For DS1000Z:
Offset Range (Probe ratio is 1X)
1 mV/div to 499 mV/div: ±2 V
500 mV/div to 10 V/div: ±100 V
As far as I could see, DougM wants to look at part of 40V P-P signal at better than 1V/DIV or less..
No digital scope I know of will have dynamic reserve of 200x of full screen sensitivity... And probably only +-8 to 10 DIVS...
That depends entirely on how the input circuit is designed and it is not 200x but 5x (40V/8V full display range @ 1V/div = 5x). If the oscilloscope allows a large offset then the input circuitry should be designed to handle a signal which swings between the maxima of the offset. It seems the Rigol has a design flaw and thus has overdrive problems.
Yep, the devil is in the detail.
For the new Siglent X-E series:
Offset ranges 1x input.
500uV~118mV/div: ±2V
120mV~1.18V/div: ±20V
1.2V~10V: ±200V/div
2V/div will give sufficient offset for DougM's needs and using the minor divisions ~250mV resolution.
Tautech, thanks for the info..
It is not about offset. That was mentioned in passing, not relevant for original question. It is a good reminder to the fact that all scopes are different, though..
For instance from 500mV to 1.2V Rigol has 100V offset compared to Siglent 20V. If that is of significance to anybody...
But that is only relevant if you have a small AC component riding on top of a large DC component. In which case you can use AC coupling too.
Most of the time, DC offset is simply used to move channels vertically on screen for visual separation.
Problem is something else.
If you have 10 vertical DIVs on screen... you put in 40V AC P-P, and put it on 5V/div you see nice +/-4 divs signal.
Then you realize a small something, maybe 100mV anomaly at zero cross. So, you put it on 50mV/div so anomaly would be at least +/-1 div (2 div P-P) vertically so you can have a better look. At this point, full screen dynamic range of input is +/-500 mV (1V P-P).. You are overdriving it 40X... On top and bottom.. not where you are looking.
Old analog CRT scopes had different circuitry, working on much higher internal voltages (even full solid state integrated ones). They had to actuate electrostatic plates on CRT tube. That gave them much larger linear dynamic range and also many times a more graceful clipping and recovery... They were better at this.
Regards,
Sinisa