Re: GDS-2000A series New economic oscilloscope by GW Instek comes to market
« Reply #125 on: April 02, 2013, 05:38:36 pm »
It seems that this scope GDS-2000A is the same case as Tektronix DPO2000 series. The update rate is quite good (6200 waveforms/s) up to 100ns/div, and then at 20ns/div it drops dramatically down to only 160 waveforms/s.
How are you drawing that conclusion?  We don't have any numbers other than their "ideal" at this point.
According to the setting that I had in this video... But it is not a conclusion, it's only a hypothesis.
Offline icon
« Reply #126 on: April 02, 2013, 07:01:37 pm »
Just to clarify, you're not saying that the GDS-2000A actually has comparable update rates to the Tek DPO2000, but that, like the Tek, there's a 'headline rate' (80k wfm/s, in the GDS-2000A's case) which rapidly (more rapidly than the Agilent) drops off as you move away from the optimum? Also, am I imagining things or does the Rigol DS2000 do a similar thing? I'm sure I saw a table somewhere, but I can't find it now.
Isn't it fun playing oscilloscope Top Trumps?
Cheers
John
Offline marmad
  • Country: aq
    • DaysAlive
« Reply #127 on: April 02, 2013, 07:14:27 pm »
Hi John,
Sure, all scopes have a wfrm/s rate that changes - always dropping towards the slower timebases because of the limiting factor of time: e.g. at 1ms/div, assuming the display is 10 divs, each sweep is 10ms - meaning the absolute maximum rate possible (with an impossible 0% dead time) is 100 waveforms per second. At the upper timebases, it appears that most DSOs (aside from the Agilent X series with their new ASICs) have a peak setting - and then begin to drop again to the smallest timebase setting. As far as the GDS-2000A series goes, we have no hard evidence of what it does in any range - except Instek's own demonstrations which appear to make it clear that it's peak setting is at 500ns/div.
« Reply #128 on: April 02, 2013, 09:12:16 pm »
Yes, the top waveform update rate of GDS-2000A is better than Tek DPO2000.
Here is the waveform update rate as measured by marmad. It also drops down at shorter timebase settings, but it is although quite good.
Here are the results of a quick test I did measuring the Rigol DS2072 waveform update rates at all timebase settings and memory depths (also attached in Excel format). If you compare these to the Agilent 2000X series published rates, it's obvious the Agilent is the clear winner - although it doesn't have anything close to the possible memory depths.
Single channel - 1MHz square wave to Channel 1 - Trigger Out to frequency counter. I took best-case rate when it was fluctuating:
14kPts 140kPts 1.4MPts 14MPts 56MPts
5ns 15,000 13,150 1,412 142 36
10ns 9,400 9,400 1,412 142 36
20ns 50,012 13,515 1,416 142 36
50ns 25,003 13,515 1,416 142 36
100ns 17,859 13,159 1,412 142 36
200ns 11,365 11,360 1,408 142 36
500ns 5,434 5,435 1,336 142 36
1us 5,263 2,890 1,126 139 35
2us 5,054 1,506 846 133 35
5us 4,425 1,176 733 130 35
10us 3,789 1,157 720 130 35
20us 2,945 992 442 117 34
50us 1,326 639 414 114 34
100us 683 421 306 94 32
200us 347 245 200 69 28
500us 140 109 97 39 21
1ms 70 56 52 29 15
2ms 35 29 27 19 10
5ms ~14 ~13 ~11 ~9 ~6
10ms ~7 ~6 ~6 ~5 ~3
20ms ~4 ~4 ~3 ~3 ~2
50ms ~2 ~1 ~1 ~1 ~1
100ms  ~1 ~1 ~1 ~1 ~1
https://www.eevblog.com/forum/testgear/first-impressions-and-review-of-the-rigol-ds2072-ds2000-series-dso/30/
Offline grego
« Reply #129 on: April 02, 2013, 09:38:49 pm »
I've made a command decision and I will be getting the 2204A. My reasoning is price/performance.
