Author Topic: Hacking the Rigol DHO800/900 Scope  (Read 1595976 times)
Awp and 24 Guests are viewing this topic.
Offline sebyonTopic starter
« on: September 23, 2023, 03:54:38 am »
Placeholder thread for hacking the Rigol DHO800/900 scope.
The general discussion thread can be found here.
https://www.eevblog.com/forum/testgear/rigols-new-dho800-oscilloscope-unbox-teardown/
@hubertyoung has provided a DHO804 FW1.14 image with the DHO924 vendor file preloaded. It can be extracted using 7zip then flash using HDD Raw Copy Tool (compressed image).
https://mega.nz/file/UjBC3KRY#Kqv1BCHNQdPcUGMfR8IqbuUwHUsUhU4GpO1keTAXqf8
@Luc7777 provided some guidelines on how has has achieved this.
  • This is what I've done:
    1. Run the Win32 Disk Imager
    2. Backup the SD
    3. Flash the SD with the image from the link
    4. Run the claibartation (offset gone) - device identifies as DHO804
    5. Connect the scope to ethernet
    6  Run adb:
        6.1 adb devices
        6.2 adb connect 192.xxx.x.xxx:55555
        6.3 "adb pull /rigol/data/vendor.bin"
        6.4 backup the generated vendor bin file from the adb folder to a new location
        6.5 copy in the adb folder the DHO924 image
        6.6 "adb push vendor.bin /rigol/data"
MOD EDIT:
Here is the latest instructions:
DHO800/DHO900 UNLOCK TOOLS
1) Install GOLang distribution
2) In the "run_DHO_Tools.bat":
- set the GO installation directory path
- set the IpAddress variable (your scope's address is on the IO tab of the "Utility" window)
- change options list, if DHO900
- change scopeID, if DHO900
- if you don't want to create a backup file and pull it to the computer, delete line 35, or make it comment like this:
rem call "adb\05 make Backup And pull it - adb rm updateGEL, sh buildGEL, pull.bat"
3) Run the "run_DHO_Tools.bat"
4) Send the generated SCPI commands to the scope via the SCPI browser tab, opened by the script. Common command view:
:SYSTem:OPTion:INSTall DHOX00-<option>@XXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXXX
Scope reboot is not needed.
5) Check BW limit on the "About" tab and the memory depth (for the DHO804) on the "Options" tab of the "Utility" window.
PS: To remove installed options use the "adb\03 adb remove ALL options.bat" file or the ":SYSTem:OPTion:UNINstall" command from p268 of the DHO800/DHO900 Programming Guide.
UPDATE REASON: extending the description text.
« Last Edit: December 16, 2023, 08:00:16 am by sebyon »
The following users thanked this post: EEVblog, Mechatrommer, lowimpedance, thm_w, Mortymore, bkw, dreamcat4, Serg65536, Vovas
Offline EEVblog
Re: Hacking the Rigol DHO800/900 Scope
« Reply #1 on: September 23, 2023, 04:01:00 am »
From here:
https://www.eevblog.com/forum/testgear/rigols-new-dho800-oscilloscope-unbox-teardown/msg5072767/#msg5072767
Here is the DHO804 FW1.14 image (Thanks to @hubertyoung) with the DHO924 vendor file preloaded. Extract using 7zip then flash using HDD Raw Copy Tool (compressed image). If there is an vertical offset then use self cal or extended self cal.
and
    2. Backup the SD
    6  Run adb:
        6.1 adb devices
The following users thanked this post: Mechatrommer, thm_w, dreamcat4, Serg65536, sebyon
Offline dreamcat4
« Reply #2 on: September 23, 2023, 08:48:11 am »
initial thoughts for hacking external awg support, link to a prior discussion. just to layout the main options:
well yes, there is the choice of hacking the firmware alone, to trigger some actions that then communicates over one of the general purpose digital interfaces on the scope (whether that is via the ethernet, or usb, or something else). the advantage of that is then should not require any hardware mods / hacking whatsoever. which gives itself a potential to keep a viable product warranty...
then there's the idea of minimally hacking the pcb awg section there, to install some low level mcu digital interface within the scope....
