Some time ago I bought a fluke 189 for cheap. Was advertised as non functional but after some troubleshooting I managed to get it working.
On first inspection I noticed that 2 case screws were missing, the battery terminals seemed to have leakage corrosion, the high amp fuse was not the correct one (25A instead of 11) and the 440mA fuse was blown. Despite this I proceeded with the preliminary inspection.
First I decided to address the leakage in the battery terminals. I cleaned them a little with a cotton swab and vinegar to discard battery power issues and begin testing. After cleaning the terminals I put on some batteries and the meter turned on! But alas there were more issues...
When putting the meter in the different functions via the rotary knob the screen showed some segments, but the display then started to fade. I suspected dirty screen contacts or some issue with the LCD power supply. I then decided to take the meter apart.
After dissasembling the case I noticed that one screw post was broken. I checked the board but the meter did not seem to have any visible electrical damage (I even checked the supercap for leakage or shorts, but it was fine). However, two of the screen tabs were broken.
I suspected that was part of the LCD issue, so I cleaned the contacts anyways and taped the tab sections of the screen to the board using electrial tape. I put the meter back together and used the two screws on the top of the case to make sure that it was tight enough for the screen to make contact with the PCB.
Success! after putting the batteries on again the display worked! It displayed all segments and every function appeared correctly on the display. However the switch function button didn't seem to work. I plugged in some leads and started making some measurments...
But no reading was shown on the meter, it appeared as if I weren't measuring anything. Voltage oscillating near 0 when in the voltage ranges, OL when in ohms, open when measuring temperature. However, the amp ranges seemed to detect lacking fuses and incorrect lead placement.
Since everything seemed to work well I suspected lead continuity to the meter, so I opened it again, connected a lead to eack jack input and tested continuity to the pcb traces to which they were connected. The A/mA, uA and V jacks seemed to test ok, but the COM jack tested open. I soldered a wire to the COM trace to make low voltage, resistance and capacitance measurments, and it worked! So I decided to take a closer look at the input jack assembly.
On closer inspection i discovered that it was damaged, and it seemed a repair was attempted using superglue. I researched more on jack assembly issues and stumbled upon MrModemhead's article on Fluke 80 series jack repair (
http://mrmodemhead.com/blog/fluke-80-series-jack-assembly-rebuild/), which seemed very similar to the one on the 189.
I decided to desolder the jack assembly and see if the issue was similar to the one on MrModemhead's article (this was one of the most time consuming parts of the project). Both tabs of the COM jack and one of the V jack were broken, as well as the plastic sleeves of both.
I rebuilt both jacks as per the instructions on MrModemhead's site and the meter started showing readings taken with leads connected to its inputs.
I fixed the non working switch function button with some double sided tape and aluminum foil, and I managed to clean the case using only dish soap, water and a brush.
I attached before and after pictures of the meter. In the second picture the meter is measuring a 5v LT1236.
I will add more posts with more pictures and descriptions of the process.