Author Topic: Enabling options on Agilent ESA series E4402B E4404B E4405B E4407B  (Read 109704 times)
Online tv84
  • Country: pt
Re: Enabling options on Agilent ESA series E4402B E4404B E4405B E4407B
« Reply #100 on: September 11, 2020, 06:17:29 pm »
My memory dump is attached.
FYI, my unit is running the latest A.14.06 firmware.
In binary format.
This Andrews's dump is a dump of ESALOADR (size 0xB5B48 bytes). Not ESAFW.
« Last Edit: September 11, 2020, 06:26:05 pm by tv84 »
Online tv84
  • Country: pt
« Reply #101 on: September 11, 2020, 06:28:46 pm »
what if we powered up normally.  once up and firmware loaded we insert the ESALOADER disc and reset
DRAM should still be loaded.
we can then break into monitor and try to dump memory?
That is a nice idea. Please try it.
Online Gribo
« Reply #102 on: September 11, 2020, 06:42:25 pm »
Ctrl+C is also the Break key, you can try that.
I am available for freelance work.
Offline smgvbest
    • Kilbourne Astronomics
« Reply #103 on: September 11, 2020, 07:02:13 pm »
Not working.
What BootRom Version are you on
Mine is E4401 Bootrom, 5.00
Sandra
(Yes, I am a Woman :p )
Offline smgvbest
« Reply #104 on: September 11, 2020, 07:07:40 pm »
Just to be sure we're all on same Page
This is the Debug menu and not what we need
----- System/pSOS Debug commands: -----
    '?' - this help message.
    'j' - drop into breakpoint.
   '^C' - Abort to monitor.
   '^P' - Process status info, and LOTS of it.
 '[dD]' - Print DLP debug information.
 '[bB]' - Big memory hog report.
 '[pP]' - Process ONLY status info.
 '[eE]' - Exchange info.
 '[gG]' - toggle breakpoint exception handlers on/off
 '[tT]' - Time log.
 '[hH]' - History log.
 '[oO]' - Memory segment ownership.
 '[mM]' - Memory segment summary.
 '[sS]' - Semaphore ownership, etc.
 '[uU]' - maximum process stack Usage.
 '[vV]' - memory Validity check.
 '[iI]' - Show psosSystemData.
 '[1]' -  Show NVRAM contents.
 '[wW] <process name>' - Show process stack trace.
This is the Monitor Menu and what we're after.
bc      [<hex boot config>] - set the bootrom configuration (see bchelp)
bootvars- display bootrom variables
bs      - force a breakpoint when starting
dbyte   [<hex start address> [num bytes]] - display memory using bytes
dlong   [<hex start address> [num bytes]] - display memory using longs
dmem    [<hex start address> [num bytes]] - display memory using bytes
dword   [<hex start address> [num bytes]] - display memory using words
gbreak  - force a gdb breakpoint
gdb     - enable gdb trapping of exceptions
gu      [<hex start addr>]      - go to start address
hmon    [device] - download into memory
rty test routine
sbyte   <hex start address> <hexchars> - set memory using bytes
slong   <hex start address> <hexchars> - set memory using longs
smem    <hex start address> <hexchars> - set memory using bytes
sword   <hex start address> <hexchars> - set memory using words
version - display bootrom version
Sandra
Online tv84
  • Country: pt
« Reply #105 on: September 11, 2020, 07:19:24 pm »
Please dump from 0x0400 0000 to 0x0401 1000.  Just to check if this is BOOTROM-related.
I think this can be done with ESALOADR. No need for ESAFW.
Online tv84
  • Country: pt
« Reply #106 on: September 11, 2020, 07:20:50 pm »
How do you trigger DEBUG MENU?
For information:
DEBUG MENU is inside ESALOADR
MONITOR MENU is inside BOOTROM   (called from within DEBUG MENU via ^C)
« Last Edit: September 11, 2020, 08:02:56 pm by tv84 »
Offline andrew9875
« Reply #107 on: September 11, 2020, 08:12:52 pm »
Using this procedure:
  • Boot from ESALOADR floppy, drop into ROM monitor menu (j, CTRL+C)
  • Retry test routine ('rty')
  • Test routine will fail/hang, then remove floppy and power cycle the unit
  • Boot from flash will fail and drop you back into ROM monitor
My memory dump (0x04011000-0x04900000) looks nearly identical to the concatenated ESAFW file that Sandra shared earlier with only a handful of addresses differing, and differs quite a lot from the ESALOADR.
I believe this dump is the ESAFW from flash. On the second boot (step 4 of my procedure), the ESALOADR disk is not present so the unit must load from flash.
Offline smgvbest
« Reply #108 on: September 11, 2020, 08:15:20 pm »
when you boot normally the Debug Menu "----- System/pSOS Debug commands: ----- " is always there.
when you boot from the ESALOADER you don't initially see a menu.   Pressing ? when show same menu above, you can CTRL+C and it causes an exception then you're in the Monitor menu where you can dump memory
I've hooked up a reset switch,   booted normally,   entered AXZ for the opt and 0123456789ABC for the code.
inserted the ESALOADER,  and hit reset.
once rebooted I got into the mornitor and did 2 dumps (ones still dumping)
first is will be
second will be
0x0401 1000 to 0x0490 0000
soon as done i'll edit and post to this message
My HostID is 29611027
My BootRom is V5.00
Sandra
Online tv84
  • Country: pt
« Reply #109 on: September 11, 2020, 08:32:11 pm »
This dump is exactly Sandra's ESAFW with 8 bytes different (in the middle of the code  :-//).
