Thanks for the manual.
OK an update since realising that my measurements were not with respect to the common (neg probe) terminal)

As per 4-16 in the series 1 service manual, with meter on the Vdc setting:
TP1 measures 3.05v
TP3 measures 2.5v (included although I didn't notice a mention of it in the manual)
The crystal is oscillating at 32768 hz and the sine wave amplitude I measure at 348mV on my meter and on the scope it looks like about 1v p-p which is a little high according to the manual
1v ref voltage looks ok
V Middle is 1.6v
Couple of other checks I made
Battery draw is 426 uA
Also getting a clock in to U2 on pin 23
Also just noticed that the display bezel is cracked
Thinking that the display might have been damaged during disassembly I also tried testing the continuity buzzer but it didn't peep.
Anything else I can check