« Reply #25 on: June 03, 2020, 07:25:27 pm »
Found this big list of model numbers and short descriptions.
https://www.davmar.org/temodel.html#fluke
Radiomuseum doesn’t have an 8125a yet!
https://www.radiomuseum.org/m/fluke_usa_en_1.html
« Last Edit: June 03, 2020, 07:43:37 pm by bluey »
Offline tggzzzTopic starter
« Reply #26 on: June 03, 2020, 07:38:14 pm »
I might as well put the 12 page "TB 9-4935-365-35" military calibration manual here, for posterity.
No, I can't remember where I got them from.
The following users thanked this post: bitseeker, alm, tonyalbus
Offline bluey
« Reply #27 on: June 03, 2020, 07:51:23 pm »
8100a manual is here - might be close
http://www.nixiebunny.com/fluke8100b/fluke8100b.html
Offline bluey
Offline bluey
« Reply #29 on: June 07, 2020, 07:30:59 am »
Fluke 1974 catalog includes 8125A. $1995 or with battery $2195!
http://doc.xdevs.com/doc/Fluke/Fluke_Catalog_1974.pdf
« Last Edit: June 07, 2020, 07:38:49 am by bluey »
Offline picburner
« Reply #30 on: June 07, 2020, 04:40:28 pm »
There is one for sale here but the lid is missing.
« Reply #31 on: June 07, 2020, 06:04:04 pm »
I suspect the mains lead is also missing, in the lid. The connector is non standard.
Offline tooki
« Reply #32 on: January 03, 2021, 12:17:17 am »
Nonstandard as in non-IEC, you mean? The connector itself is practically guaranteed to be a standard military circular connector. Easy enough to order, but very, very rough on the wallet...
« Reply #33 on: April 11, 2022, 02:05:42 pm »
Well, I sold it.
It was only taking up space in the garage, and I've got another Fluke DVM with nixies, so I preferred that someone else would "cherish it". A nice chap at bought it at a hamfest over the weekend.
Now that it has gone, of course I regret it. It was  a peculiar device, and my remaining nixie DVM is nowhere near as quirky.
Offline MegaVolt
  • Country: by
« Reply #34 on: April 11, 2022, 03:16:35 pm »
What is the reference voltage source in it?
« Reply #35 on: April 11, 2022, 04:14:11 pm »
The manual says only that Q11, R15, R22 are ordered as a matched set.
The following users thanked this post: MegaVolt
Online TimFox
« Reply #36 on: April 11, 2022, 04:16:39 pm »
The old "reference amplifier" (Q11) chips were individually characterized by Fluke for the appropriate current that gave zero tempco.
Resistor R15 sets the current, and resistor R22 gives a gross adjustment for the output voltage of the amplifier.
Offline MegaVolt
  • Country: by
« Reply #37 on: April 12, 2022, 08:04:31 am »
I have seen the diagram. This is very similar to SZA263. But maybe they used a different zener. Is its name visible in the device? Perhaps there is a photo?
Offline alm
« Reply #38 on: April 12, 2022, 11:39:18 am »
Fluke used the TI DH80417A refamp before the SZA263 in calibrators, voltage references (731A and early 731B) and multimeters (8800?) in the seventies. So that would be my guess.
« Reply #39 on: April 12, 2022, 12:56:02 pm »
I've sold the DVM, so the only photos are in this thread.
Offline alm
« Reply #40 on: April 12, 2022, 06:12:36 pm »
If the TO-39 IC in the attached crop from one of tggzzz's pictures is indeed the voltage reference, then it might be indeed the DH80417A or DH80417B refamp. The picture is too low resolution to tell for sure, but it seems to have a logo on top and marking on the side, like in the pictures here and here (look for 731A pictures).
Offline Dr. Frank
« Reply #41 on: April 12, 2022, 07:24:30 pm »
Reference Amplifier DH80417B from my old 332B/AF:
https://www.eevblog.com/forum/testgear/fluke-332baf-in-the-slaughterhouse/msg393627/#msg393627
This marking is different, and not centered on top. The component seems to be from Siemens, that's what the logo looks alike.
A bigger capital S and a smaller capital H, for Siemens & Halske. That logo was used on the TCA365 also, and Siemens made OpAmps in the seventies, afair.
It's better visible, when you download the jpg and zoom in.
Frank