Author Topic: CMU200 erratic levels outside 1200-2200 MHz  (Read 1648 times)
Offline digitalrubbishTopic starter
« on: February 24, 2016, 12:56:11 pm »
Dear friends
I acquired a CMU200 for my HAM radio lab.  The unit however has an issue.  The generator works fine but the spectrum analyser shows about 16 dB too low from 0 till 1200 MHz and above 2200 MHz.  Between 1200 and 2200 all is fine.
Can anybody give me a hint what component could cause this.  I am not afraid of digging into the boards as R&S repair service is way beyond my budget.
Thanks for all help and schematics of the input boards or mixer boards where obviously this fault is located.
Pedro Wyns  ON7WP-AA9HX
direct mail to pedro.wyns@  the famous gmail dot com
Offline _Wim_
Re: CMU200 erratic levels outside 1200-2200 MHz
« Reply #1 on: February 24, 2016, 06:21:53 pm »
Just send you the service manual. A few questions:
- do you have the same results on both inputs?
- did you do a factory reset off all settings to be sure no correction table is loaded for the inputs (you can see this under "Connect Control" => "Ext. Att. Input" => "RX Connection Table"
« Reply #2 on: February 26, 2016, 04:20:31 pm »
I have this on all 3 inputs.
no correction table
