Since mnem got a new model with the gimmick capacitor on the input, I decided to order another one to compare.
The 2 probes. I replaced the original pogopin and used shorter one I had.
![](https://www.eevblog.com/forum/testgear/hi-z-probe-for-50-ohm-spectrum-analyzer/?action=dlattach;attach=1263481;image)
![](https://www.eevblog.com/forum/testgear/hi-z-probe-for-50-ohm-spectrum-analyzer/?action=dlattach;attach=1263487;image)
I changed my setup a little bit to be able to measure S21 on top of S11. The VNA has been previously calibrated with the small test PCB and the probe is connected near the load.
![](https://www.eevblog.com/forum/testgear/hi-z-probe-for-50-ohm-spectrum-analyzer/?action=dlattach;attach=1263493;image)
Old probe (yellow) VS the New one (blue):
![](https://www.eevblog.com/forum/testgear/hi-z-probe-for-50-ohm-spectrum-analyzer/?action=dlattach;attach=1263499;image)
New probe (yellow) VS the Old one (blue):
![](https://www.eevblog.com/forum/testgear/hi-z-probe-for-50-ohm-spectrum-analyzer/?action=dlattach;attach=1263505;image)
Overall, the new one is an improvement. Perform relatively well up to 1G and input capacitance is a lot lower (1.2 to 1.5pF compared to 3.5 to 8.5pF for the old one). The gain is unfortunately not linear and starting at -25dB (and not -20dB as mentioned in the Elektor article).