I just came across this FLUKE 1630-2 Earth Ground Clamp for loop resistance.
This must be one of the most clever instruments I have come across in years.
It looks like a clamp meter but it measures impedance.
https://www.fluke.com/en/product/electrical-testing/earth-ground/fluke-1630-2-fc
What I wonder is how this instrument works.
Maybe one coil introducing a current in to the conductor under test and another coil is measuring the response?
Anyone here having used such instrument?
Offline Grandchuck
Re: FLUKE 1630-2 Earth Ground Clamp for loop resistance
« Reply #1 on: May 06, 2023, 12:35:21 pm »
Here is a clue:
« Reply #2 on: May 06, 2023, 01:00:52 pm »
Yes. The resistance mode uses an AF pulse in the coil. The coil has to be completed when the jaws close, so unlike common current clamps these have fine leaves of metal (not sure what, could be stainless steel or plated copper) which mesh together to make the circuit when the clamp jaws close.