The main fuse is in spec: 250mA.
Disconnected secondary side of transformer.
Resistance is
primary:
1-2 240 Ohm
2-3 0 Ohm
3-4 270 Ohm
secondary:
5-6 31 Ohm
6-7 34 Ohm
Voltage selector is OK:
1-4 is 230V
Secondary Voltage is
5-6: 44 V RMS
6-7: 44 V RMS
Disconnected Trafo stays cold.
Scope with differential probe shows same voltages ...
As R52 and R53 were getting hot, I followed David's (factory) advice and checked C29 (orange).
It's 2.2µF bang on and not shorted.
With connected transformer and seperated +15V and -15V lines between PSU and amp,
I put a 2k resistor as a load in between.
Output is +14.6 and -16.2V.
Trafo stays cold.
But R2 and R3 (yellow) are getting hot quickly (100°C).
Next step is to better understand this PSU.
And to put an lab PSU at the amp and see, what happens.