OK, I opened the K181, gave it a good visual inspection, Visually, all looked good, nothing burnt or broken, soldering looked ok and showed no signs of rework, although a previous owner has been inside, either to troubleshoot or calibrate ( front panel has some plastic pieces broken off and a few screws appear to be non original )
Next I measured the supply voltages: The +15 volt supply measured +14.85 and the -15 volt supply measured -15.30. I then used my 88A esr meter and checked all the electrolytic caps. I also checked the titanium caps on the low voltage input board for shorts / opens and all is well there too. I used my IR thermometer to check if anything is overheating, nothing over 90 deg. F.
I then hooked up a 15 volt variable power supply to see if the K181 would measure anything or give any reading other than displying OFLO using only the V input binding post front panel input. I applied various voltages and only had OFLO displayed on all 4 ,2V, 20V, 200V, and 1000V ranges.
Now for something interesting: While applying voltage while on the 2 volt range, I noticed that if I exceeded 2 volts, the display started going back and forth between OFLO and 000000. It would not display the input voltage being applied , it displayed 00000 ! and OFLO . As I mentioned, the LED on the Zeroing switch will not stay on, but sometimes does after 20 or more attempts.
I am thinking that the problem is related to an input board related problem. A previous owner most likely blew something on that board. I will doing some further checks to see if I can find a problem there. Another observation is that input board 2 screws that hold it down are slotted non originals which I believe should be philips type screws. Looks like a previous owner also removed this board at some time.
Any ideas , comments and /or suggestions would be greatly appreciated, Thanks to everyone !