The power supply voltage (in this case, the laptop battery at 19V) seems a reasonable worst case voltage that is likely to appear on the pads in a worst case fault condition, assuming you don't do anything crazy that can step that voltage up even higher.
Assuming you actually do have a technical/scientific foundation for your argument, you are failing miserably to present it in an understandable way.
And the problem is, 19V is lethal under certain fault conditions for a Type CF Applied Part (IEC 60601), which is what you have.
Let me repeat that in more blunt terms, with emphasis: YOU CAN STILL KILL SOMEONE IF YOU SCREW THIS UP BADLY ENOUGH.
Look. I do this for a living. You can not and should not under any circumstances put that eval kit on a patient. You might put it on yourself -- we have done that before. And in that case, we were literally in the same building as the nearest hospital, knew where the two nearest defibrillators were, had skilled staff monitoring the whole time (both engineers and first responders), nothing was anywhere near the heart, proper isolation was observed at all times, the person with the electrodes on volunteered themself, and we still thought it was crazy dangerous and there was no way we would ever, ever, ever ask one of our research patients to do it.
Face it. Eval kits don't go on patients. Period. End of bloody discussion.
Best to stop this nonsense now before your IRB sees any of it.
We are having a violent agreement... 19V is too much, so a battery powered laptop is not a guarantee of safety, which is the point I have been trying to make.
We are also violently agreeing that no patient / consumer should EVER be exposed to a medical electronic device that has not been designed, tested, and approved according to all relevant regulations and best practices.
With that out of the way, I don't see why the subject of medical electronics as a hobby should be banned completely from any discussion - why is it worse than making Tesla coils, huge capacitor banks, X-ray machines, microwave beams, laser cutters, or any other potentially dangerous project? Why is it worse than driving a 700hp muscle car or riding a turbo motorcycle? If Darwin has your number, you will have a bad day no matter what your hobby, pretty much.
If you want to play, you have to understand the risks and know what you are doing. Personally, I did not know (until this thread) that the safe limit of current through the body can be as low as 50uA. It has increased my respect for electricity vs biology, not decreased it... anyone thinking of playing with this kind of stuff needs to know those limits.