Hi group,
I am looking for the Master Disk for a HP 4155B Semiconductor Parameter Analyzer. When I got the equipment it was displaying garbage on the LCD screen. I reseated one of the display connectors and now it shows:
Can anybody supply an image of the disk?
Thank you for your help.
Jay_Diddy_B
Offline photon
Re: HP 4155B Semiconductor Parameter Analyzer - SMU - Need Master Disk
« Reply #1 on: July 19, 2015, 07:21:04 am »
Hi Jay_Diddy_B,
I am not sure this is what you need. Just in case you haven't looked at it yet...
http://www.keysight.com/main/software.jspx?ckey=213376&lc=eng&cc=US&nid=-11143.0.00&id=213376
« Reply #2 on: July 19, 2015, 02:35:21 pm »
Hi Jay_Diddy_B,
Thank you very much. I hadn't realized that the same firmware is used on the B and C models.
I downloaded the zipfile from Keysight. Copied the files onto four floppies. Installed the firmware into the machine and now I have a working unit.
Here is the configuration screen after installing the firmware:
Here is an IV plot of a green LED:
I will have to work how to make the test fixtures.
Thank you again!!
Regards,
Jay_Diddy_B
« Last Edit: July 19, 2015, 02:45:55 pm by Jay_Diddy_B »