Today I received a new motherboard for TC3 and installed it into my TC2. In general it works properly, however, there are several strange bugs.
1. Generator. The output frequency is exact only in Square and Pulse modes, in other modes (sine, triangle, sawtooth) it's way off, up to 2%, and changes only in large steps.
2. Oscilloscope. On high voltage ranges (1 V/div and higher) in DC mode the zero line is off by 100..300 mV in a random direction (depending on the range), usually up. Can this be adjusted?
3. Oscilloscope. Low time per division ranges (5 μs/div and less) have incorrect X scale. For example, a 500 kHz wave on the 2 μs/div range must be EXACTLY 1 large box wide, actually I see 14..15 waves per 12 boxes. A 100 kHz wave on the same range must be 5 boxes wide, actually it's 4 boxes (is it 2.5 μs/div???). The 10 μs/div range seems to work properly.
4. Voltage. There is a calibration mode with two points: 5 V and 20 V. After calibrating (using AD584 for 5 V and a step-up converter for 20 V), it displays 5 V, 7.5 V and 10 V correctly (+/- 5 counts), but 2.5 V is 1% off (high). Why?