Hello,
a few weeks ago I scored a Fluke 3330B on ebay. The seller said, that the switch on the
very left side for selecting the voltage/current does not work in position "8".
It turns out, that the switch is working perfectly well but the relay K3 on the A4A10
ladder driver no. 1 is hanging. I got the relay out, hooked it up to an external power
supply and after some on/off switching it was working again. Now the 3330B was able again
to select all voltages between 0 and 1111.111V. The current mode I did not check yet.
Then I set it to 1mV and checked with my 34401A and saw some dancing in the 10µV range.
I looked into the schematics and checked the voltage reference board. All voltages stood stand
still perfectly, so I checked the following chopper amplifier and differential amplifier.
On both boards were several electrolytic caps and one tantalum. I changed them all and the
dancing was gone, except for a +/- 1µV (the 3330B was working then for several hours).
To me this seems to be pretty good :-)
But this brings me to this: while fiddling with the cards in the 3330B I learned,
that there is something like an extender or riser board. But those boards seems to be
pretty hard to find, so I found this:
So i made the extender with connectors from the parts from Frank's 332B:
Does anybody know what's the name of these golden connectors (brand, name etc.) and probably where to buy?
I would like to build a riser board of my own, so that measuring will be much easier.
Every hint which may come will be helpful and really appreciated.
Thanks a lot.
Btw, I found also this very nice website with another 3330B repair attempt:
http://www.spurtikus.de/basteln/repair/fluke_3330b/index.htmlThe site is in german but the pictures are really nice.