Here's a demonstration, what an SDS2000X Plus can show with an emulated ripple with asynchronous spikes riding on it if optimal probing (without additional noise pickup) is applied. 2.5 mVpp 1 MHz ramp with 300 µV 6.000001 MHz 10 ns wide spikes riding on it.
Nice example (pic worth a thousand words). How did you generate such signal?
It is a two channel AWG (SDG6052X in this particular case), both channels combined by means of a resistive wideband (DC-12.4 GHz) power combiner/splitter. This will provide 6 dB attenuation for each channel.
Channel 1 generates a 5 mVpp ramp with 5% symmetry at 1 MHz.
Channel 2 generates a 10 ns wide pulse with 3 ns rise/fall times at 6.000001 MHz with 60 mVpp amplitude and this signal is fed through a 1 GHz precision step attenuator, set at 40 dB.
The output of the power combiner is what you see.
First thing, HiRes on MSO5000 has been around for some time now, and it's not big help...
They had to downsample 32 times to get something and then not much.
The manual says that the
bandwidth cannot exceed 1/32 the samplerate, so this sounds like 16 times oversampling to me.
This would be roughly equivalent to ERES 2.0 bits on a Siglent, where the resolution enhancement in ENOB can be set from 0.5 to 3.0.
In 10 bit mode together with ERES 2.0, we get a total resolution enhancement to 14 bits (and a theoretical ENOB of 11 bits). The trace gets very thin with this, see attached screenshot.
SDS2354X Plus_Ramp_2.5mV_1M_Pulse_300uV_6000001Hz_ERES2.0
I use two two pass trough terminators to 1k resistors from siggen, other side connected together, and grounded with 100 Ohm resistor. I have that one because we had some previous discussion about two tone testing, and someone did it with that one so I made it exactly like that so we can compare results...
That might have been me since I was interested in the scopes DR performance. The 1K series R is to isolate the two AWG outputs from each other so they don't "see" the other signal as much with the ~26dB reverse isolation looking back from the shunt 100 ohm resistor. Was concerned about how the AWG output behaves in the presence of another signal and how this might affect the AWG output amp linearity. With this high an isolation the AWG output amplifier effects should be minimal and the resultant two tone IDM representing the scopes performance. BTW using the digital combining isn't as good, since the signal is created and then passed thru the AWG amplifier chain to the output, thus exposing the signal to the chains linearity effects. However may be OK since we are only looking at 65~75dB IMD with these scopes, but with Performa01 Picoscope results of ~100dB some of the AWG amplifier chain effects may contribute if using the AWG digital combining for the two tones.
For the high 3rd order dynamic range on the Picoscope 4262 I did not need to use any additional attenuators apart from the power combiner itself, which only provides 6 dB isolation between its ports. At frequencies as low as 1 MHz, an SDG6000 amplifier output quite obviously cannot be intermodulated that easily. But maybe I should try one more time with additional attenuation – who knows, maybe the Picoscope can do even better…
EDIT: Done - and found no difference. The SDG6052X performs well enough for this test at 1 MHz, even without additional isolation between the outputs.
The integrated digital combiner does indeed generate some IMD products on its own, so it is not suitable for reliable 3rd order dynamic tests. IMD can be as bad as 45 dBc at high output levels.
EDIT: Using the internal digital combiner, the third order intermodulation intercept point is about +21 dBm at an output level of +9 dBm and a frequency of ~50 MHz.
EDIT2: Unsurprisingly (since it's done in the digital domain), the third order intermodulation intercept point is nearly the same (+21.5 dBm) for a -18 dBm output level at 1 MHz.
I got a box delivered yesterday and I find that the SDS2000X+ is only very slightly better--if at all--than the SDS1104X-E as far as noise and minimum trigger. It isn't able to match the Tek 2221A, but it gets pretty close--200uVrms vs 150uVrms for a stable display.
There is almost no difference indeed. If anything, the SDS2000X Plus might show less or lower spurious signals, but the noise level should be about the same. In fact, the little SDS1000X-E even has a slight edge here, because it provides a true full resolution 500 µV/div sensitivity, whereas the SDS2000X Plus is limited to 1 mV/div. Yet the 500 µV/div gain setting is not useless there either, because of the 16 bit display interface. This means you can make use of it in 10 bit mode (which becomes 9 bit at 500 µV/div) and also the results of math functions, especially ERES of course.