Author Topic: Fluke multimeter calibration  (Read 2707 times)

0 Members and 1 Guest are viewing this topic.

Offline bterrierTopic starter

  • Newbie
  • Posts: 7
Fluke multimeter calibration
« on: July 26, 2018, 10:18:02 pm »
Hello everyone,

This is my first post here, just found this forum today.  Looking into possibly purchasing a Fluke 8840a.  The guy is selling it “as is.”

It does power up but without any leads connected, it shows a voltage of about 2.5v.  Are these easily calibrated by the user or does it need to be sent in to Fluke?

I am hoping this is the only issue but I really have no way of knowing.  I have some experience with electrical and electronics work, but I am still very much a beginner.  Are these meters easily repairable?

I apologize if these questions have been asked prior, but I could not locate anything specific to calibration etc.


Sent from my iPad using Tapatalk
 

Offline Fungus

  • Super Contributor
  • ***
  • Posts: 18062
  • Country: 00
Re: Fluke multimeter calibration
« Reply #1 on: July 26, 2018, 11:15:14 pm »
 

Offline coromonadalix

  • Super Contributor
  • ***
  • Posts: 7958
  • Country: ca
Re: Fluke multimeter calibration
« Reply #2 on: July 27, 2018, 02:32:48 am »

Does it have the ac-08 option ???  Is the vfd display bright ??    If not  don't bother to buy it


At the section 6 you have self tests procedures and calibration in section 6.6 , the dmm doesn't need to be opened

Tell the seller to do a self test, if its show errors  move away from it ...    sorry if you are "newbie" don't take the chance to have more problems, it may get costly  ... hard to find parts or else ... 

BUT you have schematics in the manual and a there is knowledgeable people here.

Normally you need a dc calibrator IE: fluke 5700 series ... or very stable/precision voltage sources and high precision resistors


Some ic's have normal part number, others have the fluke ones

You have a tear down here :  https://www.eevblog.com/forum/testgear/fluke-8840a-bench-multimeter-teardown/
« Last Edit: July 27, 2018, 02:40:53 am by coromonadalix »
 

Offline bterrierTopic starter

  • Newbie
  • Posts: 7
Re: Fluke multimeter calibration
« Reply #3 on: July 27, 2018, 03:24:19 am »

Does it have the ac-08 option ???  Is the vfd display bright ??    If not  don't bother to buy it


At the section 6 you have self tests procedures and calibration in section 6.6 , the dmm doesn't need to be opened

Tell the seller to do a self test, if its show errors  move away from it ...    sorry if you are "newbie" don't take the chance to have more problems, it may get costly  ... hard to find parts or else ... 

BUT you have schematics in the manual and a there is knowledgeable people here.

Normally you need a dc calibrator IE: fluke 5700 series ... or very stable/precision voltage sources and high precision resistors


Some ic's have normal part number, others have the fluke ones

You have a tear down here :  https://www.eevblog.com/forum/testgear/fluke-8840a-bench-multimeter-teardown/

I believe you are referring to “Option 09.”  If that is the case, he says it has option 09.  Can I tell by looking at the meter if this is true?  or is it only internals that make a difference.

I am not sure that it does need calibration.  I only saw a photo with no leads connected and the voltage was reading at 2.5.  Is this normal?  My handheld meter stays pretty close to zero.....it drifts slightly but not 2.5 volts. 


Sent from my iPad using Tapatalk
 

Offline Fsck

  • Super Contributor
  • ***
  • Posts: 1157
  • Country: ca
  • sleep deprived
Re: Fluke multimeter calibration
« Reply #4 on: July 27, 2018, 06:50:59 am »
high impedance meters display a relatively high voltage when they're left open circuit
"This is a one line proof...if we start sufficiently far to the left."
 

Offline Fungus

  • Super Contributor
  • ***
  • Posts: 18062
  • Country: 00
Re: Fluke multimeter calibration
« Reply #5 on: July 27, 2018, 09:14:34 am »
I am not sure that it does need calibration.  I only saw a photo with no leads connected and the voltage was reading at 2.5.  Is this normal?

It might be.

If you look at the manual I posted is says the input impedance is >10 GOhms on some ranges. In one of those ranges it will be perfectly normal to show numbers on screen when the leads aren't connected to anything.



Look at the photos where you see 2.5V, if it's set to a high impedance range then worry about something else, eg. the display and whether or not it passes self-test.


My handheld meter stays pretty close to zero.....it drifts slightly but not 2.5 volts. 

Your handheld probably only has 10MOhm impedance.
« Last Edit: July 27, 2018, 09:19:10 am by Fungus »
 

Offline coromonadalix

  • Super Contributor
  • ***
  • Posts: 7958
  • Country: ca
Re: Fluke multimeter calibration
« Reply #6 on: July 27, 2018, 10:19:54 am »
Well if the meter has an very high impedance over 20 megs in the giga ohms range below 20 volts, it can be normal,  ask the seller to short the leads, if it gets to zero, it would be Okay 

High impedance can lead / or act like an antenna, even leaving a probe(s) plugged can give some parasitic values / results

Normally for the 8040 it's the ac-08 option
For the 8842 it would be the ac-09 option ??

If you don't have it, you wont be able to measure Ac voltages ... 

And finally the other option is the gpid interface board IEEE-05 option

Having those 2 board in it would be very useful to you

If you end buying it, check for bad solders on the transformer,  Redo them all, I've seen lots of complaints of flickering display brightness.
« Last Edit: July 27, 2018, 10:28:20 am by coromonadalix »
 

Offline threephase

  • Regular Contributor
  • *
  • Posts: 205
  • Country: gb
Re: Fluke multimeter calibration
« Reply #7 on: July 28, 2018, 08:30:54 am »
A 2.5V reading could well be due to a high input impedance of the meter. My 8846 when switched to high input impedance will read anywhere between 1.5V and 6.5V. It does tend to have higher readings when the leads are plugged in.

As already stated, if the input is shorted, the reading should drop to zero.

With regards to calibration, you can check the calibration yourself if you have access to a higher spec instrument, preferably with better than 10x the accuracy of the instrument you are checking, but it is acceptable to have an instrument 3x better accuracy. You then use that instrument to verify the voltage / current / resistance being applied to the instrument being checked. You will find however that measurement technique will also start to come into play, but it is a great way to learn about the instruments and their limitations.

The high spec calibrators are really just for commercial calibration labs to get the instruments through the door and back out as efficiently as possible with recorded results. You can get just as good accuracy yourself with the right application and techniques used.

If you have to adjust calibration, then that can be a different matter. You will need the specific values at the required accuracy to follow the manufacturer's procedure and you generally find that you need all of them as you have to go through the whole calibration procedure, otherwise it will be invalid. You will also need the correct stable ambient conditions.

You can of course send it out to a cal lab, but be careful what you are getting. In my country, calibration means checking, if adjustments are required, then that is an extra charge. UKAS calibration (checks) of my 8846 costs circa £120, standard calibration checks are around £80. It has never needed adjustment, so do not know what that would cost.
 