I wonder if there's a way to read them out via the GP-IB?
In short, no, not in a way that I would rely on.
There is a GPIB command called "CAL DUMP" which contains most, but not all, of the needed cal constants for backup along with a pile of other data which I don't see any documentation for. I also don't see a GPIB command to restore the data to the SA from this dump format.
I would stick with the filling-in-the-chart method, since that's also used in the procedure to restore the constants. It doesn't take very long, otherwise you're in uncharted territory with your cal backup.
The pun, it burns!
Or I could just change the battery quickly, hoping nothing goes wrong.
Just kidding. Charts it is.
Re incomplete GPIB 'CAL DUMP'. Gee, it's almost like they deliberately laid a trap, knowing some users would assume CAL DUMP did something useful. Then would end up sending the instrument back for some very expensive HP attention, after they found they didn't actually have a workable calibration backup. But no, I can't imagine a corporation doing something sneaky and profit motivated like that. (Yes I can.)
You could do CAL DUMP as a secondary method to make sure you have the numbers written down right, since most of the numbers are embedded in there, or the time-tested method of taking a video of the screen while you step through the constants if you're feeling lazy.
I'm a big fan of reliable, never-fail paper records.
8590E Series Programmer Guide:
http://literature.cdn.keysight.com/litweb/pdf/08590-90235.pdf
Yesterday I spent over an hour downloading almost everything on the keysight pages for the 8590. Even ones with no relevance to my interests. They are a trove of RF practice.
And so I already have that pdf, thanks. Plus joyfully, an original paper manual among those coming with the instrument. Ha, it's still in the original shrink-wrap.
All currently in a truck driving over the Rockies, I gather. Am praying to the gods of foam-in-place.
Speaking of phase noise, here's a list of the HP Application notes I have. In paper originals. Anyone with a stack of them they were about to downsize, please let me know.
Some are relevant to spectrum analyzers (and phase noise.) Now I have (touch wood) something to play with, it'll be worth reading them.
Can anyone recommend other good reads on care & use of spectrum analyzers?
Hewlett Packard application notes
H prefix = 'I have physical copy'
> prefix = 'ordered physical copy'
p prefix = I have a pdf copy
#4 The PIN diode
#52 Frequency and time standards
H #57 Noise Figure Primer 1965
#57-1 Fundamentals of RF and Microwave Noise Figure Measurements 1983 8970A
#57-1 Agilent Fundamentals of RF and Microwave Noise Figure Measurements 2000 859X-NFA
#57-2 Noise Figure Measurement Accuracy 1988 8970B
#57-2 Noise Figure Measurement Accuracy - The Y-Factor Method 2001 8590X-346X
#57-3 10 Hints for Making Successful Noise Figure Measurements 2000
H #63 Spectrum Analysis 1968
#63A More on Spectrum Analysis
#64 Microwave Power Measurement
H #64-1 Fundamentals of RF and microwave power measurements
> #64-2 Extended application of automatic power meters 380442352194
H #64-3 Accurate and automatic Noise Figure Measurements 1980
#65 Swept Frequency Techniques, printed Aug. 1965
H #73 Calibration of a Gamma Ray Spectrometer
H #77-1 Transistor Parameter Measurements
#77-2 Precision Frequency Comparison
H #77-3 Measurement of Complex Impedance
#77-4 Swept Frequency Group Delay measurement
#83 Increased Output Resistance for DC Regulated Power Supplies
#84 Swept SWR Measurements in Coax
#86 Using Vector Impedance meters
H #89 Magnetic tape recording handbook
#90 DC POWER SUPPLY HANDBOOK
H #90B DC Power supply handbook (book)
H #91 How vector measurements expand design capabilities
H #92 Network Analysis at Microwave Freq !!!! uses 8410.
#94 Connector design employing TDR techniques
#95 S-Parameters - circuit analysis and design
#95-1 S Parameters Techniques for faster design
#99 8541A Automatic Network Analyzer Measurements capabilities
#102 Program controllers
#107 Guard Circuits
#112-2 Using 675/675A Network analyz as an education tool
#114 A2A Video Transmission System Alignment 1969
#116 Precision Frequency Measurements
H #117-1 Microwave network analyzer applications (book) (uses 8410) x3
#117-2 Stripline component measurement with 8410a
#122 EMI Measurement Procedure. Calibration & Operation of the HP 851/ 8551B
Spectrum Analyzer to Measure Electromagnetic Interference
H #123 Floating Measurements and Guarding, published June 1970.