The problem is that you didn't run the app before taking the dump. We need the dump after the app has run/is running. Because all the rest of the mem is 0x00s.
Online tv84
  • Country: pt
« Reply #110 on: September 11, 2020, 08:38:49 pm »
It MUST be possible to abort into Monitor mode from within ESAFW because ESAFW also has the DEBUG MENU in the code (with the ^C option). Just saw that in the code.
Offline andrew9875
« Reply #111 on: September 11, 2020, 08:41:18 pm »
Got it, didn't quite grasp what the issue was before. The application was definitely not running when I created the new dump.
Looking forward to seeing Sandra's results
Offline smgvbest
« Reply #112 on: September 11, 2020, 08:41:40 pm »
When I boot from the ESALOADER and see the Menu ^C works fine
when I boot normally and see the Menu,  even though ^C is in the menu all I see is a message that it doesn't recognize the keypress
Sandra
Online tv84
  • Country: pt
« Reply #113 on: September 11, 2020, 08:45:28 pm »
Ohhh. That's another story. Let's wait for your dumps. I've got my fingers crossed.
If not successfull, I'll ask you to do me a log dump of all the submenu options of the DEBUG MENU (just the 1st screen) so that I can crosscheck the functions.
Offline smgvbest
« Reply #114 on: September 11, 2020, 08:53:04 pm »
still dumping,  it's up to 0x040Dxxxx and it's all zero's in this area
i'm trying to figure our device addressing
the max address is 0x07FFFFFF  only ADDRESS BIT 0..27 are used
21,22,23 are used to address the FLASH memory (thru a 74138)
U55 controls the addressing which is the communications controller which goes to the enable pin on the 74138
Sandra
Online tv84
  • Country: pt
« Reply #115 on: September 11, 2020, 09:01:36 pm »
What is the specific msg?
Online tv84
  • Country: pt
« Reply #116 on: September 11, 2020, 09:05:23 pm »
I saw somewhere that the app and mem would not go upper that 0x06000000. Maybe in your boot logs...
That doesn't mean that physically it couldn't go to that limit mentioned by you.
Offline smgvbest
« Reply #117 on: September 11, 2020, 09:09:56 pm »
From the hardware side it's physically limited to 0x07FFFFFF (A0..A27) the remaining bits are NC
Sandra
Offline smgvbest
« Reply #118 on: September 11, 2020, 09:14:03 pm »
Unknown debug char: 'C' (0x43).  Press '?' for help.
if I do CTRL+C its
Unknown debug char: ' ' (0x03).  Press '?' for help. (i think)   still dumping so I can't check
I do not miss dialup speeds
Sandra
Offline smgvbest
« Reply #119 on: September 11, 2020, 09:16:17 pm »
There are no sub menus
each option is a direct action in the menus
the may take parms but that is passed with the option you want
Sandra
Offline smgvbest
« Reply #120 on: September 11, 2020, 09:27:28 pm »
Is anyone able when booting normally (no ESALOADER Disc) able to do the CTRL+C and enter the Monitor Program?
Sandra
Offline smgvbest
« Reply #121 on: September 11, 2020, 09:30:02 pm »
Sure, all I want is a dump of the 1st page of those direct actions (just to identify the strings in it).
You don't need to try other keys. The DEBUG in ESAFW doesn't have the code for "^C". So we would need to patch it.
BTW, the DEBUG MENUs of ESAFW and ESALOADR have slight differences. I think ESAFW has more options.
I do have some spare FLASH memory for the BootLoader so if a patch isnt to hard maybe thats the way to go???
These are the menu options seen when booting
This is the Debug menu when booting Normally, you have to do a ? to see it
 '[tT]' - Time log.
This is the Monitor Menu from when using ESALOADER.
rty test routine
« Last Edit: September 11, 2020, 09:32:13 pm by smgvbest »
Sandra
Offline suj
« Reply #122 on: September 11, 2020, 09:30:17 pm »
I tried but it did not work.
I didn't get FDD todayso I still have older firmware.
Online tv84
  • Country: pt
« Reply #123 on: September 11, 2020, 09:31:38 pm »
that was these then
No, maybe I explained wrong:
I want the next steps in each of those options (just DEBUG MENU).
Offline smgvbest
« Reply #124 on: September 11, 2020, 09:37:12 pm »
that was these then
if I type a option it executes the command so if thats what you want heres some
is this what you're after?
***** Mosquito Bootrom *****
Copyright 1988-1997,
Hewlett-Packard Company, all rights reserved.
@(#)HEWLETT-PACKARD, E4401 Bootrom, 5.00
@(#)LDS Rev: 3.02 - Module Incremental (Sep  9 2003)
@(#)Linked: Sep  9 2003 14:46:44
Bootrom Checksum ...
Bootrom DRAM:     Testing 69632 bytes at 0x04000000
Non Destructive SRAM Test ...
Main Firmware DRAM:     Testing 33484800 bytes at 0x04011000
Main FW Checksum ...
Self-tests complete.SRAM selftest results:
        Start  = 0xa000000
        End    = 0xa007fa3
        Errors = 0x0
DRAM selftest results:
        Start  = 0x4011000
        End    = 0x6000000
        Errors = 0x0
hpibPort = 0x8005000
hpibPort = 0x8005000, bus Address = 19
Cache Enabled
16MBytes of FLASH
Download to Flash Selected
>>> mainMain()
text segment:           0x4011000 thru 0x4435e14 ( 424e14 bytes)
data segment:           0x4600000 thru 0x476dd88 ( 16dd88 bytes)
bss  segment:           0x476dd88 thru 0x48bcce8 ( 14ef60 bytes)
ROM size:               0x00592b9c ( 592b9c bytes of 4194304 max.)
memory pool (all):      0x048bcce8 thru 0x05ffffff (24392472 bytes)
Calling start_psos() ...
