Most of the time when I switch mine on at normal room temperature it is with in +/-1mV of Zero volts. (especially after tweaking VR3).
Hence no point using zero button. The button should never be used for timed tests.
This unit is not designed for 'measurements' below 50mA. It gives an indication below 50mA but not something I would trust, given the noise threshold of +/-10-15mA.
Drift is caused by temperature change, using the Zero PB, or physically moving the clamp.
So eliminating those issues will reduce drift to within +/-1mV.
Given that most long term measurements involve measuring the change then initial zero is not critical.
Adding an output filter for slow measurements to reduce the +/-10-15ma noise is also fine.
Remember this unit is not a precision instrument. It is a cheep automotive clamp mostly used to measure currents above 100mA or to show the relative timing of switching events. ie injectors ect...
Precision instrument amplifiers, circuitry & design layout COST & are hard to duplicate when the clamp heads core is included.
This serves the purpose for which is was designed very well & is the cheapest reasonable option I have found. Pico sell a unit, with what I believe is basically the same circuit, for 2-3 times the cost & one can spend 5x the cost on other units for no gain.
At about 10x the cost some units start to out perform it, but one needs to pay at least $700US to get a unit that may be designed to minimise the noise & achieve good accuracy, sensitivity & frequency response. These units will usually have two separate but complimentary measuring systems for the DC & AC components.
One gets what one pays for has never been truer!
Cheers,
John.