H #124 True RMS measurements
H #126 Theory and Applications of Wave Analyzers, published November 1970. x2
H #136 Understanding & Operating the 8555A Spectrum Analyzer and 8445A Preselector. 1971
H #139 Stabilizing Gamma-Ray Spectrometer Systems 260571717091
#140-0 Fourier Analyzer Training Manual
H #144 Understanding Microwave Frequency Measurements
H #150 Spectrum Analysis... Spectrum Analyzer Basics (Uses HP 141T)
#150B Spectrum Analysis... Using the 8557A and 8558B Spectrum Analyzers. pub Feb 1975.
H #150-1 Spectrum Analysis... Amplitude and Frequency Modulation. Nov 1971
H #150-2 Spectrum Analysis... Pulsed RF. (uses 141T) (slide rule in back pocket)
H #150-3 Swept frequency measurements and selective frequency counting with a tracking generator. (uses 141T)
#150-4 Spectrum analysis... Noise measurements Apr 1974
H #150-5 Spectrum Analysis... CRT Photography and X-Y Recording Techniques
#150-6 Spectrum Analysis... CATV Proof of Performance
#150-7 Spectrum Analysis... Signal Enhancement
#150-8 Spectrum Analysis... Accuracy Improvement
#150-9 Spectrum Analysis... Noise Figure Measurement
#150-9 Spectrum Analysis... Field Strength Measurement
#150-10 Spectrum Analysis... Distortion Measurement
#150-11 Spectrum Analysis... Distortion measurements (uses 141T)
H #150-12 Spectrum Analysis... Using the HP 11517A External Mixer to 40GHz Nov 1977.
H #150-13 Spectrum Analysis... Using the HP 8565A Spectrum Analyzer from 2-18GHz Aug 1978.
#150-14 Spectrum Analysis... Using External Waveguide Mixers Above 40 GHz
#155-1 Active Device Measurements with the HP 8755 Frequency Response Test Set, printed Aug. 1978
#155-2 100 dB Dynamic Range Measurements Using the HP 8755 Frequency Response Test Set, printed Sept. 1977
#155-3 Automating the HP 8755 Scalar Network Analyzer, printed Jan. 1981
#158 Selecting The Right DVM
H #162-1 Time interval averaging
H #163-1 techniques of digital troubleshooting
H #163-2 The IC troubleshooters - New techniques of digital troubleshooting
H #167-5 Trouble shooting in the data domain is simplified by logic analyzers
H #167-6 Mapping, a dynamic display of digital system operation
H #167-7 Supplementary data from map displays without changing probes
#167-9 Functional analysis of the Motorola M6800 uP system
H #171-1 HP 8640A/B & HP 8405A Application note - crystal testing (1923820255)
H #173 Recent advances in pulsed RF and microwave frequency measurements
#173-1 Dynamic measurement of microwave voltage controled oscillators with the 5345A
#174-1 Transfer Characteristic of a VCO
H #174-3 5345 counter. Measuring Integral Nonlinearity of a VCO
#174-4 5345 counter. Measuring dual VCO tracking error.
H #174-6 5345 counter. Measuring the Stability of a Freq Source
#174-8 5345 Measuring FM Peak-to-Peak deviation
H #174-10 5345 Electronic counter - meas. electrical length of cables
#174-12 5345 Measuring frequency sweep linearity of sweep generators
H #175-1 Differential phase and gain at work
H #183 High freq swept measurements.
