Hi
I'm looking for a HP34401A on ebay and noticed that several of them have dots in the middle after every digit
e.g. zoom in
http://www.ebay.it/itm/122529702180
is it normal or just a symptom of problems with the display driver or other parts of the circuit?
thanks
Offline floobydust
Re: HP34401A display with dots after every digit
« Reply #1 on: June 08, 2017, 06:18:31 am »
I've worked with the HP34401's and never seen that.  Would drive me nuts  |O  You can program the comma separator on/off.
In the eBay ad, looks like segment E shorted to colon? The colon is on with any digit that has segment E on.
I would say trouble with the VFD driver IC or solder bridge on the PCB. NEC uPD7527 to drive the VFD.
Hope I'm wrong?
Offline xemax
« Reply #2 on: June 08, 2017, 08:16:10 am »
You need a new display.
I had two 34401 with the dots problem. After replacing the display they are like new.
« Reply #3 on: June 09, 2017, 04:42:45 am »
Hope I'm wrong?
thanks for your reply
I also noticed that the 4W indicator is on too while it should be off since the multimeter is likely set in V measurement.
So I guess all those multimeters should be considered as defective even if they are sold as in full working condition.
« Reply #4 on: June 09, 2017, 04:45:24 am »
so that is quite a common problem I guess
BTW does the display module integrate the driver circuit?
How much did you spend for those displays and where did you got them?
thanks
Offline xemax
« Reply #5 on: June 09, 2017, 07:43:09 am »
No. Just the display is faulty.
I ordered this:
http://www.ebay.de/itm/1x-New-vacuum-Fluorescence-Display-Display-VFD-For-HP-AGILENT-34401A-/131861099316
The following users thanked this post: not1xor1
« Reply #6 on: June 10, 2017, 02:04:49 pm »
I ordered this:
I think somebody replaced a similar display with alphanumeric LEDs, but do not remember what multimeter model...
I have to check with google
it should be cheaper, but would probably require more work
Online coromonadalix
« Reply #7 on: June 10, 2017, 02:53:25 pm »
it was for a fluke 45 if i recall   a thread is here
https://www.eevblog.com/forum/repair/fluke-45-vfd-display-broken/
« Last Edit: June 10, 2017, 02:55:36 pm by coromonadalix »
Offline Gyro
« Reply #8 on: June 10, 2017, 03:58:53 pm »
I ordered this:
Just the display? That's a bit weird.  Surely that must indicate an internal short (assuming it's not a soldering fault).
If you can work out which pins and are actually able to measure a low impedance path (after removing it), then there ought to be a good chance of clearing it with a bit of current.  :-\
Best Regards, Chris
Offline floobydust
« Reply #9 on: June 10, 2017, 10:15:21 pm »
Being a VFD display fault doesn't make sense to me   ???
Segments are basically a phosphor-coated electrode with a trace out to the external pins.
If we assume electrical short between segments, the NEC VFD driver MCU should have push-pull outputs? Oddly it MCU runs off +13V, -18V, +18V rails.
Offline usagi
« Reply #10 on: June 12, 2017, 07:21:52 pm »
The following users thanked this post: edavid, not1xor1
Offline floobydust
« Reply #11 on: June 12, 2017, 07:47:24 pm »
those dots are caused by leaky VFD drivers. they are TI parts. obsolete and replaced by pericom . i have posted the xref before on the forum.
these are 32 bit high voltage shift registers in plcc44 package. replace em and the problem is solved.
Note : if you have a really old 34401 then the VFD is driven by a upd75xxx custom cpu with integrated drivers . That one is unobtainium....
and the fronts are not interchangeable... a newer 8051 based frontpanel does not work in a machine that needs the upd75xxx based one.
This from that thread conclusion there is the VFD driver IC is leaky- not a VFD display tube problem
free_electron believes it is the VFD driver IC.
xemax says replacing the VFD fixed the problem, and it was the VFD, in another thread below:
https://www.eevblog.com/forum/repair/hp-34401a-dmm-with-leaking-segments/
https://www.eevblog.com/forum/repair/hp-34401a/
https://www.eevblog.com/forum/buysellwanted/ebay-many-faulty-hpagilent-34401a-multimers/msg811170/#msg811170/
« Last Edit: May 17, 2021, 07:57:59 pm by floobydust »