Regarding triggering on a pattern: in most cases you don't know what the problem is so you can't really trigger on something you don't know.
Decoding the full memory can give you a list with all decoded messages in an acquisition. This usually can be saved to a CSV file which can be processed on a computer to (for example) find specific or malformed messages quickly using search features of Excel (or a similar program).
Ah, so you are saying you don't scan for errors in the oscilloscope screen, but instead do post-processing on a PC. That is what I would think it is easier. In theory the Rigol could do the same if you perform a "zoom out", which will allow the decoder to work, and then save all this data. I will try this at a later time.
This is where having two instruments or a mixed signal instrument can be nice. Use the scope to check for signal integrity, use the logic analyzer to capture and decode the data. In most cases you don't really need to know what the data is in order to see that it's garbled, distorted by ringing, noise or of the wrong amplitude.
That is always my discomfort of having protocol decoding on a pure waveform oscilloscope - is this simply trying to shoehorn a non-target functionality into an instrument designed for another purpose? Obviously that the more features the merrier, but if they negatively impact the core functionality of the equipment than it should be left to more specialized brethren in the product line. All in all, I think it is already a feat on a very low cost to account for such advanced functionality, let it be on screen or on the buffer.
Usually, when you're looking at decoded data in such a way that you're interested in more than just what's on the screen at the time, you're doing so with the scope stopped. There's no good reason for not decoding the entire acquisition under those conditions, and there's nothing that says that you have to compromise the waveform update rate for it. I can understand why you might want to limit decoding to the displayed portion of the waveform during capture, particularly if your trigger conditions don't involve that decoding, but that's no longer valid once the scope is stopped.
Ok, so I agree with you in theory but, as I mentioned before, if the decoding process takes too long to scan through the entire buffer due to processing limitations, then I suspect the UI may be perceived as "slow" (you never know, it is hard to please everyone
![Smiley :)](https://www.eevblog.com/forum/Smileys/default/smiley.gif)
). Obviously that you can get a more powerful oscilloscope with an Intel processor in it, but that is certainly not entry level anymore.
I can only speak for the Rigol scope but recently I had to debug a bit banged SPI flash ROM programming system. The problem with the Rigol is even though you can capture a lot of the SPI data and zoom in on it, as soon as the start frame (ie nCS goes low) moves off the screen as you pan through the data the decoding switches off which kind of defeats the purpose.
BTW People using the Rigol for protocol analysis I found the delayed triggering and zoom was much more useful than changing the timebase and panning the data contents using the horizontal position control because it allows you to trigger at the left side of the display and zoom in around the trigger point much more easily ![Wink ;)](https://www.eevblog.com/forum/Smileys/default/wink.gif)
Oh, so I see you already tried this with the Rigol. I will see what happens with the method you described on the DS4000. Thanks for sharing the procedure!
(me? I'd do it by sending a short message over and over and using pass/fail - a longer serial decode wouldn't make any difference to that)
Unfortunately that doesn't always work. Sometimes it is a specific message which can cause a device to misbehave every now and then. At some point I had to deal with a problem where a message went wrong once every 30 minutes to an hour.
I agree with nctnico here - sometimes the sequence of bits on the wire is what triggers and error (transmission losses, ringing, etc.), not the line itself.
(couldn't yet read the rest of the post. Thanks for the discussion and OP, please apologize for hijacking the thread.)