Author Topic: Flood of new Agilent 82357B GPIB-USB adaptors on eBay the real deal?  (Read 171544 times)
Offline HighVoltage
Re: Flood of new Agilent 82357B GPIB-USB adaptors on eBay the real deal?
« Reply #375 on: October 06, 2018, 09:20:26 am »
I did not expect this....
I just looked at my original genuine "Keysight" branded 82357B and the "Functional Test Certificate" that came with it. Interestingly enough, the serial numbers don't match !
The serial number on the device can be traced on the Keysight website.
MY563...
The serial number on the certificate can NOT be traced on the Keysight website.
MY555...
Hmmm, who knows, may be something strange is still going on with this device at the Keysight production or packaging facility.
The following users thanked this post: Rob2
Offline Zucca
  • EE meid in Itali
« Reply #376 on: October 08, 2018, 02:00:42 pm »
Seems like Keysight has a feeling for that  ^-^:
Source:
http://literature.cdn.keysight.com/litweb/pdf/82357-90003.pdf
More over RTFM, it looks like the serial number is used in the SW, so it must be stored in somewhere inside the 82357B.
« Last Edit: October 08, 2018, 02:02:31 pm by zucca »
Offline tkamiya
« Reply #377 on: October 08, 2018, 08:36:16 pm »
I had a conversation with an owner of a local calibration lab.  I told him I ended up with a counterfeit HP82357B.  He laughed and said all of them are fake!  (I'm sure he meant on secondary market)  According to him, they usually work just as well and so much cheaper.  I guess this is a well known thing in the industry.
I guess there are some good fakes and bad fakes?
He showed me a copy of an HP cable that even said "Made in USA".  According to him, it's a fake.  Whomever copied the original did so EXACTLY.
Offline bitseeker
« Reply #378 on: October 09, 2018, 02:03:16 am »
Possible. Even in the earlier discussions of this adapter, there were photos of some units that were more obviously fake. In other product categories (e.g., watches or handbags), there are different grades of counterfeits, with corresponding prices reflecting the quality of "fakemanship."
Offline HighVoltage
« Reply #379 on: October 09, 2018, 08:11:34 am »
Interesting, Zucca
In Figure 1-1 of this manual page, they forgot to show the  "Functional Test Certificate"
And in the text above it, they call it the "82357B Certificate of Calibration"
With all this confusion, it seems easy to bring fake items on to the market.
As to the working ability of the fake adapters:
I had two of them and one was working perfectly and the other one always gave me problems and I thought it was software related, until this thread was opened.
But at the beginning of this thread, I still believed my fake one was real, until I opened them all! :-DD
« Last Edit: October 09, 2018, 08:13:24 am by HighVoltage »
Offline ekeppel
« Reply #380 on: December 09, 2018, 01:59:34 am »
MY563...
MY555...
@HighVoltage
Let me preface this post by saying that I have *not* had occasion to examine a genuine Keysight-branded GPID adapter personally, so take this with a grain of salt please. :)
Having said that, did you happen to open this one up and see if everything is up to Keysight standards?  I'm not entirely clear from your post if this was purchased directly from Keysight or through another vendor.  It may be my own 'counterfeit paranoia' talking, but to me something just doesn't look right about the way that copper paint is applied near the screw.  I've noticed that counterfeiters will often modify their products to make them appear more legit once notable differences from the real thing become known to consumers, even gleaning such information from forums like this one, so...
I don't know why, but the impression that I get when I see your second photo is that the paint was sprayed on there 'for show', and may not truly extend to the inside of the device.  On the less-paranoid side, the differences in paint application from the genuine Agilent branded ones that I've seen may be simply due to slight changes made when the Keysight re-branding took effect. :-//
I hope that I'm proven wrong here, since it would eliminate the easiest method (that I'm aware of) for differentiating the genuine article from the fakes without opening the case.