I actually just got off the phone with the regional sales manager here in the US (a nice gent named Frank Paz) in Chino, CA.  I had a couple of questions for him which he answered satisfactorily to me.
1. Where is it serviced if there is a problem?
It's actually serviced here in the States.  They have a repair facility in Chino, CA to handle service.
2. If it needs recal where does it have to go?
Again, their US service facility will handle the recal. Negative for them is the recommended annual recal vs Agilents 2 or 3 year cycle but whatever. It's done here in the US and has a 2 week turnaround.
3.  CAN decode.
As I mentioned earlier its in their next firmware update I believe. Either way its coming and it's not an addon. I actually use this so it's kind of important if I have a LA that it can do CAN.
4.  Fine vertical (or lack thereof)
This is on their firmware roadmap with no scheduled release date.  Frank mentioned he had just heard back from the factory from my earlier query about this. Instek has a decent rep here in the US so I don't think he's blowing smoke up my ass. Then again there's no release date scheduled so it could be tomorrow or 2 years from now.  Not a deal breaker though.
From tequipment.net I can get this for $1765 and add in the LA a little later after I mess with the scope.  That's $100 less than the best price listed for then 2000x used on eBay, and when you add in all the upgrades for the Agilent to bring it to par with the Instek it's a significant amount (memory upgrade, segmented memory, decodes, etc).
Stay tuned.
Greg
Offline marmad
  • Country: aq
    • DaysAlive
« Reply #130 on: April 02, 2013, 10:00:03 pm »
Good for you, Greg - early adopters rule!  :D  I'm looking forward to hearing about it, and I hope you'll consider doing some kind of review here.
Offline tequipment
    • TEquipment
« Reply #131 on: April 04, 2013, 05:50:49 pm »
If you want EEVblog price for this GDS unit drop me an email, salesteam@tequipment.net  Ref the EEVblog and me, Evan Cirelli.
I will authorize a discount.
One again everything thanks for the business.
Evan Cirelli
Vice President of TEquipment.NET
Offline dr.diesel
« Reply #132 on: April 04, 2013, 06:12:32 pm »
Evan Cirelli
Does this apply to other manufactures as well?
Offline tequipment
    • TEquipment
« Reply #133 on: April 04, 2013, 06:44:27 pm »
Yeah for the most part.  I think I am going to set up a EEBlog coupon or discount.  In the future we can give EEBlog pricing as a login but for now email us for a quote.
I want to be as honest and open with you guys as possible with pricing.  We want your business.
Thanks
Evan
Offline dr.diesel
« Reply #134 on: April 04, 2013, 06:49:48 pm »
Thanks
Evan
You are my new best friend!    :-+
Offline marmad
  • Country: aq
    • DaysAlive
« Reply #135 on: April 05, 2013, 02:02:18 pm »
While looking through the manual carefully, I noticed something which is buried in there (page 110):
The 2MPts of sample memory are only usable in single channel - single trigger mode (or as segments). When using normal/auto triggered modes, it's 1MPt for a single channel - or 500k when using 2 channels.
Granted, 1M/500k is adequate for most tasks so it's not necessarily a problem. It's just that GW-Instek perhaps should make this a little more clear in documents/adverts for the DSO - otherwise it seems slightly disingenuous.
« Last Edit: April 05, 2013, 02:47:18 pm by marmad »
« Reply #136 on: April 05, 2013, 04:15:35 pm »
Someone should measure the waveform update rate for more timebase settings. I am afraid it will be worse than Rigol DS2000.  :(
Offline grego
« Reply #137 on: April 05, 2013, 04:34:45 pm »
It's not.
http://www.ittsb.eu/GDS-2102A%20Wfms%20measurments.html
Just use table 3 as your data as we're discussing the validity of the other 2 tables.
« Reply #138 on: April 05, 2013, 04:55:58 pm »
Thanks for the link. http://www.ittsb.eu/GDS-2102A%20Wfms%20measurments.html The Trig Out is quite weird, but it might not be a problem... Unless you want to use it synchronise another instrument... It should be like 0...5V signal.
c) The Trigger-Out has an voltage range not higher than 200 mV AC, which varies significantly depending the time base of the scope and usually ranges from 120 mV to 5 mV and even less than that.
