Author Topic: EEVblog 121GW Multimeter Issues  (Read 710795 times)
Online EEVblog
Re: EEVblog 121GW Multimeter Issues
« Reply #1400 on: September 11, 2023, 11:51:59 pm »
To which email address?
If i click email in the forum, i simply don't get an email address.
I also mailed to your eevblog.com contact address but no reaction yet.
I got it, just very busy right now.
Offline hainjedaf
« Reply #1401 on: September 11, 2023, 11:59:02 pm »
Ok. I'll wait. no hurry.
At least I know contact established.
Kind Regards,
Marout Yasuo Sluijter-Borms
Amsterdam NL
Offline DavePrime
« Reply #1402 on: October 26, 2023, 01:55:58 am »
AC volts function stopped working. When testing mains I initially got a good reading, then over the course of a few seconds voltage readings dropped to 0 and frequency reading would briefly flash 60hz then back to 0. A few hours later I testedmains again to document on video and voltage only read 75-80 volts, briefly made it to 100 and then dropped to 0 volts over a few seconds, same flashing behavior on frequency after this occurred. Now reading ~ 0 volts plugged into mains, the frequency reading will alternate between 0 and very briefly flashing 60. Any ideas?
Video link
https://youtu.be/Xmz-QeD6Tpw?si=PQZffWXd8nblFL4A
https://youtube.com/shorts/xEucjcGsZ94?si=BhtVPb4WIAV2M0B4
Offline bicycleguy
« Reply #1403 on: October 26, 2023, 02:05:42 am »
Don't recognize that + - symbol below Auto
Offline J-R
« Reply #1404 on: October 26, 2023, 03:51:06 am »
I'm sure he's kicking himself as we speak...
Offline DavePrime
« Reply #1405 on: October 26, 2023, 04:20:38 am »
Yes I did kick myself, and that did fix the issue. Changed the batteries a few weeks back and somehow one bad cell got mixed in and never noticed the warning was still on the screen.
Offline hainjedaf
« Reply #1406 on: October 26, 2023, 08:49:25 am »
I can understand the 'being busy' part.
Buut... We're a month in so... is there any progress?
Maybe something turned up in the two bins of multimeters just researched?
I'd be very happy with a broken 121GW PCB to unsolder good working input jacks from...
Kind Regards,
Amsterdam NL
Online EEVblog
« Reply #1407 on: October 26, 2023, 10:27:44 pm »
Sorry. It's in the shipping system now so I can't forget.
Offline KungFuJosh
  • TEAS is real.
« Reply #1408 on: October 27, 2023, 12:17:22 am »
Is this model a Back to the Future reference?
The following users thanked this post: mwb1100
Offline hainjedaf
« Reply #1409 on: October 27, 2023, 11:55:20 am »
Thanks a milion. :) Hope my 121 GW will be a happy camper soon :)
Love that DMM, especially with ProbeMaster leads
Kind Regards,
Amsterdam NL
Online EEVblog
« Reply #1410 on: October 28, 2023, 08:38:44 am »
Of course.
Offline hainjedaf
« Reply #1411 on: November 14, 2023, 02:03:02 pm »
Hi Dave,
Austraila Post knocked on my door this morning.
Need to work tonight so tomorrow I'll rejuvenate the 121GW.
Thanks a Million. I'll post the result here :)
Kind Regards,
Amsterdam NL
Offline aesteve
« Reply #1412 on: February 22, 2024, 11:42:34 am »
Hi Dave. I recently got a broken jack in my 121GW |O. Found this thread through a search engine. Do you still send replacements? I've sent you an email.
Online EEVblog
« Reply #1413 on: February 25, 2024, 11:09:21 pm »
Yep, send me an email.
Offline aesteve
« Reply #1414 on: March 04, 2024, 08:26:48 pm »
I already did. Maybe you missed it because I used a different mail account from what I used to register in this forum, or maybe I didn't send it to the right account. I'll send you it again from the other account. dave@eevblog.com eevblog.official@gmail.com <- is any of these correct?
