Author Topic: Multimeter selection for production testing  (Read 1879 times)
Offline Antonio90
Re: Multimeter selection for production testing
« Reply #25 on: April 26, 2024, 07:51:26 pm »
For what is worth, the Siglent 3045X seems to have two shunts, and does interrupt the current path when switching from 2A range to anything lower.
« Last Edit: April 26, 2024, 07:53:52 pm by Antonio90 »
Offline J-R
« Reply #26 on: April 27, 2024, 07:30:35 pm »
How about a 1 Ohm shunt with a 6.5 digit DMM?
Offline tszabooTopic starter
« Reply #27 on: April 27, 2024, 08:23:51 pm »
100uA would give 100 uV. That's 0.1% of the smallest range. ADCs are not linear int these ranges, thermocouple effects are significant, and offset as well.
On the other end of the  scale, it would drop 2V on when the peak current happens. That's too much, the device would brownout.
Offline J-R
« Reply #28 on: April 27, 2024, 11:35:43 pm »
You didn't mention the voltage range of the device, and I assumed it had some regulation and could handle a 2V drop.  Maybe battery powered, such as nominal 4x1.5V or 12V.  But if your device runs at 3V I agree 100% with that value being a problem, in which case dropping to 0.1 Ohm might be an option.
100uV is trivial for a 6.5 digit DMM, 10uV is pretty easy with just some sanity checks, and even 1uV is possible if it's decent.  The DMM can autorange between 100mV/1V/10V without interrupting the supply.
Thermocouple effects are not much of an issue because you have the same number and type of connections going in as you do out.  But ideally still build a shunt configuration that is not going to get warm.
