Upd: I put thumbnails inline, and the full size images are attached. These are probably too wide to be inlined. So please scroll to the end for full-size images.I also was not able to reproduce this in the 1V range or with fast changing signal.
But it easily happens in the "battery voltage monitoring", which is basically 1-1/X .
I set up my signal generator like this:
![](https://www.eevblog.com/forum/index.php?action=dlattach;topic=101807.0;attach=886102;image)
Set logging with 0 interval and left it on for 16 hours:
![](https://www.eevblog.com/forum/index.php?action=dlattach;topic=101807.0;attach=886106;image)
I did check that the LCD still showed full cycle before I turned it off to extract SD card.
But there is something more. File is corrupted in several places.
Let's check the string length in the file:
$ awk '{print length}' 19120701.CSV | uniq -c
1 27
1 14
1 18
1 32
1 60
9 24
90 25
900 26
9000 27
11476 28
1 173
78518 28
18848 29
1 419
17752 29
1 239
2927 29
1 389
44207 29
1 389
20757 29
1 89
87220 29
1 25
1 21
So there is header (first 5 lines of length 27, 14, 18, 32 and 60 characters).
Then there are 9, 90, 900, 9000 lines with increasing index.
But then after
9+90+900+9000+11476 = 21475 lines appears line of 173 characters. What are these?
All the bad lines were longer than 29 characters.
$ awk 'length > 30 { print }' 19120701.CSV
,MAIN,,,SUB-1,,,SUB-2,,,Remark,
No. ,Func. ,Value,Unit,Func. ,Value,Unit,Func. ,Value,Unit,
21476,DCV,009.121481,DCV,009.185,V,,,,,,,,
118118861,DCV,009.120,V,,,,,,,,
136614,DCV,009.136621,DCV,009.120,V,,,,,,,,
139561,DCV,009.120,V,,,,,,,,
183769,DCV,009.160,V,183781,DCV,009.120,V,,,,,,,,
204539,DCV,009.166,V,,,,,204541,DCV,009.120,V,,,,,,,,
That's interesting, but where are the 173 and 419 characters? Here they are:
$ awk 'length > 30 { print }' 19120701.CSV | xxd
00000050: 2c56 616c 7565 2c55 6e69 742c 0d0a 3231 ,Value,Unit,..21
00000060: 3437 362c 4443 562c 3030 392e 3100 0000 476,DCV,009.1...
00000070: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000080: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000090: 0000 0000 0000 0000 0000 0000 0000 0000 ................
000000a0: 0000 0000 0000 0000 0000 0000 0000 0000 ................
000000b0: 0000 0000 0000 0000 0000 0000 0000 0000 ................
000000c0: 0000 0000 0000 0000 0000 0000 0000 0000 ................
000000d0: 0000 0000 0000 0000 0000 0000 0000 0000 ................
000000e0: 0000 0000 0000 0000 0000 0000 0000 0032 ...............2
000000f0: 3134 3831 2c44 4356 2c30 3039 2e31 3835 1481,DCV,009.185
00000100: 2c56 2c2c 2c2c 2c2c 2c2c 0d0a 3131 3800 ,V,,,,,,,,..118.
00000110: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000120: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000130: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000140: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000150: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000160: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000170: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000180: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000190: 0000 0000 0000 0000 0000 0000 0000 0000 ................
000001a0: 0000 0000 0000 0000 0000 0000 0000 0000 ................
000001b0: 0000 0000 0000 0000 0000 0000 0000 0000 ................
000001c0: 0000 0000 0000 0000 0000 0000 0000 0000 ................
000001d0: 0000 0000 0000 0000 0000 0000 0000 0000 ................
000001e0: 0000 0000 0000 0000 0000 0000 0000 0000 ................
000001f0: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000200: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000210: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000220: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000230: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000240: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000250: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000260: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000270: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000280: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000290: 0000 3131 3838 3631 2c44 4356 2c30 3039 ..118861,DCV,009
000002a0: 2e31 3230 2c56 2c2c 2c2c 2c2c 2c2c 0d0a .120,V,,,,,,,,..
000002b0: 3133 3636 3134 2c44 4356 2c30 3039 2e00 136614,DCV,009..
000002c0: 0000 0000 0000 0000 0000 0000 0000 0000 ................