Online thm_w
  • Non-expert
« Reply #1415 on: March 04, 2024, 11:28:27 pm »
I already did. Maybe you missed it because I used a different mail account from what I used to register in this forum, or maybe I didn't send it to the right account. I'll send you it again from the other account. dave at eevblog.com eevblog.official at gmail.com <- is any of these correct?
The first one is correct. His instructions previously in this thread were to "put 121GW connector in the title".
Offline crispus
« Reply #1416 on: March 10, 2024, 02:17:15 pm »
Hi everyone,
I cleaned it, still doesn't work but the shim is not the issue as one could see the above pictures  ;D but that lead me to the issue.
At close inspection it doesn't seem that any other component misses the magic smoke, ...
(Attachment Link)
After some fail attempts to source a new board (contacted the distributor (Welectron), they sent me to Brymen, Bryman sent me to eevblog and then was silence) I would like to give it one more try before send it to a better place.
But before going that rabbit hole, I have some questions:
  • beside the schematic is there any other documentation to speed up the process?
  • does any know if there are internal traces around the affected area?
  • some higher resolution photos in the affected area (for a good one :D)?
  • pcb parts placement would be great
Any tip will help.
Thank you.
« Last Edit: March 10, 2024, 02:22:55 pm by crispus »
I know I'm numskull, but I look around me and I feel better.
Offline Kean
    • Kean Electronics
« Reply #1417 on: March 10, 2024, 05:17:04 pm »
Hi everyone,
(Attachment Link)
Any tip will help.
Thank you.
That is significant damage, and I did something similar to my 121GW as shown back in post 1314 and in another thread at https://www.eevblog.com/forum/testgear/eevblog-121gw-discussion-thread/msg1599418/#msg1599418 (and earlier).
In my case I had mistaken measured a high voltage (1200V 25kHz from memory) that burnt my PCB, but after I cleaned the PCB to remove the burnt section it worked fine again.  For you to have a similar burnt PCB (although on a different section of the switch) indicates that your 121GW very likely was also exposed to a high voltage to cause the arc over and burnt FR4.
So, the first thing you need to is dig away at that burnt PCB surface a bit, and then wash it with some IPA.  The carbonised section is going to be conductive and is likely causing leakage currents between the pads of the rotary switch.
If that doesn't help (or even if it does), then you might want to check the state of the protection components close to the HY3131 chip, e.g. D13, Q1/Q2, and Q3/Q5.  If those seem OK (not leaky or shorted) then it is possible that your HY3131 chip is damaged.  The parts may not show visible damage, but still be causing your false readings.
« Last Edit: March 10, 2024, 05:18:44 pm by Kean »
The following users thanked this post: EEVblog
Online EEVblog
« Reply #1418 on: March 11, 2024, 12:37:23 am »
Any tip will help.
Was this back in 2021 as per original post?
There is no other documentation I'm afraid.
As Kean said, it seems that would have to be dug out and potentially re-epoxied.
Offline Kean
    • Kean Electronics
« Reply #1419 on: March 11, 2024, 02:01:11 am »
Any tip will help.
In that other thread I linked, there are some good tips to read starting at my original post and over the following week or so.
https://www.eevblog.com/forum/testgear/eevblog-121gw-discussion-thread/msg1580530/#msg1580530
There was some helpful info reverse engineering the range switch contacts, but also Joe did some great video content showing what would be needed to cause such damage which then triggered my memory of the potential incident that lead to the damage.
Like you, after the PCB damage I was getting unexpected high readings (166V or OFL when measuring 12V DC).  This is a high impedance circuit, and unexpected leakage currents could cause the weird readings.  Cleaning up the burnt PCB sorted it all out.  Amazing nothing else was needed, and the protection components seemed to have done their job.