>>>> debug() process starting
DLP Loaded - Power Suite Utilities, A.06.05, Nov 21 2003 15:45:40
 '[tT]' - Time log.
 '[9]' -  Show Exception Report.
>d
                                DLP LIST
Name         State      Text            Data            BSS
c:dlp\ps2\ps2.o Loaded  0x5c3dcc8/1751472       0x5c3bcb4/8192  0x5b6abe0/856256
c:dlp\pn\pn.o Unlicensed        0x0/0   0x0/0   0x0/0
c:dlp\catv\catv.o Unlicensed    0x0/0   0x0/0   0x0/0
Currently 1 DLP's loaded
>b
Memory HOG report - oink oink
caller PC    count      bytes
0x04338902   17072    5349060
0x04361be2       1    1751492
0x00000000     111    1049872
0x04361bfe       1     856276
0x05d00d06     194     798504
0x05d01948      83     341628
0x043ebcfe      83     294836
0x05cefb16      61     245220
0x05cf28b6      40     160800
0x05ceceb2      33     132660
0x05cf564a      15      60300
0x043eb0f0     339      39892
0x042b66a2       1      32792
0x05cea37a       7      28140
0x04345af2      53      26632
0x05d0235c       3      12348
0x042def9e       1      11108
0x04361bf0       1       8212
0x042ec426       1       2596
0x042ec3e2       1       2148
0x0412a3d0       1       2068
0x042ec404       1       1196
0x04364636       1       1032
0x042ec3c0       1       1028
0x042def8c       1        812
0x0414b2a2       3        624
0x0a0008f6       1        532
0x0a000446       1        532
0x0a0005ea       1        532
0x0a000626       1         84
>p
    pid    PNAME  STAT/M PRI GID POS  TIX MEMORY STK CPU
0x048ca38c  SWFI   RUN    51  0   *     1    0kB  8% 25%
0x048cc0f0  AAFI   RDY    51  0   1     1    0kB 12%  0%
0x048c9714  IDLE   RDY     0  0   2     1    0kB  6%  8%
0x048ccf30  CLOK  paus   100  0   .     0    0kB 16%  0%
0x048c9ffc  DIst  paus    52  0   .  3106    0kB  3%  0%
0x048c9aa4  DRST  paus    80  0   .    61    0kB 47%  0%
0x048c99c0  FMOT  paus   100  0   .    43    0kB 47%  0%
0x048cd014  UPDT  xblk   249  0   .     1    0kB 23%  0%
0x048cce4c  MAXM  xblk    60  0   .     1    0kB  6%  0%
0x048ccc84  LLMR  xblk    60  0   .     1    0kB  6%  0%
0x048ccba0  PRNT  xblk    60  0   .     1    0kB  8%  0%
0x048ccabc  DSPF  xblk    51  0   .     1    0kB  5%  0%
0x048cc9d8  DSPM  xblk    60  0   .     1    0kB  6%  0%
0x048cc8f4  DMFI  xblk    51  0   .     1    0kB  5%  0%
0x048cc810  DMMR  xblk    60  0   .     1    0kB  6%  0%
0x048cc648  FCFI  xblk    51  0   .     1    0kB  5%  0%
0x048cc564  FCMR  xblk    60  0   .     1    0kB  6%  0%
0x048cc480  ANSQ  xblk    52  0   .     1    0kB 13%  0%
0x048cc39c  ANFI  xblk    51  0   .     1    0kB 13%  0%
0x048cc2b8  ANMR  xblk    60  0   .     1    0kB  6%  0%
0x048cc1d4  AASQ  xblk    52  0   .     1    0kB 21% 28%
0x048cc00c  AAMR  xblk    60  0   .     1    0kB  6%  0%
0x048cbf28  SYMR  xblk    60  0   .     1    0kB  8%  0%
0x048cbe44  SGMR  xblk    60  0   .     1    0kB  6%  0%
0x048cbd60  ZMKR  xblk    60  0   .     1    0kB  6%  0%
0x048cbc7c   MKR  xblk    60  0   .     1    0kB  6%  0%
0x048cbb98  DEF3  xblk    51  0   .     1    0kB  5%  0%
0x048cbab4  SNFI  xblk    51  0   .     1    0kB  5%  0%
0x048cb9d0  SNMR  xblk    60  0   .     1    0kB  6%  0%
0x048cb8ec  DEF2  xblk    51  0   .     1    0kB  5%  0%
0x048cb808  LGDT  xblk   251  0   .     1    0kB  8%  0%
0x048cb724  LGDE  xblk    60  0   .     1    0kB  6%  0%
0x048cb640  DSFI  xblk    51  0   .     1    0kB  5%  0%
0x048cb55c  LGDT  xblk   251  0   .     1    0kB  8%  0%
0x048cb478  LGDS  xblk    60  0   .     1    0kB  6%  0%
0x048cb394  SWFI  xblk    51  0   .     1    0kB  5%  0%
0x048cb2b0  LGST  xblk   251  0   .     1    0kB  8%  0%
0x048cb1cc  LGSW  xblk    60  0   .     1    0kB  6%  0%
0x048cb0e8  DEFI  xblk    51  0   .     1    0kB  5%  0%
0x048cb004  DEMT  xblk   251  0   .     1    0kB  8%  0%
0x048caf20  DEMR  xblk    60  0   .     1    0kB  6%  0%
0x048cae3c  DMZF  xblk    51  0   .     1    0kB  5%  0%
0x048cad58  ZDMT  xblk   251  0   .     1    0kB  8%  0%
0x048cac74  ZDMR  xblk    60  0   .     1    0kB  6%  0%
0x048cab90  SIFI  xblk    51  0   .     1    0kB  5%  0%
0x048caaac  SIMT  xblk   251  0   .     1    0kB  8%  0%
0x048ca9c8  SIMR  xblk    60  0   .     1    0kB  6%  0%
0x048ca8e4  DZFI  xblk    51  0   .     1    0kB  5%  0%
0x048ca800  DZMT  xblk   251  0   .     1    0kB  8%  0%
0x048ca71c  DZMR  xblk    60  0   .     1    0kB  6%  0%
0x048ca638  DSFI  xblk    51  0   .     1    0kB  5%  0%
0x048ca554  DSMT  xblk   251  0   .     1    0kB  8%  0%
0x048ca470  DSMR  xblk    60  0   .     1    0kB  6%  0%
0x048ca2a8  SWMT  xblk   251  0   .     1    0kB  8%  0%
0x048ca1c4  SWMR  xblk    60  0   .     1    0kB  6%  1%
0x048ca0e0  MIME  xblk    79  0   .     1    0kB  6%  2%
0x048c9f18  FPLP  xblk   250  0   .     1    0kB  9%  0%
0x048c9e34  DCAS  xblk   251  0   .     1    0kB  8%  0%
0x048c9d50  RLCN  xblk   230  0   .     1    0kB  8%  0%
0x048c9c6c  REMT  xblk   250  0   .     1    0kB  5%  0%
0x048c9b88  PCKB  xblk   230  0   .     1    0kB  8%  0%
0x048c97f8  DISP  xblk   253  0   .     1    0kB  8%  4%
0x048c98dc  APPS  xblk   230  0   .     1    0kB 25%  0%
0x048c9630  ROOT  xblk   230  0   .     1    0kB  4% 32%
           >> EVENTS: W(0x0) S(0x2000)
