Author Topic: Rigol DP811 remote sensing  (Read 135 times)

0 Members and 1 Guest are viewing this topic.

Offline terranceTopic starter

  • Contributor
  • Posts: 34
  • Country: us
Rigol DP811 remote sensing
« on: Today at 05:36:45 am »
I got my DP811 calibrated, and the voltage output and measurement is within spec now.

but, when I toggle "sense", voltage output (externally measured) jumps between 2 values, is 1.8% lower when "sense off", but internal readings are the same. There's 0 load, and sense+ sense- is short to output +/- at the terminal.


- set: 1.000 V,  meas 0.982 V
- set: 10.000 V, meas 9.819 V
- set 40V. meas 39.284V
 

Offline terranceTopic starter

  • Contributor
  • Posts: 34
  • Country: us
Re: Rigol DP811 remote sensing
« Reply #1 on: Today at 06:43:44 am »
 >:( this drives me nuts, I did a "sense off" calibration, and now, it is 1.8% lower when "sense on" than "sense off"   :wtf:

And the "ADC cal" (page 2) fails, and "sense cal" has only 1 calibration point.

I'll try a "sense" calibration in script.
« Last Edit: Today at 06:49:02 am by terrance »
 