Poll
How many cycles will the KeySight U1281A's detent spring last?
0-2000
7 (16.3%)
2k-4k
5 (11.6%)
4k-8k
16 (37.2%)
8k-16k
8 (18.6%)
>16k (most rubust meter ever made)
7 (16.3%)
Total Members Voted: 40
Author Topic: Handheld meter robustness testing  (Read 1261985 times)
0 Members and 52 Guests are viewing this topic.
Online rsjsouza
  • Eternally curious
    • Vbe - vídeo blog eletrônico
Re: Handheld meter electrical robustness testing.
« Reply #2400 on: March 04, 2018, 12:54:40 pm »
MOVs would give up if limiar overvoltage is applied for too long - the circuitry could probably withstand 1.1~1.2kV, but the MOVs would suffer thermal stress. All in all, I don't think any meter would be too different in this regard (unless they used higher voltage MOVs).
I've had people suggest similar things but it really makes no sense to me.  Maybe you could explain why you feel this way.  What is causing all of this thermal stress you mention?
If a MOV rated (loose definition of "rated" here) for 1kV is subjected to, say, 1.2kV, it will try to clamp this voltage differential using the only means it knows: by dissipating the surplus energy through its body. But that you already know.
In the scenario above, considering the output impedance of the source is low enough, the amount of the surplus energy that needs to be dissipated depends on the waveform. In a single pulse transient, the extra energy is perfectly contained with minor (if at all) stress to the MOV. On the other hand, a 1.2kVAC at 50 or 60 Hz will demand the MOV to continuously dissipate the energy contained in the upper and lower cycles of the sinewave. By the same logic and what was reported in the thread I mentioned, 1.2kVDC is the worst scenario as there is no time for the MOV to cool.
In the AC scenario (and to a much lesser extent to DC), the survivabilty of the MOV is highly dependent on its physical characteristics, as well as its environmental (temperature and humidity) and the surrounding heatsink ability of its PCB (large copper areas, clearance, etc.). That is why I mentioned that most (if not all) DMMs would have the same outcome as reported in the linked thread.
Offline WhichEnt2
  • Country: ru
« Reply #2401 on: March 04, 2018, 01:27:22 pm »
They are changed 2 smd caps above the "fluke and linear" branded IC. On your video small caps are installed, but on early teardown photos they use big caps.
In my 87V which is manufactured at feb 2017 they are also small.
Short pieces, high value, small period, huge amount, long delay.
Offline joeqsmithTopic starter
« Reply #2402 on: March 04, 2018, 02:37:03 pm »
Because I am currently looking at the 87V,  let's just use it for an example.  A few things to consider.  The one MOV has a 1Meg resistor in series with it.  Even with 2KV applied, will limit the current to 2mA assuming the MOVs were shorted.    The second leg uses a PTC and surge rated resistor for the drive side.  With enough DC the PTC would eventually kick in but may not limit the current enough to prevent long term damage (assuming the secondary clamp is not engaged).    However...
I think the standards require 1.1 times the meter's rating but I would need to double check that.  In any case the F87V uses S05K575 MOVs.  These have a 785VDC rating.   You have two in series.  So even if the voltage is at 1.5KVDC, the MOVs are not doing much of anything.  I've looked at a fair number of meters now.  Granted, there is some cheap junk out there but let's ignore that class.  For the meters rated for 1KV, looking over my note,  I don't see any that clamp a couple hundred over 1KV.   One meter was sort of a hybrid using a combination of GDTs and a MOV.   The MOV in this case had a DC rating of 670 volts.  However, the GDTs would not fire until 1200.  With a DC source, if you did manage to trip the GDTs they are going to stay on until the current drops which would put a strain on the MOV.  But again, we have to get he GDT to turn on in the first place.
Obviously if the voltage was so high that the MOVs were conducting and we left it like this for an extended time, I fully agree that the MOV will degrade but I just don't see this happening at the sub 1.5KV.
Assuming there is anything left of the 87V when I am done with it, I would be willing to attach it to a 1.2KVDC power supply and let it sit for a long term test.  Say a week with the meter in the DC volts setting.