The only multimeter which had the proper sensitivity as frequency counter due their low mV range, plus the Min/Max/Average function, is the U1270A Series.
(...) 3) Avoid using a multimeter if possible, very few are capable for this task, if you have a sensitive and fast dedicated frequency counter capable for 4-5 measurements at the 100 KHz range … Use it.
What? Marmad used his multimeter to measure waveform update rate of his Rigol DS2000 and it was OK. http://youtu.be/gAY1GQEjrfc?t=2m51s
« Last Edit: April 05, 2013, 05:10:42 pm by Hydrawerk »
Offline marmad
  • Country: aq
    • DaysAlive
« Reply #139 on: April 05, 2013, 05:02:15 pm »
Well, I can't find any specifications from either Rigol or GW-Instek about what the levels are for their Trigger Output terminals on those DSOs. I wonder if they might be optically-coupled, requiring a pull-up resistor to get a decent higher level voltage. But in any case, my Fluke 87 III doesn't seem to have much trouble reading the Rigol's output.
« Reply #140 on: April 05, 2013, 05:12:55 pm »
Rigol doesn't specify the Trigger output level, but it will be maybe 0 to 5V?
At Agilent it is like this:
« Reply #141 on: April 05, 2013, 05:15:07 pm »
At a normal /non GW Instek/ scope it is no problem to measure the Trig Out with a low-end counter or multimeter... You've probably seen this my video.
Offline marmad
  • Country: aq
    • DaysAlive
« Reply #142 on: April 05, 2013, 05:22:51 pm »
Just measured the Rigol and it appears to have a Trigger Out voltage level of ~3.5V DC into 1M; ~1.5V into 50 Ohm.
« Last Edit: April 05, 2013, 05:27:21 pm by marmad »
Offline grego
« Reply #143 on: April 05, 2013, 05:56:15 pm »
Well it's moot for me now.  I just finished ordering mine from tequipment.net -- great pricing.
I should have it in my grubby little hands by the middle of next week.
Offline marmad
  • Country: aq
    • DaysAlive
« Reply #144 on: April 05, 2013, 06:19:49 pm »
Well, if you dare take it apart, I'm sure there are loads of members who'd love to see what's inside  ;)
« Reply #145 on: April 05, 2013, 08:12:12 pm »
That's how it should be. I suspect that there is a hardware bug in early batch of GDS-2000A. The manufacturer knows it. The Taiwanese businessman at the trade fair told me not to use a multimeter to measure the Trig Out.
I will not buy any GW Instek product, because the company has no subsidiary in Europe.  :-- Are you from USA, grego? Then it's OK. The scope is nice and feature rich. Suitable for those who need the mask pass/fail test.
« Last Edit: April 05, 2013, 08:15:09 pm by Hydrawerk »
Offline marmad
  • Country: aq
    • DaysAlive
« Reply #146 on: April 05, 2013, 08:55:41 pm »
Well, I like the feature set of the Instek - especially some things sorely missing from the DS2000. But I don't care about the expandability because I would never buy any of that stuff (I like a separate LA and AWG). I definitely prefer the design and interface of the Rigol by a large margin - even the type and method of displaying screen information, IMO, is better on the Rigol (the GW-Instek doesn't even show you the sample rate on the screen). And I love the low-noise 500uV front end and large memory of the Rigol.  So I think if I was choosing 2-channels today, I would probably still go with the Rigol.
But as far as 4-channels goes, the GW-Instek is significantly cheaper than the DS4000 for a similar DSO - so it's much more attractive.
Offline grego
« Reply #147 on: April 05, 2013, 09:08:00 pm »
Yep.  I'm outside Chicago here in the US of A.
« Reply #148 on: April 05, 2013, 09:20:08 pm »
GDS-2000A has no High resolution mode, only 8 bit all the time.
Offline grego
« Reply #149 on: April 05, 2013, 09:27:30 pm »
Correct - that was a known for me.  Not a huge deal for me considering the other options that are of use to me.