On the topic of fakes, it's getting really bad out there with counterfeits, even down to the component level.  I recently ordered some ST-branded 2N3055s on eBay, and they were all fake.  I still had a few genuine ones from Mouser to compare to, and even the external differences were obvious.  Sure enough, though rated for over 15A, the eBay ones consistently blew at just a couple of amps when tested under load.  :--  Had I used them to repair the PSU I was working on at the time, they would have likely fried anything hooked it.  I went looking around online, and sure enough there are YouTube videos of folks cutting open the metal cases of these things to see what was inside.  Compared to the real ones, the fakes contained only a tiny undersized piece of silicon with no chance of handling their rated current.
I now buy components like that only from a trusted source such as Digi-Key, Mouser, etc.  Lesson learned.  Unfortunately, most hobbyists such as myself can't afford to pay Keysight's retail price for something like this GPIB adapter, so we're forced to navigate the land of make-believe in hopes of finding the real thing.   :palm:
« Last Edit: December 09, 2018, 02:05:55 am by ekeppel »
Offline coromonadalix
« Reply #381 on: December 09, 2018, 03:39:15 am »
Are the prologix gpib adapters compatible with hp equipement  ?,  if so im not aware of any counterfeit item in this brand ?
« Reply #382 on: December 09, 2018, 04:27:36 am »
When a US company that has no Chinese distributors and you can buy a product for 25% less than from the manufacturer out of China I suspect there may be a bit of an issue with cloning.
Online TheSteve
  • Living the Dream
« Reply #383 on: December 09, 2018, 04:32:36 am »
gpib = gpib, so they they will talk to the equipment no problem. However if you wanted to use any HPAK produced software etc it won't work with Prologix adapters.
VE7FM
Offline alm
« Reply #384 on: December 09, 2018, 09:34:34 am »
The instrument side will be very similar and mostly compatible (IEEE 488.1 and related standards). The computer side not so much. There are some industry standards here (the gpib-32 low-level library, the GPIB resource in VISA), but last time I checked, Prologix didn't implement any of them.
This means that generally software developed for the Prologix interface will not work without modifications on a Keysight / NI / ICS / Tektronix interface, and the other way around. The latter includes pretty much any software made by an instrument manufacturer (e.g. BenchVue), any LabVIEW driver and IVI drivers. Open source / hobby initiatives seem split, but in general the industry standard ones seem to be better supported there.
Offline HighVoltage
« Reply #385 on: December 09, 2018, 10:02:31 am »
Yes, I did open this one up and it is a genuine Keysight part that came from a reputable distributor.
Probably some unlikely but possible paperwork mixup at the Keysight factory.
Offline ekeppel
« Reply #386 on: December 09, 2018, 11:41:55 pm »
Excellent.  So at least for now we are safe if we see the copper overspray.  Thanks! :-+
Offline macboy
« Reply #387 on: December 10, 2018, 05:05:13 pm »
The Prologix adapters can talk to the instruments, but can your software talk to the Prologix adapter? Every other GPIB adapter on the market, from Agilent, NI, or whoever, comes with APIs and DLLs conforming to standards, and this allows compatibility between software and adapters. Prologix is the exception, they don't have a VISA layer or a GPIB-32.DLL or any other standardized interface. Some homebrew software supports the Prologix adapters but don't expect most professional software to do so.
In this case, I'd lean toward the cloned 82357b. It may be fake, but it is a 1:1 copy, including the GPIB bus driver ICs. These are glaringly missing from the Prologix one. Instead, it depends on the logic levels of the microcontroller inputs (are they true TTL levels as required?), and the drive strength of the micro's outputs (can they sink 48 mA of current on all lines at the same time? Hm...), and the ESD input protection of the micro as well. These are the things that those bus driver ICs are designed to handle.
The following users thanked this post: E Kafeman
Offline Mr Simpleton
  • Not the sharpest knife in the drawer
« Reply #388 on: February 14, 2019, 09:59:43 am »
 |O
Bugger, I was one of those fooled by the hype and got one of the 82357B from Nettcy168... and of course it was a fake. Now while waiting for the darn thing to show up I managed to get a used NI-USB HS dongle that works like charm. So when the Agilent fake showed up it was put on the shelf and promptly forgotten (should have initiated a refund but...), until yesterday. I had some work done using a 82357A, and suddenly remembered my fake one. Interestingly this one is not even detected by my computer.... Dead as a door nail.  Should keep me busy all friday evening looking to see why....