Online rsjsouza
  • Eternally curious
« Reply #2403 on: March 04, 2018, 03:41:56 pm »
That sums it up. I should have prefaced my post with the assumption the MOVs are directly in parallel with the inputs - in other words, I was completely illiterate about the 87V's input circuitry. :)
That would be an interesting test.
« Reply #2404 on: March 04, 2018, 04:15:34 pm »
I saw a meter with a place holder for a MOV right across the inputs. I think it may have been sold by Tek Power.  The MOV was not populated for good reason.   Imagine the stress that would put on the leads, PCB, MOV....   If anyone ever comes across a handheld meter that uses a MOV without any additional limiting device/s, please let me know.
Consider it done.  I will need to set up some sort of containment so the inquisitive critters are kept safe.   Maybe measure the input impedance before and after the test.  Let me think about it.
Offline HKJ
    • Tests
« Reply #2405 on: March 04, 2018, 04:19:24 pm »
I have seen a spark gab across the input terminals, but I do not remember what meter it was.
Offline HKJ
    • Tests
« Reply #2406 on: March 04, 2018, 05:05:38 pm »
I am VERY interested in knowing what meter this is.
It was a cheap meter with the usually bogus CAT rating:
My review: http://lygte-info.dk/review/DMMPeakmeter%20MS8248S%20UK.html
« Reply #2407 on: March 04, 2018, 05:20:31 pm »
Speechless
Offline Cliff Matthews
    • General Repair and Support
« Reply #2408 on: March 04, 2018, 06:18:20 pm »
No shunt? er.. maybe it's the fuse..  :-DD  I've seen nube's everywhere raving on any product that gets rid of that "nasty range switch"
Offline Hydrawerk
« Reply #2409 on: March 04, 2018, 10:14:39 pm »
OK, Fluke 87V is not a very good product. What about Fluke 289, 28II or 179??
Offline Fungus
« Reply #2410 on: March 05, 2018, 04:28:15 am »
Does this look like a multimeter shootout thread or a "Which multimeter should I buy?" thread?  :-//
The 87V is supposedly the yardstick meter, the industry standard for excellence, the meter that can do no wrong.
So far joe's only unboxed it and noted a clunky user interface (how many secret power-on modes does it have?) and cheap fit and finish (poor plastics, bent shield).
Testing's not over yet. Not by a long way.
« Last Edit: March 05, 2018, 04:48:49 am by Fungus »
Offline tautech
« Reply #2411 on: March 05, 2018, 05:00:46 am »
Yeah right !
You're new here aren't you ?
Tucked way back on P13 of this 100 pages of goodness:
https://www.eevblog.com/forum/testgear/hear-kitty-kitty-kitty-nope-not-that-kind-of-cat/msg708183/#msg708183
2 1/2 years ago !
Offline Fungus
« Reply #2412 on: March 05, 2018, 05:23:44 am »
Yeah right !
Please pay attention.
There's a whole bunch of people who believe the Fluke 87V is perfect in every way, they insist the previous tests must have been done with faulty meters (or something).
The clue is at the top of the page you're reading:
« Last Edit: March 05, 2018, 07:14:56 am by Fungus »
« Reply #2413 on: March 05, 2018, 01:35:22 pm »
Funny, that poll was a fair amount of time ago as well.   Yes, I have read several comments where people felt something was wrong.  Indeed, it may have been.  That's partly why I am doing this series in smaller chunks as well.  It gives people time to digest what is going on and ask questions.
I still need to run the continuity tests and make sure I don't see any susceptible sweet spots.   Then it's onto more destructive testing.  My plan for the next part is to run the tests in the following order:
open fuse test
AC line
piezo grill starter
ESD gun
transient test to failure (if it can be repaired, I will go ahead and do that)
rotary switch cycle testing
long term 1.2KVDC for rsjsouza
Fungus like the drop tests so I may drop this thing on the block of wood as well
Again, all of this could change depending how the testing goes....
