You can't have any meter that measures precise esr or capacitance in circuit. you can only be close to.
I check the esr with Jay_diddy esr meter adapter I made with a digital display. only for magnitude order
if I read a high esr , then I grab the de5000 to measure capacitance, which can be close but never the real one
if the esr is too high, and the capacitance not the one marked on the cap I'm testing, then I desoldier the cap and measure it with the de5000
then only I get a precise value of the cap.
the in circuit method is there to narrow the bad caps, not to measure precisely.
I could do all of this with the de5000, but the jay_diddy esr meter is protected for charged caps, not the de5000...