They still sell this crap on e-bay and they have upped the offer by showing "Functional Test Ceryificate". Never got a ceryificate with mine so bets are thats why it doesn't even show up at the device manager.
Offline HighVoltage
« Reply #389 on: February 14, 2019, 10:12:48 am »
It is a surprise to me that after so many years, Keysight is not doing anything against these fake offers.
« Reply #390 on: February 14, 2019, 10:20:42 am »
Given they are a relatively simple device perhaps if Keysight were less greedy than $900 AU Pesos retail we may not buy the clones at $130 AU pesos.
Much as mine worked fine and continues to do so having Paypal as a backstop for evilbay's claims process makes it a reasonable gamble if you get a Dud.
Personally if the cloning clowns produced a functional equivalent in a different wrapper I would still buy one to give it a go so it is a shame they think copying is the way to the most sales.  ::)
« Reply #391 on: February 14, 2019, 10:54:57 am »
Could not resist cracking this open... and was surprised how "neat" the inside looked. Much much better workmanship than some showed us years ago. Still lacking the metalization on the inside, and the PCB clearly is a fake, but soldering looks OK, but a few puddles of soldering flux.
Juggling the mini-USB connector brought the bugger back to life, and I managed to connect to a HP8753E. Closer inspection shows a  very very crappy connector on the cable end. Will be interesting to see how long the interface will stay OK :) For any critical work I'll reveret back to my NI USB-HS...
Offline HighVoltage
« Reply #392 on: February 14, 2019, 10:55:27 am »
For those of you, who are interested:
I am selling my genuine National Instrument GPIB to USB Adapter
https://www.eevblog.com/forum/buysellwanted/(fs)-national-instruments-gpib-to-usb-adapter-new-ni-187965b-01/
Offline bitseeker
« Reply #393 on: February 15, 2019, 01:45:21 am »
Yeah, I had a good laugh when I first saw those showing up. Some also use the wrong fonts. I reported the fakes to eBay for over a year, but they can't (or won't) do anything about it, even though I tell them why it's obviously fake. So, I gave up.
Offline rkeller
« Reply #394 on: April 25, 2019, 11:53:41 am »
Did we ever come up with a consensus as to what GPIB to USB was worth buying for under $200 US?
I see the real deal on Keysight and Newark’s websites for $594.
https://www.keysight.com/en/pd-851808-pn-82357B/usb-gpib-interface-high-speed-usb-20?cc=US&lc=eng
https://www.newark.com/keysight-technologies/82357b/high-speed-usb2-0-gpib-interface/dp/31M9424
But... is there a reasonable clone or copy that works?
Offline HighVoltage
« Reply #395 on: April 25, 2019, 12:48:57 pm »
The clones seem to work ok under windows but have problems under linux.
I suggest to get a real Keysight or Agilent Adapeter. This way you will not need to worry if your possible problems come from the adapter.
Offline 2N3055
  • Country: hr
« Reply #396 on: April 25, 2019, 01:15:17 pm »
One I have is completely indistinguishable from original and works perfectly in Linux an windows. I use it with Raspberry PI no problem too..
Offline rkeller
« Reply #397 on: April 25, 2019, 09:45:44 pm »
Which one might that be?
Offline 2N3055
  • Country: hr
« Reply #398 on: April 25, 2019, 10:22:43 pm »
This one..
Offline notfaded1
« Reply #399 on: August 27, 2019, 04:21:04 pm »
I decided to look for a used real Keysight labeled GPIB.  I think I found one and it's on it way from guess where?  Fingers crossed it's real.  It's hard to tell from the pictures but I'm pretty sure it's got the shielding around the thumb screws and he swears it was a real full priced Keysight he bought with a MM from a distributor.  I tried hard to get him budge on price but 110$ delivered was as low as he would go.  He has a LOT of real Keysight hardware as well and I've seen pictures of them so after a lot of back and forth I think I'm in the money.  I was going to buy a fake but I'd rather have a real one.
Regards,
Bill
.ılılı..ılılı.
notfaded1