Yep, no doubt - and why I mentioned that in my post. Still a useful, low cost, low voltage differential probe.
The P6046 is very or even irreplaceably useful if you actually have an application which needs something like it. This probe was Tektronix's longest lived product.
There is a big difference in performance between low voltage and high voltage differential probes. The Sapphire and similar high voltage differential probes do not include a complete CMRR versus frequency or noise (or recovery time) specifications for good reason; they cannot compete with the P4046 in this respect or even the old Tektronix 7A13 and Preamble 1822 (bought by LeCroy) designs when they are used with the proper external probes. Except for input voltage range, even an properly adjusted oscilloscope in add and invert mode can be competitive. (1)
On the other hand, they are perfectly suited for line side applications like off-line switching power supplies where a higher voltage range is required and are simpler to use and less expensive than a 7A13 or Preamble 1822. And if capt bullshot's story about calibration is common, then most users do not understand or know how to accurately use them anyway.
As far as their seemingly high price, I suspect it is just a situation of what the limited market will bear and the design and validation is tricky for such a low volume product.
(1) It is a little eerie using an analog oscilloscope in add and invert mode while adjusting it for maximum CMRR. You can *see* the displayed noise decrease when it is adjusted correctly. The same goes for adjusting the external probes on a 7A13 or Preample 1822. The movie opens on a fog shrouded hill; the fog clears as the audience yells, "Focus!"