I hope this helps you sort yours out too.
Offline tszaboo
« Reply #1420 on: June 17, 2024, 08:25:17 pm »
Hello EEVBlog!
I feel honored to now be a part of this awesome community. What makes me very sad is that my first post has to be a negative one.
I purchased my 121GW meter on July 26th 2018, during a small batch at official EEVBlog store, after being informed about its availability by the newsletter. Soon I discovered how big mistake I have made.
After receiving the meter and unboxing it, few things immediately stood out. The meter was scratched in many places (especially the knob and in the area around lead sockets) - see attachment. Scratches on two flat sides of the knob clearly came from someone's else fingernails. Part of the knob you grab by your fingers was noticably shinier than other areas. Batteries have also already been put inside the meter - From what I've seen on unboxing videos, these should be included seperately in the pouch.  Conclusion is simple - I recieved a used meter. I thought - "Ok, fine. This is not that bad. I can use this meter nevertheless." And I could - for a month. After that strange things started to happen.
Before making the purchase I did some research online (mainly this forum) and discovered that at some point a common problem with 121GW meters was the knob itself. Apparently there was a recall done at some point regarding the contact issues (knob replacement and addition of a shim). This took place months earlier. Common sense suggested that my meter should have already been free of this issue. This couldn't be farther from the truth.
After a month of very, very light use the meter is basically unusable. The knob wipers stopped making a good contact with the PCB (i think) and the meter reboots constantly while switching modes. Resistance mode cannot be turned on at all. The meter just turns off when I select it. The issue is present for every single knob position.
I made a short video where the issue is clearly visible:
As you can see, this is clearly not a 121GW revision with a fixed knob. I must have received an older, flawed version. This is absolutely unforgivable for a meter this expensive (and every other for that matter). What I find the most frustrating is that the recall was done some time before I bought my meter from the EEVBlog store. At that point I should expect this issue to be already resolved but in the end I recieved a meter with this exact defect still present.
I've sent Dave an email describing these issues on 5th October. He replied a week later and promised to send me a new meter. I was really happy about that. So why am I even writing this post, you may ask? Well, because the new meter has never been shipped and Dave has not replied to any of my emails since then :palm:. I've been trying many times to contact him through at least 3 EEVBlog email adresses and sending him PMs on the forum, asking why the new meter hasn't been shipped. Three weeks have passed and I have not recieved any reply. I have no idea why.
I can only hope that this post will grab Dave's attention and that my meter will get replaced eventually. Do you know if there is any other way to contact Dave or to fix the meter? I just want to be able to use the product i paid for.
Thank you for your help in advance,
A VERY frustrated customer.
So the meter I have developed an issue very similar to this video. Only in the resistance range, moving the knob will reset it, it's fiddly. Seems to be a physical issue. Is this something which is fixable by user?
Offline J-R
« Reply #1421 on: June 17, 2024, 10:04:50 pm »
When/where did you purchase the DMM?  Some of the early units needed a shim added to the selector switch.
It's also possible there is some damage or contamination.
You could disassemble it to visually inspect the internals.  Be careful not to break the small plastic clip that holds the two halves together near the bottom corner.  (Although honestly if you break it it's not a massive issue.)
Online EEVblog
« Reply #1422 on: June 18, 2024, 12:49:44 am »
As J-R mentioned, could be an older unit, what's the serial number?
I can send a replacement knob with shim if needed.
Offline tszaboo
« Reply #1423 on: June 24, 2024, 08:31:53 am »
It was purchased a few months ago.
SN: 230102474
When I switch it to ohms range, it is off, until I wiggle the knob a bit.
Let me know what I (or you) can do, I don't want to RMA it, because it's a PITA to do at work, plus I don't want to cause unnecessary expense to you.
Offline jeslaughter
« Reply #1424 on: June 26, 2024, 01:21:19 am »
Need email address to order replacement jacks for my GW121.  Thanks in advance.