[more]...
this was after some people had been asking about the real internal awg module
https://www.eevblog.com/forum/testgear/rigols-new-dho800-oscilloscope-unbox-teardown/msg5045164/?topicseen#msg5045164
Offline Nikki Smith
    • Raspberry Pi, ARM SBCs & electronics hackery
Logic Analyser for DHO800/900
« Reply #3 on: September 23, 2023, 11:00:26 am »
For use with the DHO900 (and hopefully the DHO800 after a few mods?) I've created a new v3.1 of the 16-channel LA clone board that is cheaper to make and easier to hand-solder:
https://climbers.net/sbc/clone-pla2216-logic-probe-analyzer/
The PCB has dual footprints for the quad-channel SN65LVDS391, so you can use SOIC or TSSOP packages depending on which is cheaper (or available in stock).
KiCad source files included, as are GERBERs and BOMs for LCSC & Mouser. Should be around US$15 incl components and PCB.
!! I haven't tested it !! Just waiting for the components to arrive...
The DHO800 at a minimum will need a hole cut into the front panel for the 50-way IDC connector, and a socket soldered onto the motherboard. Some people have said the 2 missing memory chips are required to use the logic analyser functionality, but I don't think that has been confirmed yet? The missing chips are GigaDevice GDP2BFLM-CA, 96-ball FBGA:
https://www.gigadevice.com.cn/Public/Uploads/uploadfile/files/20230704/DS-00823-GDP2BFLM-Rev1.1.pdf
The following users thanked this post: EEVblog, simonboydfoley, dreamcat4, AceyTech, NikNik
Offline Nikki Smith
Cloning the DHO900 AFG
« Reply #4 on: September 23, 2023, 11:10:43 am »
I've started an attempt to clone the plugin AFG module from a DHO900.
In theory this could be used to add the same AFG functionality to a hacked DHO800. At a minimum 2x5-pin and 2x20-pin 1.27mm SMD headers would need to be soldered to the motherboard, and possibly other missing components.
I've gone as far as I can just working from photos: the attached DHO900 AFG.zip contains KiCad schematic & PCB layout, hi-res photos and spreadsheet of the major components. Hopefully other people with the actual AFG hardware will be mad enough to continue the effort, tracing the tracks and measuring the capacitors & inductors.
It is a busy board with a lot of passives, but I haven't seen any expensive/custom components yet.
« Last Edit: September 25, 2023, 07:18:52 am by Nikki Smith »
The following users thanked this post: Mechatrommer, richfiles, nunojpg, Houseman, dreamcat4, AceyTech, vd1992, NikNik
Online tv84
  • Country: pt
« Reply #5 on: September 23, 2023, 12:36:28 pm »
This vendor.bin thing was shown a year ago in the DHO1000 thread (HDO1000 at the time).
The vendor_dho924.zip that is being shared contains the following info:
00000000  File CRC32: A2A69654  [00000008-000000D3]  CRC OK
00000004 File Length: 204  Size OK
00000008    NameSize: 4
0000000C        Name: E_CFG_MODEL_RAW
00000010   FieldSize: 56
00000014       CRC32: 0ADE0274  [0000001C-0000004B]  CRC OK
00000018    DataSize: 44  Size OK
0000001C        Data: DHO924
0000004C    NameSize: 4
00000050        Name: E_CFG_SN_RAW
00000054   FieldSize: 56
00000058       CRC32: 6184E52C  [00000060-0000008F]  CRC OK
0000005C    DataSize: 44  Size OK
00000060        Data: DHO9A252500008
00000090    NameSize: 4
00000094        Name: E_CFG_MAC
00000098   FieldSize: 56
0000009C       CRC32: 8292C492  [000000A4-000000D3]  CRC OK
000000A0    DataSize: 44  Size OK
000000A4        Data: 0019AFA00004
  File processed OK
I've been away from DHO investigations as, I think, all is done since DHO1000 came out.
With that in mind, I ask if people have tried to just change the model in their vendor.bin to DHO924 and see if that is enough for a model conversion? Given previous Rigol cases I think it should not be necessary to flash whole new CF 924 images.
But I might be wrong...
As in the past, there should be a SCPI command to change params in vendor.bin:
:VENDor:CONFigure
If anyone with a DHO8xx wants to try this SCPI method, I can try to revive my dusty tool and create the command file.
« Last Edit: September 23, 2023, 12:47:21 pm by tv84 »
The following users thanked this post: EEVblog, bulba99, iMo, Martin72, mwb1100, akkk44
Offline akkk44
  • Country: cn
« Reply #6 on: September 23, 2023, 02:39:33 pm »
For the first batch of DHO804/814s with firmware version 01.14: yes, replacing only the vendor.bin will be enough.