        64 Process(s) (27 avail); Total time: 4569 ticks.
>e
    xid    XNAME  TYPE  ACC  maxQ  Qlen BLOCKED
0x048c1584        fifo  any    1    0    REMT
0x048c1562  UPDI  fifo  any    1    0    UPDT
0x048c1540  MAXM  fifo  any  inf    0    MAXM
0x048c151e  Sign  fifo  any  inf    0
0x048c14fc  LIMI  fifo  any  inf    0    LLMR
0x048c14da  DSPM  fifo  any  inf    0    DSPM
0x048c14b8  DSPS  fifo  any    1    0
0x048c1496  DSPF  fifo  any  inf    0    DSPF
0x048c1474  SPEC  fifo  any    1    0
0x048c1452  SPEC  fifo  any    1    0
0x048c1430  DMMR  fifo  any  inf    0    DMMR
0x048c140e  DMFI  fifo  any  inf    0    DMFI
0x048c13ec  SPEC  fifo  any    1    0
0x048c13ca  FCMR  fifo  any  inf    0    FCMR
0x048c13a8  FCFI  fifo  any  inf    0    FCFI
0x048c1386  SPEC  fifo  any    1    0
0x048c1364  ANSQ  fifo  any  inf    0    ANSQ
0x048c1342  ANFI  fifo  any  inf    0    ANFI
0x048c1320  ANFS  fifo  any  inf    0
0x048c12fe  ANOW  fifo  any   20    0
0x048c12dc  ANMR  fifo  any  inf    0    ANMR
0x048c12ba  SPEC  fifo  any    1    0
0x048c1298  AARS  fifo  any    1    0    AASQ
0x048c1276  AAFI  fifo  any  inf    0
0x048c1254  AAFS  fifo  any  inf    0
0x048c1232  AAMR  fifo  any  inf    0    AAMR
0x048c1210  SPEC  fifo  any    1    0
0x048c11ee   CMR  fifo  any  inf    0
0x048c11cc  SPEC  fifo  any    1    0
0x048c11aa  SIGT  fifo  any  inf    0    SGMR
0x048c1188  ZMKM  fifo  any  inf    0    ZMKR
0x048c1166  MKMR  fifo  any  inf    0    MKR
0x048c1144  DEF3  fifo  any  inf    0    DEF3
0x048c1122  SNFI  fifo  any  inf    0    SNFI
0x048c1100  SNMR  fifo  any  inf    0    SNMR
0x048c10de  SPEC  fifo  any    1    0
0x048c10bc  DEF2  fifo  any  inf    0    DEF2
0x048c109a  LGDT  fifo  any    1    0    LGDT
0x048c1078  LGDE  fifo  any  inf    0    LGDE
0x048c1056  SPEC  fifo  any    1    0
0x048c1034  DSFI  fifo  any  inf    0    DSFI
0x048c1012  LGDT  fifo  any    1    0    LGDT
0x048c0ff0  LGDS  fifo  any  inf    0    LGDS
0x048c0fce  SPEC  fifo  any    1    0
0x048c0fac  SWFI  fifo  any  inf    0    SWFI
0x048c0f8a  LGST  fifo  any    1    0    LGST
0x048c0f68  LGSW  fifo  any  inf    0    LGSW
0x048c0f46  SPEC  fifo  any    1    0
0x048c0f24  DEFI  fifo  any  inf    0    DEFI
0x048c0f02  DEMT  fifo  any    1    0    DEMT
0x048c0ee0  DEMR  fifo  any  inf    0    DEMR
0x048c0ebe  SPEC  fifo  any    1    0
0x048c0e9c  DMZF  fifo  any  inf    0    DMZF
0x048c0e7a  ZDMT  fifo  any    1    0    ZDMT
0x048c0e58  ZDMR  fifo  any  inf    0    ZDMR
0x048c0e36  SPEC  fifo  any    1    0
0x048c0e14  SIFI  fifo  any  inf    0    SIFI
0x048c0df2  SIMT  fifo  any    1    0    SIMT
0x048c0dd0  SIMR  fifo  any  inf    0    SIMR
0x048c0dae  SPEC  fifo  any    1    0
0x048c0d8c  DZFI  fifo  any  inf    0    DZFI
0x048c0d6a  DZMT  fifo  any    1    0    DZMT
0x048c0d48  DZMR  fifo  any  inf    0    DZMR
0x048c0d26  SPEC  fifo  any    1    0
0x048c0d04  DSFI  fifo  any  inf    0    DSFI
0x048c0ce2  DSMT  fifo  any    1    0    DSMT
0x048c0cc0  DSMR  fifo  any  inf    0    DSMR
0x048c0c9e  SPEC  fifo  any    1    0
0x048c0c7c  SWFI  fifo  any  inf    0
0x048c0c5a  SWMT  fifo  any    1    0    SWMT
0x048c0c38  SWMR  fifo  any  inf    0    SWMR
0x048c0c16  SPEC  fifo  any    1    0
0x048c0bf4  shrL  fifo  any    1    0
0x048c0bd2  ACTV  fifo  any    1    0
0x048c0bb0  hihr  fifo  any    1    0
0x048c0b8e  hihr  fifo  any    1    0
0x048c0b6c  hihr  fifo  any    1    0
0x048c0b4a  hihr  fifo  any    1    0
0x048c0b28  MENU  fifo  any    1    0
0x048c0b06  MENU  fifo  any    1    0
0x048c0ae4  MENU  fifo  any    1    0
0x048c0ac2  MENU  fifo  any    1    0
0x048c0aa0  MENU  fifo  any    1    0
0x048c0a7e  MENU  fifo  any    1    0
0x048c0a5c  MENU  fifo  any    1    0
