HP 8482A -HP 5347A
« on: Today at 11:21:34 am »
Hello, colleagues. Please advise me if I can use the HP8482A sensor for the HP 5347A counter. Thank you very much!
 

Offline Tony_G

  • Frequent Contributor
  • **
  • Posts: 951
  • Country: us
  • Checkout my old test gear channel (link in sig)
    • TGSoapbox
Re: HP 8482A -HP 5347A
« Reply #1 on: Today at 03:30:13 pm »
You should be able to - I suspect that the reason it isn't listed in the manual is that the 5347A is a "20Ghz instrument" and the 8482A is only a 4GHz sensor (from memory, somewhere around that as compared to the 8481A which goes to 18GHz).

The connection is the same but I haven't used the 5347A (I think I have a similar instrument around here somewhere but I don't use it) so there might be something specific about the calibration factor entry but that should be fairly easy to overcome by just putting in a temp value if it needs all them to run.

TonyG