The issue here is that newer DHO804/814s ship with firmware version 01.00, simply replacing the vendor.bin will cause a 5mV-10mV offset to appear, and it can't be eliminated by self-cal.
Thanks to @hubertyoung, we now have access to SD card images with firmware version 01.14, we can basically "downgrade" the newer scopes back to the state it was in during the first batch, which is easily hackable.
Sidenote: Someone has claimed that he managed to perform the hack on firmware version 01.00, but he refuses to provide the method, so I don't know if there's any credibility to his claims.
I am curious about how the vendor.bin is decrypted, and it seems that you mentioned a method for changing the content of it with the SCPI command? It is certainly interesting.
« Last Edit: September 23, 2023, 02:48:04 pm by akkk44 »
Online tv84
  • Country: pt
« Reply #7 on: September 23, 2023, 02:54:11 pm »
But isn't v01.14 a newer version than v01.00?  :-// :o
What about just replacing the cal files and let the FW do the self cal?  Maybe the "newer" cal files are so off that the self cal can't correct them...
I just think that it has to be an easier way.
« Last Edit: September 23, 2023, 02:57:44 pm by tv84 »
Offline ebastler
Re: Logic Analyser for DHO800/900
« Reply #8 on: September 23, 2023, 04:12:35 pm »
I think what has been confirmed is that (a) the DHO900 models have these chips populated, and (b) the DHO800 models can be upgraded to 50 Mpts memory via just a software hack, i.e. the extra memory chips are not required for sample storage. That's what led to the assumption that the extra memory supports the logic analyser functionality.
Online TurboTom
« Reply #9 on: September 23, 2023, 06:27:25 pm »
True, but there's something that doesn't make much sense: Why does Rigol add two 256MBits * 16 RAM chips to sample 16 digital channels while one of these chips would have been more than sufficient for this job. Considering the ADC delivers a 12bit datastream at 1.25GSa/s and a single such RAM chip, interfaced via the Zync FPGA/SoC is well fast enough to cope with this data rate, we may in future DHO900 units only see one of the additional footprints populated with a RAM or .... there may be hacking possibilities that as yet nobody has on their list. At least, all the front end hardware including ADC and ADC sampling clock PLL are capable of 2GSa/s. And two such RAM chips, if you swap the ADC / MSO data streams, and the FPGA section of the Zync has enough I/O connectivity ... well I guess you know what I mean ;D
Of course that's something that isn't done at home easily, but maybe even Rigol has plans for a further uprated model based on this hardware. Pure speculation...  ;)
The following users thanked this post: Fungus, Nikki Smith
Offline ebastler
« Reply #10 on: September 23, 2023, 07:30:16 pm »
Well, there is that strange mismatch in the DHO924 specs, with four channels at 250 MHz bandwidth and still only 1.25 GSamples/s. They even ship four fast probes with it, which one can't fully use.
Maybe the DHO800/900 was indeed designed with a higher sample rate in mind. Then either some technical challenge got in the way late during development, or product management intervened and wanted to protect the more expensive product lines for the time being. The latter is a more likely explanation in my opinion, and might mean that we will see faster sampling a year from now, maybe even software-upgradable in the current scopes...  ::)
Online Martin72
  • Testfield Technician
« Reply #11 on: September 23, 2023, 09:26:24 pm »
The maximum you can get out of one ADC for all channels can be seen with the DHO1000.
This model is really "reduced by half" compared to the DHO4000 with its 2 ADCs.
It is interesting to see why the bandwidth of the DHO900 is 250Mhz and not 200Mhz like the DHO1000, which seems more "natural".
(Kierkegaard)
Offline EEVblog
« Reply #12 on: September 23, 2023, 10:55:16 pm »
The latter is indeed the most likely explaination. But it does also have a different FPGA which could potentially impact that.
Offline EEVblog
« Reply #13 on: September 23, 2023, 10:57:45 pm »
I don't get that either. Mine is 1.00
Offline EEVblog
« Reply #14 on: September 23, 2023, 11:01:01 pm »
I suspect that few will go to the effort to hack their DHO800 to this extent, but the probe should be popular with DHO900 owners given the Rigol probe costs a crazy $400.
The following users thanked this post: Nikki Smith
Online Fungus
« Reply #15 on: September 24, 2023, 04:07:04 am »
I think it's more likely it was the FPGA that decided it.
The cheaper FPGA in these probably can't keep with the 2GHz sample rate.