0x048c0a3a  MENU  fifo  any    1    0
0x048c0a18  MENU  fifo  any    1    0
0x048c09f6  ACTV  fifo  any    1    0
0x048c09d4  SDRL  fifo  any    1    0
0x048c09b2  SDIL  fifo  any    1    0
0x048c0990    R2  fifo  any    1    0
0x048c096e    R1  fifo  any    1    0
0x048c094c    R0  fifo  any    1    0
0x048c092a  isLk  fifo  any    1    0
0x048c0908  dtLk  fifo  any    1    0
0x048c08e6  mSTM  fifo  any  inf    0    MIME
0x048c08c4  mMIN  fifo  any  inf    0
0x048c08a2  mMCL  fifo  any  inf    0
0x048c0880  mMCR  fifo  any  inf    0
0x048c085e  mMDA  fifo  any  inf    1
0x048c083c  mMSA  fifo  any  inf    0
0x048c081a  mDVL  fifo  any  inf    0
0x048c07f8  dest  fifo  any  inf    0
0x048c07d6  mLDS  fifo  any  inf    0
0x048c07b4  FNSL  fifo  any    1    0
0x048c0792  DDET  fifo  any  inf    0
0x048c0770  DTRG  fifo  any  inf    0
0x048c074e  DSWP  fifo  any  inf    0
0x048c072c  Didi  fifo  any  inf    0
0x048c070a  cntw  fifo  any    1    0
0x048c06e8  cntx  fifo  any    1    0
0x048c06c6  dRes  fifo  any    1    0
0x048c06a4  Dlp   fifo  any    1    0
0x048c0682  CalC  fifo  any    1    0
0x048c0660  Scpi  fifo  any    1    0
0x048c063e   LG1  fifo  any    1    0
0x048c061c  ANON  fifo  any    1    0
0x048c05fa  GPIB  fifo  any    1    0
0x048c05d8  PCkb  fifo  any    1    0    PCKB
0x048c05b6  DISP  fifo  any    1    0    DISP
0x048c0594  OMMG  fifo  any    1    0
0x048c0572   UNS  fifo  any    1    0
0x048c0550  RLDS  fifo  any    1    0    ROOT
0x048c052e    BW  fifo  any    1    0
0x048c050c  GLds  fifo  any    1    0
0x048c04ea  BLds  fifo  any  inf    1
0x048c04c8  CISW  fifo  any    1    0
0x048c04a6  MRLK  fifo  any    1    0
0x048c0484  HWLK  fifo  any    1    0
0x048c0462    FP  fifo  any    1    0    FPLP
0x048c0440  ADCF  fifo  any    1    0
0x048c041e  DIRg  fifo  any    1    0
0x048c03fc  ANON  fifo  any    1    0
0x048c03da  DIRf  fifo  any    1    0
0x048c03b8  ANON  fifo  any    1    0
0x048c0396  SIOB  fifo  any    1    0
0x048c0374  CALM  fifo  any    1    0
0x048c0352  DIRe  fifo  any    1    0
0x048c0330  ANON  fifo  any    1    0
0x048c030e  DIRd  fifo  any    1    0
0x048c02ec  DIRc  fifo  any    1    0
0x048c02ca  DIRb  fifo  any    1    0
0x048c02a8  DIRa  fifo  any    1    0
0x048c0286  APPS  fifo  any  inf    0    APPS
0x048c0264  DLLK  fifo  any    1    0
0x048c0242  SUBL  fifo  any    1    0
0x048c0220  LCSH  fifo  any    1    0
0x048c01fe  FBUF  fifo  any    1    0
0x048c01dc  DCAS  fifo  any    1    0    DCAS
0x048c01ba  RLCN  fifo  any    1    0    RLCN
0x048c0198  MROF  fifo  any    1    0
0x048c0176  MRON  fifo  any    1    0
0x048c0154  SWSP  fifo  any    1    0
0x048c0132   SRQ  fifo  any    1    0    SYMR
0x048c0110  PRTH  fifo  any    1    0    PRNT
        155 Exchange(s) (245 avail).
        2 Msg buffer(s) (1022 avail).
>g
Breakpoint handler installed
>i
PsosSystemData (0x048bcce8):
        (0x048bcce8) OS_PCB   *runningPCB    = 0x048ca38c
        (0x048bccec) OS_PCB   *readyList     = 0x048c97f8
        (0x048bccf0) OS_PCB   *pauseList     = 0x048ccf30
        (0x048bccf4) OS_PCB   *pcbActiveHead = 0x048cd014
        (0x048bccf8) OS_PCB   *pcbFreeHead   = 0x048ccd68
        (0x048bccfc) OS_XCB   *xcbActiveHead = 0x048c1584
        (0x048bcd00) OS_XCB   *xcbFreeHead   = 0x048c15a6
        (0x048bcd04) OS_Message *mgbFreeHead = 0x048c3750
        (0x048bcd08) void  *sstackEnd        = 0x048c0110
        (0x048bcd0c) short kernelLevel       = 0
        (0x048bcd0e) short reserved1         = 0
        (0x048bcd10) int   reserved2         = 1280
        (0x048bcd14) int   phileData         = 620765184
        (0x048bcd18) int   probeEntry        = 71205862
        (0x048bcd1c) OS_PCB   *memQHead      = 0x048bcd1c
