I have not ran a test like this with any of the other handheld meters I have. None of them can capture data this fast.
Last night I ran a crude test on my BM869s by driving the data interface cable as fast as possible with a test program. I was able to obtain readings over the cable at a rate of about 200 ms per sample. This was done by starting the next read as soon as the previous read was completed and measuring the interval. I may post more on this in another thread after I do a few more experiments.
Dr Frank was suggesting a higher test current for the low ranges.
I do not have this metter, but majority of multimeters I touched have always problem with low and high margin of ranges, especialy for lowest range is a good test for mesuremant capabilities, e.g. current like 100nA, or 1nA. Even if spec is claiming 0.5 % or so at marginal values for ranges this might be quite high like 10% for lowest range and test current like 10nA, or 1nA , or 1uA at 1000V etc.
I wonder if GW121 is much different from other mutimeters? I would think that it might be actually worst than others due to low burden volatege (?)
You have to look at the full specifications. A tolerance may be 0.5% + 4, the 4 is a very important part of the tolerance specification when measuring low values.
There is also a detail about AC, there the tolerances is usual not valid below 5% to 10% of the range.
I wrote a bit about it here:
https://lygte-info.dk/info/DMMTolerances%20UK.html
I can not find test with nA test current. Did I missed this?
My lowest test current is usual 1uA, I did mention that you need to use REL mode.
.....range and test current like 10nA, or 1nA , or 1uA at 1000V etc.
Are you trying to apply 1000V with a supply that is limited to a uA and measure the voltage? If you are trying to measure current, where does the 1KV fit in? I am not sure what you are asking.
Not uA and not like this, as simple as that:
- 100nA at any voltage (better 10 or 1nA, no idea how 121GW sensitive is)
- 100nA (better 10 or 1nA, no idea how 121GW sensitive is) at 1kV (1kV relative to ground , not accross multiemer)
That would be outside the specifications for the meter, it is only rated for 600V to earth.
I would not care for specifiation for such tests, but other hand I cannot ask anyone for potentialy destructive tests.
I would also not buy 121GW at this moment for own tests, it looks not mature to me.
Are you looking to find leakage current of the voltage range? If so, the leakage at 1kV isn't likely to fit on a straight line curve down to zero, so we would need a few different points.
Hi!
Got my 1.21GW a few days ago.
My first impression was like
![Face Palm :palm:](https://www.eevblog.com/forum/Smileys/default/facepalm.gif)
when i saw the glonky springs trying to hold the batteries in place.
Oh my god, what a bad quality battery holder for a 240 EUR Multimeter.
Without aranging the Springs after puting the batries in the holder they would almost jump out of it.
But the Multimeter itself looks pretty god
![ThumbsUp :-+](https://www.eevblog.com/forum/Smileys/default/icon_smile_thumbsup.gif)
Some measurements in relatve mode makes no sense for me, but i need some more practice.
that is true, the battery holders are terrible. That was also my first impression.
The worst springs of all my battery powerd instruments. I have aligned the springs with a small tweezer, so now it lookes a bit better.
Beside of this a great meter with some room of improvement in the firmware.
I think all of us who bought this meter were shocked at the springs fitted in the battery holder. A well made battery holder fitted with springs which are simply too long.
To use a "Dave" expression when reviewing products the springs are a "FAIL"!
Good meter fitted with the wrong length battery springs, how did this get into the production run?
Was just wondering if it would get ESR in the capacitance measurment aswell
This multimeter has in theory an 50000 counts display , than range switching upward at 55000 counts but downward at 40000 counts is supposed to be a feature ?
I find this rather annoying ... lets say I want to adjust a 5.0000V voltage refference , if I go up beyond 5.5000V I lose the last digit of resolution because is switching the 50V range , but then if I adjust downward I can't get it back since it remains in 50V range until it reaches 4V
@CDaniel - Why not use the provided range switch?
Because it has autorange ...
@CDaniel - You said you can't get it back after you turned down V-ref.. go manual to desired range, then flip it back to auto.
Sure I could , but I don't think is normal to be 2 thresholds , that is my point .
And that was just an example .
I would consider this quite normal, though the hysteresis is a bit on the wide side.
This is what happens with my Fluke 87V:
Counts | Up at | % | Down at | % |
6000 | 6.600 | +10 | 5.40 | -10 |
20000 | 19.999 | +0 | 16.80 | -16 | (Of course, it can't display > 19999) |
The 121GW does it at +10% and -20%.
Sure I can , but I don't think is normal to be 2 thresholds , that is my point .
It's definitely normal! Without hysteresis you would have continuous range change at the threshold, a much worse experience.
I have 4 meters. They all have two thresholds like that. I would imagine to prevent excessive autoranging when taking measurements near the threshold value. I believe it's called hysteresis.
Ok , you are wright , I cheked my other multimeter who has 6,6V up and 6,2V down , which is much smaller and I never noticed since is not around 5V ( I don't think that's random design )
Sure I could , but I don't think is normal to be 2 thresholds , that is my point.
And that was just an example.
There will always be two thresholds to prevent "hunting" of the auto-range circuit when close to the switch point.