Re: GM328 out of order
« Reply #25 on: August 04, 2021, 11:29:45 am »
There should be a form of autotest procedure inside the ATMega, which causes the initialisation to be VERY long, and made me think about quartz problem...

That is right. If the 3 test terminals are short-circuited at start-up, the self-calibration procedure is started.
 

Offline PBhTopic starter

  • Contributor
  • Posts: 14
  • Country: fr
Re: GM328 out of order
« Reply #26 on: August 04, 2021, 06:42:11 pm »
I would sooner expect the 328P to be damaged, these testers don't have much for input protection.
For the component socket 3 inputs, GM328 DO have protection diodes TO Vcc and GND and a TVS.
 