causing this? Or is it within spec? I can't find the bit in the manual which tells me...

Amps


Volts
 

Offline sorenkir

  • Regular Contributor
  • *
  • Posts: 146
  • Country: fr
Re: HP6612C - voltage and current offset, does it need calibrating?
« Reply #1 on: February 13, 2017, 08:37:47 am »
Hi,

To me, it seems within spec. See Service Manual http://cp.literature.agilent.com/litweb/pdf/5962-8200.pdf page 23:
Low Voltage (0V) Vout +- 10mV
Low current (0A) Iout +- 1.53mA

Michel.
 