#185 Waveform Parameter Measurements using the Mircoprocessor Controlled
Oscilloscope Model 1722A. Dec 1974
#185-2 Transmission line matching and length measurements using
dual delayed sweep in the uP controlled scope Model 1722A
#185-4 Elimination of computation on analog measurements by using the
direct reading oscilloscope model 1722A
#187-3 Three HP-IB Configurations for Making Microwave Scalar Measurements
H #191-6 Precision time interval generation & measurement apps library
H #196 Automated measurements using the 436A pwer meter
#200 Fundamentals of the electronic counters
H #200-1 Fundamentals of microwave frequency counters
H #200-2 Fundamentals of Quartz oscillators
#200-4 Understanding frequency counter specifications
#201-2 Differential Non-Linearity of a VCO
#205-1 Low Freq Amp Considerations of 3042A
H #207 Understanding and measuring phase noise in the freq domain
#216 HP Digital Network and Spectrum Analysis primer
A guide to the use of the HP3570A/71A Analyzers
H #221A Automating the 8410B Microwave Network Analyzer (won 2527060511) in 8409B binder
H #222 A designer's guide to signature analysis
H #225 Measuring phase spectral density of synthesized signal sources ...
with the 5390A frequency stability analyzer
H #240-0 Digital signal analysis - time & freq domain measurements
H #243 The Fundamentals of Signal Analysis
H #243-1 Dynamic signal analyzer applications
H #243-2 Control system development using dynamic signal analyzers
H #243-3 The fundamentals of modal testing
#246-1 Optimizing the dynamic range of the HP 3585A spectrum analyzer
H #262 Eliminating time-base errors from oscilloscope measurements (HP 1725A/1727A)
H #286-1 Applications and operation of the 8901A modulation analyzer
#286-2 Accurate Mixer/Amplifier compression measurement using the 8901A modulation analyzer
H #292 Minicomputer analysis techniques using logic analyzers
H #297-1 8161A programmable pulse generator
p #319 Parametric characterization of digital circuits up to 50MHz
with the 8180A/8182A stimulus.response system.
H #358-1 HP 5371A frequency and Time interval analyzer
H #372-1 Power supply testing
#401-2 Programming Guide for the 59307A
#401-3 5345A/1000 Programming Example
H #401-13 HP 3325A Synthesizer/Function Generator and 1000 Computer Programmnig Guide 1979
#916 HPA GaSa Sources
#920 HARMONIC GENERATION USING STEP RECOVERY DIODES &SRD MODULES
#1304-2 Time Domain Reflectometry Theory for use with HP 54750A and HP 83480A Mainframes, printed 1988
Misc
H 1987 Tutorial description of the HP-IB interface
HP Slide rule calculators (cardboard)
H HP REFLECTOMETER CALCULATOR dB to VSWR error limits
A Hewlett-Packard reflectometer calculator / mismatch error limits calculator.
It is used to convert return loss (in dB) to (VSWR) Voltage Standing Wave Ratio
and also calculates mismatch error limits. For more information reference
HP Application Note 183 "High Frequency Swept Measurements".
Especially useful for microwave work when using directional couplers for return
loss. This calculator has been out of print for many years and is hard to find.
H HP Vector Impedance calculator / Capacitance - Inductance - Reactance calculator.
For use with HP 4815A Vector Impedance Meter.
H Pulsed RF Calculator. 1971.
In pocket at back of HP App Note 150-2, Spectrum Analysis... Pulsed RF.
BOOKS
H Microwave Theory and Measurements
Hewlett-Packard Company Engineering Staff of the Microwave Division
Bought 20120419 via abebooks.com $10 + $15 postage
H Phase noise. RF & Microwave Phase noise Measureent Seminar HP.
(A photocopy in blue ring spine w peach covers. 118 pages.)
Others
H Feeling comfortable with Logic Analyzers. 5954-2686 Printed April 1988.
References & sources
http://www.hpmemory.org/news/an150/an150_page_00.htmhttp://www.hparchive.com/Application_Notes/HP-App-Note-Index-1967-Revised.pdfhttp://mmhewpack.free.fr/ressources/resrc_an_01.htmhttp://www.hpmemory.org/ressources/resrc_an_01.htmI'd like to expand this to a complete listing of all HP app notes ever published.
Actually collecting all these old App Notes is probably an unachievable ambition. But I'd like to.
The ultimate goal being to scan them all in the high visual quality and display technology they deserve,
as opposed to the often very poor quality scans as pdf that can be found on the net now.