Active probes should be connected to a resistive load, like a 50 Ohm scope input. The 50 Ohm source should not require any compensation for this. The tip will be a fairly reactive high-impedance load, so any divider in front of it will require compensation.
I assume you're not just randomly trying things and that you downloaded the manual and followed the alignment procedure in section 5. If that is what you are doing, what equipment was substituted? Could you post a picture of your test setup?
What kind of tip is it? Do you require compensation for other tips? None of my active FET probes, although they're not Tek, have required any adjustment or compensation, and auto zero when used with the scope.
If they are normally supposed to be used with some other Oscilloscope, which perhaps has auto or manual probe compensation adjustment on the 'scope itself, perhaps they might not be a good choice for, what from the pix, looks to be a 7000 series analog Tek.
The compensation output is intended for low frequency compensation which is all that is required for an attenuator. The peak and valley shown are a higher frequency component which is not meant to be used. Do they show up with other probes?