        (0x048bcd20) OS_PCB   *memQTail      = 0x048bcd1c
        (0x048bcd24) int   timeoutTicks      = 41
        (0x048bcd28) short ticks             = 45
        (0x048bcd2a) short pad1              = 0
        (0x048bcd2c) int   time              = 292
        (0x048bcd30) int   date              = 130155777
        (0x048bcd34) char  motbl[12]         =
        (0x048bcd40) short ticksPerSec       = 100
        (0x048bcd42) short ticksPerSlice     = 1
        (0x048bcd44) char  todset            =
        (0x048bcd45) char  eventRace         = (0x048bcd46) char  unusedPad[2]      =
        (0x048bcd48) Lds_UInt32 switchProc   = 0
        (0x048bcd4c) regionInfo[0].minSeg      = 20
        (0x048bcd50) regionInfo[0].maxSeg      = 58796
        (0x048bcd54) regionInfo[0].minPend     = 6020
        (0x048bcd58) regionInfo[0].regionEnd   = 0x048dcce7
        (0x048bcd5c) regionInfo[0].regionName  = REG1
  (0x048bcd60) regionInfo[0].freeHead    = 0x048ce73c
        (0x048bcd64) regionInfo[0].freeTail    = 0x048d3e30
        (0x048bcd68) regionInfo[0].regionFlags = 0
        (0x048bcd6c) regionInfo[1].minSeg      = 20
        (0x048bcd70) regionInfo[1].maxSeg      = 24261400
        (0x048bcd74) regionInfo[1].minPend     = 24261401
        (0x048bcd78) regionInfo[1].regionEnd   = 0x05ffffff
        (0x048bcd78) regionInfo[1].regionEnd   = 0x05Bfffff
        (0x048bcd80) regionInfo[1].freeHead    = 0x048dcce8
        (0x048bcd84) regionInfo[1].freeTail    = 0x05e74208
        (0x048bcd88) regionInfo[1].regionFlags = 0
        (0x048bcd8c) regionInfo[2].minSeg      = 20
        (0x048bcd90) regionInfo[2].maxSeg      = 21844
        (0x048bcd94) regionInfo[2].minPend     = 21845
        (0x048bcd98) regionInfo[2].regionEnd   = 0x0a006393
        (0x048bcd9c) regionInfo[2].regionName  = dyna
        (0x048bcda0) regionInfo[2].freeHead    = 0x0a000e40
        (0x048bcda4) regionInfo[2].freeTail    = 0x0a000e40
        (0x048bcda8) regionInfo[2].regionFlags = 1
        (0x048bcdac) regionInfo[3].minSeg      = 128
        (0x048bcdb0) regionInfo[3].maxSeg      = 7120
        (0x048bcdb4) regionInfo[3].minPend     = 7121
        (0x048bcdb8) regionInfo[3].regionEnd   = 0x0a007fa3
        (0x048bcdbc) regionInfo[3].regionName  = nvra
        (0x048bcdc0) regionInfo[3].freeHead    = 0x0a0063d4
        (0x048bcdc4) regionInfo[3].freeTail    = 0x0a0063d4
        (0x048bcdc8) regionInfo[3].regionFlags = 0
        (0x048bcdcc) regionInfo[4].minSeg      = 0
        (0x048bcdd0) regionInfo[4].maxSeg      = 0
        (0x048bcdd4) regionInfo[4].minPend     = 0
        (0x048bcdd8) regionInfo[4].regionEnd   = 0x00000000
        (0x048bcddc) regionInfo[4].regionName  =
        (0x048bcde0) regionInfo[4].freeHead    = 0x00000000
        (0x048bcde4) regionInfo[4].freeTail    = 0x00000000
        (0x048bcde8) regionInfo[4].regionFlags = 0
        (0x048bcdec) regionInfo[5].minSeg      = 0
        (0x048bcdf0) regionInfo[5].maxSeg      = 0
        (0x048bcdf4) regionInfo[5].minPend     = 0
        (0x048bcdf8) regionInfo[5].regionEnd   = 0x00000000
        (0x048bcdfc) regionInfo[5].regionName  =
        (0x048bce00) regionInfo[5].freeHead    = 0x00000000
        (0x048bce04) regionInfo[5].freeTail    = 0x00000000
        (0x048bce08) regionInfo[5].regionFlags = 0
        (0x048bce0c) regionInfo[6].minSeg      = 0
        (0x048bce10) regionInfo[6].maxSeg      = 0
        (0x048bce14) regionInfo[6].minPend     = 0
        (0x048bce18) regionInfo[6].regionEnd   = 0x00000000
        (0x048bce1c) regionInfo[6].regionName  =
        (0x048bce20) regionInfo[6].freeHead    = 0x00000000
        (0x048bce24) regionInfo[6].freeTail    = 0x00000000
        (0x048bce28) regionInfo[6].regionFlags = 0
