There is no information on the construction of a 5.5 digit multimeter manufactured by Hantek, model HDM3055B. Therefore, I will share what is known about and what can be assumed. The purchase of such a meter is not cheap, the meter with delivery and tax fees for Poland is about 545 USD. For the amount of 639 USD you can buy the Siglent SDM3055 model. Does the 94USD difference pay off? Let's see ...
The multimeter has 2 PCBs. The analog part is the largest and occupies the entire bottom of the housing. The display part is mounted on the front panel and connected with the analog part by a connector. An interesting solution because it requires a mechanical fit and is comfortable. On the front panel PCB, we can find, among others:
- U412 MCIMX6Y2DVM05AB microcontroller (Arm Cortex-A7 528MHz)
- U410 DDR3L SDRAM 1Gb memory designated D9SFT (Micron MT41K64M16TW-107)
- U8 2G-Bit 3.3V Nand Flash Memory (W29N02GVBIAA)
- U1 STM32F030K6T6
- U2 RTC (AnalogTek company AT8307T , equivalent to DS1307)
- U417 EEPROM (24LC64 designated 4L64I)
Unfortunately, the problem of PCB cleanliness is already visible here. U1 and the capacitors at the stabilizer were soldered again. It can happen in production but can't it be washed after this process? It looks nasty.
The housing is solid, made of thick sheet metal and properly painted. All bolt fasteners are by welding nuts. The most important part in terms of measurement accuracy looks like this:
Here is also a problem with soldering.
The PCB has a pinched screen over the analog part. It covers a large area of the PCB.
And at that moment a practice that I do not understand begins. Namely, several integrated circuits have mechanically removed markings. Interestingly, later painted black ... After all, it is not a problem for electronics to find out which integrated circuit it is, or at least from which group. And so, most importantly:
- the U22 reference voltage system has no markings removed and it is ADR03A
- the ADC converter (U24) has the markings removed, in my opinion it is ADS1255
- U23 temperature tgemperature sensor is LM75
- RMS U1 converter, no markings, matches LTC1966/67/68
- a lot of analog keys from the DGxxx series
- you can see some guard rings
- U3 and U18 no markings, I don't know what it is, U18 is probably an opamp - there are a lot of passive elements (capacitors) on the bottom of the PCB, I did not take it out because there is a lot of work on it
- a fairly good ADDA fan, type AD0412LB-G70 (12V), turns on for up to 2 seconds only during boot
A multimeter with a measuring range of 100mV, 1V and 10V can have a selected input resistance of 10M or high. The thing that worries me is the lack of a precise 1776 series Caddock divider. I also can't find a divider for> 10V voltage with a resistance of 10M. I know that this divider is switched on by a relay. By the way, it's even good that there are only 3. On the bottom there is a 0R002 measuring resistor in the Kelvin circuit for the 10A range. It's such quick facts.