000002d0: 0000 0000 0000 0000 0000 0000 0000 0000 ................
000002e0: 0000 0000 0000 0000 0000 0000 0000 0000 ................
000002f0: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000300: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000310: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000320: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000330: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000340: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000350: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000360: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000370: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000380: 0000 3133 3636 3231 2c44 4356 2c30 3039 ..136621,DCV,009
00000390: 2e31 3230 2c56 2c2c 2c2c 2c2c 2c2c 0d0a .120,V,,,,,,,,..
000003a0: 3133 3935 3439 2c44 4356 2c30 3039 2e31 139549,DCV,009.1
000003b0: 3630 2c56 2c2c 2c2c 2c2c 2c2c 0d00 0000 60,V,,,,,,,,....
000003c0: 0000 0000 0000 0000 0000 0000 0000 0000 ................
000003d0: 0000 0000 0000 0000 0000 0000 0000 0000 ................
000003e0: 0000 0000 0000 0000 0000 0000 0000 0000 ................
000003f0: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000400: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000410: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000420: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000430: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000440: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000450: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000460: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000470: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000480: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000490: 0000 0000 0000 0000 0000 0000 0000 0000 ................
000004a0: 0000 0000 0000 0000 0000 0000 0000 0000 ................
000004b0: 0000 0000 0000 0000 0000 0000 0000 0000 ................
000004c0: 0000 0000 0000 0000 0000 0000 0000 0000 ................
000004d0: 0000 0000 0000 0000 0000 0000 0000 0000 ................
000004e0: 0000 0000 0000 0000 0000 0000 0000 0000 ................
000004f0: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000500: 0000 0000 0000 0000 3133 3935 3631 2c44 ........139561,D
00000510: 4356 2c30 3039 2e31 3230 2c56 2c2c 2c2c CV,009.120,V,,,,
00000520: 2c2c 2c2c 0d0a 3138 3337 3639 2c44 4356 ,,,,..183769,DCV
00000530: 2c30 3039 2e31 3630 2c56 2c00 0000 0000 ,009.160,V,.....
00000540: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000550: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000560: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000570: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000580: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000590: 0000 0000 0000 0000 0000 0000 0000 0000 ................
000005a0: 0000 0000 0000 0000 0000 0000 0000 0000 ................
000005b0: 0000 0000 0000 0000 0000 0000 0000 0000 ................
000005c0: 0000 0000 0000 0000 0000 0000 0000 0000 ................
000005d0: 0000 0000 0000 0000 0000 0000 0000 0000 ................
000005e0: 0000 0000 0000 0000 0000 0000 0000 0000 ................
000005f0: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000600: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000610: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000620: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000630: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000640: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000650: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000660: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000670: 0000 0000 0000 0000 0000 0000 0000 0000 ................
00000680: 0000 0000 0000 0000 0000 0000 0000 3138 ..............18
00000690: 3337 3831 2c44 4356 2c30 3039 2e31 3230 3781,DCV,009.120
000006a0: 2c56 2c2c 2c2c 2c2c 2c2c 0d0a 3230 3435 ,V,,,,,,,,..2045
000006b0: 3339 2c44 4356 2c30 3039 2e31 3636 2c56 39,DCV,009.166,V
000006c0: 2c2c 2c2c 2c00 0000 0000 0000 0000 0000 ,,,,,...........
000006d0: 0000 0000 0000 0000 0000 0000 0000 0000 ................
000006e0: 0000 0000 0000 0000 3230 3435 3431 2c44 ........204541,D
000006f0: 4356 2c30 3039 2e31 3230 2c56 2c2c 2c2c CV,009.120,V,,,,
00000700: 2c2c 2c2c 0d0a ,,,,..
So basically there are random blocks of zeroes and some lost data.
Do these correlate with the output problems?
I replaced corrupted lines with impossible value of -1V:
![](https://www.eevblog.com/forum/index.php?action=dlattach;topic=101807.0;attach=886110;image)
Here is magnified graph just before it stopped:
![](https://www.eevblog.com/forum/index.php?action=dlattach;topic=101807.0;attach=886118;image)
(it is attached to the next message.)
I am also attaching the original file (to the next message) if you want to play with it yourself.