        (0x048bce2c) regionInfo[7].minSeg      = 0
        (0x048bce30) regionInfo[7].maxSeg      = 0
        (0x048bce34) regionInfo[7].minPend     = 0
        (0x048bce38) regionInfo[7].regionEnd   = 0x00000000
        (0x048bce3c) regionInfo[7].regionName  =
        (0x048bce40) regionInfo[7].freeHead    = 0x00000000
        (0x048bce44) regionInfo[7].freeTail    = 0x00000000
        (0x048bce48) regionInfo[7].regionFlags = 0
        (0x048bce4c) regionInfo[8].minSeg      = 0
        (0x048bce50) regionInfo[8].maxSeg      = 0
        (0x048bce54) regionInfo[8].minPend     = 0
        (0x048bce58) regionInfo[8].regionEnd   = 0x00000000
        (0x048bce5c) regionInfo[8].regionName  =
        (0x048bce60) regionInfo[8].freeHead    = 0x00000000
        (0x048bce64) regionInfo[8].freeTail    = 0x00000000
        (0x048bce68) regionInfo[8].regionFlags = 0
        (0x048bce6c) regionInfo[9].minSeg      = 0
        (0x048bce70) regionInfo[9].maxSeg      = 0
        (0x048bce74) regionInfo[9].minPend     = 0
        (0x048bce78) regionInfo[9].regionEnd   = 0x00000000
        (0x048bce7c) regionInfo[9].regionName  =
        (0x048bce80) regionInfo[9].freeHead    = 0x00000000
        (0x048bce84) regionInfo[9].freeTail    = 0x00000000
        (0x048bce88) regionInfo[9].regionFlags = 0
        (0x048bce8c) regionInfo[10].minSeg      = 0
        (0x048bce90) regionInfo[10].maxSeg      = 0
        (0x048bce94) regionInfo[10].minPend     = 0
        (0x048bce98) regionInfo[10].regionEnd   = 0x00000000
        (0x048bce9c) regionInfo[10].regionName  =
        (0x048bcea0) regionInfo[10].freeHead    = 0x00000000
        (0x048bcea4) regionInfo[10].freeTail    = 0x00000000
        (0x048bcea8) regionInfo[10].regionFlags = 0
        (0x048bceac) regionInfo[11].minSeg      = 0
        (0x048bceb0) regionInfo[11].maxSeg      = 0
        (0x048bceb4) regionInfo[11].minPend     = 0
        (0x048bceb8) regionInfo[11].regionEnd   = 0x00000000
        (0x048bcebc) regionInfo[11].regionName  =
        (0x048bcec0) regionInfo[11].freeHead    = 0x00000000
        (0x048bcec4) regionInfo[11].freeTail    = 0x00000000
        (0x048bcec8) regionInfo[11].regionFlags = 0
        (0x048bcecc) regionInfo[12].minSeg      = 0
        (0x048bced0) regionInfo[12].maxSeg      = 0
        (0x048bced4) regionInfo[12].minPend     = 0
        (0x048bced8) regionInfo[12].regionEnd   = 0x00000000
        (0x048bcedc) regionInfo[12].regionName  =
        (0x048bcee0) regionInfo[12].freeHead    = 0x00000000
        (0x048bcee4) regionInfo[12].freeTail    = 0x00000000
        (0x048bcee8) regionInfo[12].regionFlags = 0
        (0x048bceec) regionInfo[13].minSeg      = 0
        (0x048bcef0) regionInfo[13].maxSeg      = 0
        (0x048bcef4) regionInfo[13].minPend     = 0
        (0x048bcef8) regionInfo[13].regionEnd   = 0x00000000
        (0x048bcefc) regionInfo[13].regionName  =
        (0x048bcf00) regionInfo[13].freeHead    = 0x00000000
        (0x048bcf04) regionInfo[13].freeTail    = 0x00000000
        (0x048bcf08) regionInfo[13].regionFlags = 0
        (0x048bcf0c) regionInfo[14].minSeg      = 0
        (0x048bcf10) regionInfo[14].maxSeg      = 0
        (0x048bcf14) regionInfo[14].minPend     = 0
        (0x048bcf18) regionInfo[14].regionEnd   = 0x00000000
        (0x048bcf1c) regionInfo[14].regionName  =
        (0x048bcf20) regionInfo[14].freeHead    = 0x00000000
        (0x048bcf24) regionInfo[14].freeTail    = 0x00000000
        (0x048bcf28) regionInfo[14].regionFlags = 0
        (0x048bcf2c) regionInfo[15].minSeg      = 0
        (0x048bcf30) regionInfo[15].maxSeg      = 0
        (0x048bcf34) regionInfo[15].minPend     = 0
        (0x048bcf38) regionInfo[15].regionEnd   = 0x00000000
