Fluke 867B - doesn't enter set-up mode
« on: August 08, 2017, 09:07:38 pm »
Hello,

I picked up a Fluke 867B, and it fails to enter set-up mode. In rest, it seems functional.
Does anyone have an idea what it could be?
I tried cleaning the contacts on both sides of the rotary switch, but the problem persists.
I attach a photo with how it looks in set-up mode.

Any ideas are welcome!

Thanks!
 

Offline Wytnucls

  • Super Contributor
  • ***
  • Posts: 3045
  • Country: be
Re: Fluke 867B - doesn't enter set-up mode
« Reply #1 on: August 10, 2017, 07:25:42 pm »
Mine has developed the same fault now. I haven't had time to troubleshoot the problem yet.
The maintenance manual is available online, if that's any help.
 

Offline mzacharias

  • Frequent Contributor
  • **
  • Posts: 710
  • Country: us
Re: Fluke 867B - doesn't enter set-up mode
« Reply #2 on: August 10, 2017, 07:35:52 pm »
Mine did that and I wasn't able to fix. I replaced a cap or two inside - no go. Sold "as-is" on eBay. Other than not entering setup, it worked fine; just could not change any default settings.
 

Offline mahi

  • Regular Contributor
  • *
  • Posts: 99
  • Country: 00
Re: Fluke 867B - doesn't enter set-up mode
« Reply #3 on: August 10, 2017, 07:45:43 pm »
Exactly the same screen is shown if you hold the dial between any two positions. So at first glance it looks like the multimeter does not detect the rotary switch is in SETUP position.

The circuit to read the dial position is quite simple so it should be easy to confirm whether the switch works okay.
« Last Edit: August 10, 2017, 07:54:07 pm by mahi »
 

Offline Wytnucls

  • Super Contributor
  • ***
  • Posts: 3045
  • Country: be
Re: Fluke 867B - doesn't enter set-up mode
« Reply #4 on: August 10, 2017, 09:13:34 pm »
Yes. The switch is a voltage divider, presenting a varying voltage on pin 83 of U25.
Hope U25 is not the culprit.
 

Offline Wytnucls

  • Super Contributor
  • ***
  • Posts: 3045
  • Country: be
Re: Fluke 867B - doesn't enter set-up mode
« Reply #5 on: August 11, 2017, 09:05:39 am »
Fixed!   :-DMM

Tested all the values of the resistors around the selector switch.
Found R166 (component test) reading a bit low, at 4.5K ohm, instead of 4.99K ohm.
Cleaned the resistor with electronic cleaner. The resistance only went up a little bit (4.6K ohm), but that was enough to activate the right sequence in U25.
« Last Edit: August 11, 2017, 09:10:31 am by Wytnucls »
 
The following users thanked this post: analogRF

Offline Wytnucls

  • Super Contributor
  • ***
  • Posts: 3045
  • Country: be
Re: Fluke 867B - doesn't enter set-up mode
« Reply #6 on: August 11, 2017, 09:19:17 am »
Vdd is 5.2V
R170 (Logic) 49K9 reads 12K0 on mine.
R168 (Setup) 6K0 reads 5K2.
All the other resistors have the correct values.

Smoothing cap C64 (Ceramic 10% 25V X7R 0805) should measure 0.1 uF.
I couldn't find TP22 (40 MIL test pad) to check the voltage to pin 83 (U25).
 (Update:Test pad 22 is on the reverse side of the PCB close to the selector switch)

I would recommend cleaning all resistors and PCB around the selector switch before replacing anything.
« Last Edit: August 12, 2017, 08:24:41 pm by Wytnucls »
 