This is definitely strange. I'm hoping my DHO800 can be set to 125Mhz, it seems like the sweet spot for this ADC on a 4-channel 'scope.
« Reply #16 on: September 24, 2023, 05:04:47 am »
Honestly probably more of a development / cost thing. Probably cheaper to throw in a slightly overkill probe on the top DHO900 scope if someone is willing to pay for all the features than commission a probe new to spec for only one line. The PVP2350 (the 350MHz probes included with the DHO924) is the standard probe for the MSO5000.
Offline ebastler
« Reply #17 on: September 24, 2023, 06:46:44 am »
Well, there is that strange mismatch in the DHO924 specs, with four channels at 250 MHz bandwidth and still only 1.25 GSamples/s. They even ship four fast probes with it, which one can't fully use. [...]
Oh, I was not referring to the 350 MHz spec on the probes, but to the fact that Rigol gives you four of them. The sampling rate drops to  ~300 MHz when using more than two channels, so two fast probes and two of the cheaper 150 MHz type would have been enough.  ;)
Yes, I realize that having two different pairs of probes would be a hassle, so I appreciate that Rigol provides four PVP2350s. Decent value for money for the $100 surcharge they ask for the 924 model -- if only the scope could fully use the probes...
Offline ptluis
  • Country: pt
« Reply #18 on: September 24, 2023, 08:13:50 pm »
I was sniffing Dave's DHO800 dump file and found references to 2 other models whose initial letters were swapped. Should they be the same models but launched in another country? maybe yes maybe no.
They are HDO800 and HDO900, included are also DHO800 and DHO900
Offset       0  1  2  3  4  5  6  7   8  9  A  B  C  D  E  F
01D6DB8D0                        48  44 4F 38 30 30 00 48 44          HDO800 HD
01D6DB8E0   4F 39 30 30 00 44 48 4F  38 30 30 00 44 48 4F 39   O900 DHO800 DHO9
01D6DB8F0   30 30                                              00
I also found references to codes assigned to options available or not on the oscilloscope and a type of .lic file
Maybe it will be possible to activate the options through these codes and generate such licenses the same way we did on the DS1054Z model
01D6DB7F0                            42 4E 44 00 42 57 37 54           BND BW7T
01D6DB800   31 30 00 45 4D 42 44 00  43 4F 4D 50 00 42 57 31   10 EMBD COMP BW1
01D6DB810   35 54 32 35 00 41 55 54  4F 00 42 4F 44 45 00 42   5T25 AUTO BODE B
01D6DB820   57 37 54 32 30 00 42 57  31 30 54 32 30 00 52 4C   W7T20 BW10T20 RL
01D6DB830   55 00 42 57 32 54 34 00  42 57 32 54 38 00 42 57   U BW2T4 BW2T8 BW
01D6DB840   34 54 38 00 41 45 52 4F  00 46 4C 45 58 00 41 55   4T8 AERO FLEX AU
01D6DB850   44 49 4F 00 2E 6C 69 63  00 40 00 55 6E 6B 6E 6F   DIO .lic @ Unkno
01D6DB860   77 6E 00 46 6F 72 65 76  65 72 00 64 61 79 73 00   wn Forever days
Online tv84
  • Country: pt
« Reply #19 on: September 24, 2023, 08:16:55 pm »
The HDO was Rigol's initial designation before having to change to DHO.
Regarding the lics generation, look into the HDO/DHO1000/4000 thread. It's all there so there is no need to reinvent the wheel.
The following users thanked this post: Vovas
Online Fungus
« Reply #20 on: September 24, 2023, 08:35:30 pm »
They were called "HDO" when they were announced but that turned out to be a trademark (of Lecroy?) or something so they changed it to "DHO".
The firmware probably works both ways.
Offline ptluis
  • Country: pt
« Reply #21 on: September 24, 2023, 08:37:04 pm »
Nice!  :-+
Online Fungus
« Reply #22 on: September 24, 2023, 08:38:14 pm »
Does it have power analysis?
Let's hope so. Ideally I want my DHO800 to be 125Mhz bandwidth and 50Mpts of memory but without the extra AWG/LA menus of the DHO900.
Offline ptluis
  • Country: pt
« Reply #23 on: September 24, 2023, 08:39:01 pm »
So, probably some code leftovers.
Offline ptluis
  • Country: pt
« Reply #24 on: September 24, 2023, 08:41:08 pm »
I don't have the scope (yet) so I can't tell you what each code means. Sorry