        (0x048bcf3c) regionInfo[15].regionName  =
        (0x048bcf40) regionInfo[15].freeHead    = 0x00000000
        (0x048bcf44) regionInfo[15].freeTail    = 0x00000000
        (0x048bcf48) regionInfo[15].regionFlags = 0
        (0x048bcf4c) regionInfo[16].minSeg      = 0
        (0x048bcf50) regionInfo[16].maxSeg      = 0
        (0x048bcf54) regionInfo[16].minPend     = 0
        (0x048bcf58) regionInfo[16].regionEnd   = 0x00000000
        (0x048bcf5c) regionInfo[16].regionName  =
        (0x048bcf60) regionInfo[16].freeHead    = 0x00000000
        (0x048bcf64) regionInfo[16].freeTail    = 0x00000000
        (0x048bcf68) regionInfo[16].regionFlags = 0
        (0x048bcf6c) regionInfo[17].minSeg      = 0
        (0x048bcf70) regionInfo[17].maxSeg      = 0
        (0x048bcf74) regionInfo[17].minPend     = 0
        (0x048bcf78) regionInfo[17].regionEnd   = 0x00000000
        (0x048bcf7c) regionInfo[17].regionName  =
        (0x048bcf80) regionInfo[17].freeHead    = 0x00000000
        (0x048bcf84) regionInfo[17].freeTail    = 0x00000000
        (0x048bcf88) regionInfo[17].regionFlags = 0
        (0x048bcf8c) regionInfo[18].minSeg      = 0
        (0x048bcf90) regionInfo[18].maxSeg      = 0
        (0x048bcf94) regionInfo[18].minPend     = 0
        (0x048bcf98) regionInfo[18].regionEnd   = 0x00000000
        (0x048bcf9c) regionInfo[18].regionName  =
        (0x048bcfa0) regionInfo[18].freeHead    = 0x00000000
        (0x048bcfa4) regionInfo[18].freeTail    = 0x00000000
        (0x048bcfa8) regionInfo[18].regionFlags = 0
        (0x048bcfac) regionInfo[19].minSeg      = 0
        (0x048bcfb0) regionInfo[19].maxSeg      = 0
        (0x048bcfb4) regionInfo[19].minPend     = 0
        (0x048bcfb8) regionInfo[19].regionEnd   = 0x00000000
        (0x048bcfbc) regionInfo[19].regionName  =
        (0x048bcfc0) regionInfo[19].freeHead    = 0x00000000
        (0x048bcfc4) regionInfo[19].freeTail    = 0x00000000
        (0x048bcfc8) regionInfo[19].regionFlags = 0
        (0x048bcfcc) regionSaveInfo[0] = 0x00000000
        (0x048bcfd0) regionSaveInfo[1] = 0x00000000
        (0x048bcfd4) regionSaveInfo[2] = 0x0a000004
        (0x048bcfd8) regionSaveInfo[3] = 0x00000000
        (0x048bcfdc) regionSaveInfo[4] = 0x00000000
        (0x048bcfe0) regionSaveInfo[5] = 0x00000000
        (0x048bcfe4) regionSaveInfo[6] = 0x00000000
        (0x048bcfe8) regionSaveInfo[7] = 0x00000000
        (0x048bcfec) regionSaveInfo[8] = 0x00000000
        (0x048bcff0) regionSaveInfo[9] = 0x00000000
        (0x048bcff4) regionSaveInfo[10] = 0x00000000
        (0x048bcff8) regionSaveInfo[11] = 0x00000000
        (0x048bcffc) regionSaveInfo[12] = 0x00000000
        (0x048bd000) regionSaveInfo[13] = 0x00000000
        (0x048bd004) regionSaveInfo[14] = 0x00000000
        (0x048bd008) regionSaveInfo[15] = 0x00000000
        (0x048bd00c) regionSaveInfo[16] = 0x00000000
        (0x048bd010) regionSaveInfo[17] = 0x00000000
        (0x048bd014) regionSaveInfo[18] = 0x00000000
        (0x048bd018) regionSaveInfo[19] = 0x00000000
9>
Contents of the Exception Report:
[0x0a007fa4] D0 = 0x00000000
[0x0a007fa8] D1 = 0x00000000
[0x0a007fac] D2 = 0x00000000
[0x0a007fb0] D3 = 0x00000000
[0x0a007fb4] D4 = 0x00000000
[0x0a007fb8] D5 = 0x00000000
[0x0a007fbc] D6 = 0x00000000
[0x0a007fc0] D7 = 0x00000000
[0x0a007fc4] A0 = 0x00000000
[0x0a007fc8] A1 = 0x00000000
[0x0a007fcc] A2 = 0x00000000
[0x0a007fd0] A3 = 0x00008000
[0x0a007fd4] A4 = 0x00000000
[0x0a007fd8] A5 = 0x00000000
[0x0a007fdc] A6 = 0x00000000
[0x0a007fe0] A7 = 0x00000000
[0x0a007fe4] SSP = 0x00000000
[0x0a007fe8] SR = 0x0000
[0x0a007fec] PC = 0x00000000
[0x0a007fec] FMT/VO = 0x0000
 '[tT]' - Time log.
Sandra
The following users thanked this post: tv84
