tinhead, do you know how the built-in DMM is calibrating in handhelds? Is there any file with calibration data, or there is no software ability to influence on dmm readings?
the DMM IC they using is FS9922-DMM4, afaik there is no software calibration for the DMM in the DSO firmware.
It dioes simply reads the value from the DMM IC and display it directly (after parsing them of course).
The DMM can be only calibrated (hardware wise) via 4 smd (ugly) pots:
- R45 -> AC rectification value, pin 1 (top, near pin 1)
- R34 -> REF voltage, pin 4 (top, between pin1 and two diodes)
- R27 -> capacitance measurment input RCAP1 pin19 (bottom, left from R17 and relay)
- R17 -> capacitance measurment input RCAP2 pin 20 (bottom, between relay and R27)
Hantek is using Fluke 5500A for calibration, i would say that good enought. Let's hope they spend enought time per unit.
I don't see any conformal coating on your pictures, so should be no problem to calibrate DMM (if necessary).
There is however one thing, when calibrating ensure that Handheld battery is full loaded and calibration is done
when running on battery (with external power supply disconnected). There was some drift in mV range on older models
due crap external power supplies (well, OEM parts). No idea if that changed, so better to calibrate directly on battery.
These handhelds have no True-RMS circuit (well, one can use DSO for that), so AC calibrations is anyway not that important.
The REF can be however improved, Hantek is using FS9922-DMM4 internal source, like in app note 40k2 between REFO and REFI,
and then 10k pot + 40k2 resistor in series from REFI to AGND. The internal REFO is 1.2V bandgap source, with 100ppm/°C.
One can use here external reference, e.g. REF1112 (or ICL8069A like in app note, however initial accuracy of REF1112 is much better)
and trim it with some better pots (not that crap smd thing in series). But finally one will not get much better results, but more stable yes.