Offline Muttley Snickers
  • Cursed: 679 times
« Reply #2414 on: March 05, 2018, 02:34:33 pm »
Depending on the order you might be able to put Flukes warranty to the test at the same time.   :o :-BROKE
Offline ChrisLX200
« Reply #2415 on: March 05, 2018, 02:57:08 pm »
open fuse test
AC line
piezo grill starter
ESD gun
You forgot the flame-thrower test, and maybe Truck test (run it over with your car and see if it still works) :)
Offline 3db
« Reply #2416 on: March 05, 2018, 11:17:40 pm »
I'm sure Joe has a few guns too.  >:D
Offline tautech
« Reply #2417 on: March 06, 2018, 01:12:57 am »
I could shoot it with my fuse powered cork gun.    :-DD
Piezo ignitor and Methanol upgrade required ?
« Reply #2418 on: March 06, 2018, 10:52:23 am »
After seeing the bent shield in my brand new Fluke 87V, a member with a slightly older 87V took theirs apart and provided me with this picture.  I had hoped the meter I received was a one off mistake but it appears they are bending the shield and no one is catching it.  I have not yet done anymore with the 87V and it's still sitting in parts.  The shield looks like it is designed to sit flat.  I can place it flat in the board.   The two tabs seem to be setup correctly.    Maybe they cut their QC staff to make more money?  Maybe it's supposed to be bent?
Too bad these large companies have no presence in these groups.   If it were Brymen, I would just ask them and based on all my previous experiences with them, would have an answer in a day.  I never found a contact at Danaher/Fluke.
Offline Fungus
« Reply #2419 on: March 06, 2018, 12:09:20 pm »
Maybe it's deliberate.
Does it press against the back of the case like a spring? Does it look like there's any reason to do that?
Doesn't necessarily mean it's supposed to be like that.
Offline Fungus
« Reply #2420 on: March 06, 2018, 12:14:16 pm »
Dave's running a live switch cycle on his second channel.  I watched it for a few minutes.  Tried to calculate the resistance with the formulas shown but looks like its 1.7K ohms the one way  a little better the other.  Guessing I am missing something as I think he said they could only read up to 10 ohms or so.   So much computing power and they resort to sticky notes.   :palm:
He also cleaned off the accumulated dust/debris after every bunch of cycles.  :scared:
If that's fiberglass dust then it's an abrasive, just sayin'.
Offline ChrisLX200
« Reply #2421 on: March 06, 2018, 12:22:40 pm »
I don't think it's fiberglass - solder mask more likely. In normal use the meter would be picked up and put down potentially dislodging the dust anyway after every use or so - you cannot replicate normal usage easily. These sorts of test can only ever be a rough guide to wear rate, and you would need to test many to get some sort of average.
« Reply #2422 on: March 06, 2018, 12:57:58 pm »
It's held in place by the screws.  It does not appear to spring off the back of the case.  It looks like it was designed to fit flat. There are no clearance issues I see with that would prevent it from sitting flat.  I see no reason for it to be bowed other than poor quality control.
« Reply #2423 on: March 06, 2018, 01:24:18 pm »
I was under the impression that this round of testing was done live, non-stop, 50K half cycles, no cleaning with the shim installed correctly.  I only saw the last few minutes so if you saw him cleaning it again, that's a problem.   If your pulling that out of your ass and don't really know, we should ask Dave.
I don't need to run more of the same meter to see the difference in wear.   Between the pictures and resistance metric, it's good enough for me.   Some have certainly cut deeper than mask.  Then there is that detent spring, it's not always the contacts that are the problem.
Then again, I have been told many times that I can't draw a conclusion from the transient tests I run because the sample size is one.
« Last Edit: March 07, 2018, 12:37:28 am by joeqsmith »
Offline Fungus
« Reply #2424 on: March 06, 2018, 01:47:32 pm »
So basically the metal tabs in the middle of the shield just weren't in their PCB holes?
Oh, dear.
« Last Edit: March 06, 2018, 01:49:33 pm by Fungus »