HP 70601A RF ( 0...26,5GHz )
HP 70906A Preselector ( 50k...26,5GHz
and 70900A, 70903A, 70902A, 70004A 70311A
I intend to improve its parameters by adding HP 70620A (Preamplifier 2-22GHz).
My question is whether my HP 70k will work with this type of preamplifier (in its operating range 2..22GHz) ?
Marian, I would think it would at least partially work, but your readings could possibly be a bit off, depending on how the 70900 deals with the preamp being in the system. There's not documentation on non-standard configs like that, so it's mostly trial and error to see how it will behave. In general though, the system usually "tries" to work even with non-standard setups, but you sometimes get strange or off behavior.
The 70620 sends gain/flatness information to the 70900 so it can give you an accurate amplitude reading over the full frequency range. Whether or not the 70900 interprets that data correctly in the frequencies below 22GHz with a 70620A depends on how that data is used vs how it is done with the 70620B. I have a feeling it will work correctly, but I have not tried it.
I have a 70908A 22GHz RF section, and a 70600A 22GHz preselector that was intended to go with a 70905A RF section. I tried putting the 70600A in the chassis with my 70908, hoping that I could use the 70600A preselector independently via the tune voltage input. As I mentioned above, the system "tried" to work, but it wasn't a useful result. The 70900B saw the unit, and did not complain about it, but IIRC it blanked the trace below 3 GHz and above a certain range (maybe for the first band of the preselector---I don't remember the details exactly). The tune voltage did control the preselector correctly within one band, but the 70900 would not make it switch bands, since it wasn't paired with a 70905 input section, even though my 70908A is also 22GHz.
But I do not believe there are any band changes or anything like that with the 70620A preamp, just the calibration data as I mentioned above, so it is likely to work, and in the worst case probably just give some amplitude error (the microwave preamp is spec'd at +/-2.8dB over the range from 2-22GHz.) If it did show amplitude errors, you could measure the gain at various frequencies yourself with another instrument, and print out a correction table, or even generate a correction line that you could load in the analyzer and add to the trace to make it display accurately. Give it a shot and let us know how it goes!
![Cheesy :D](https://www.eevblog.com/forum/Smileys/default/cheesy.gif)
Edit: Also, make sure your 70900A has a firmware version late enough (900314 or later) to work with the 70620A.