Hello,
according to the specification of the "PULSE CURRENT MEASUREMENT OPERATION", there is a burst mode, which features current measurements with higher rates.
BURST MODE:
Measurement Aperture: 33.3μs.
Conversion Rate: 3600/second, typical.
Number of Samples: 1 to 5000.
Transfer Samples Across IEEE Bus in Binary Mode: 4800 bytes/ second, typical.
I did not find in the manual how to set up such a burst measurement. Does anybody know or has an idea?
Online ch_scr
Re: Keithley 2306 (2300 series) burst mode
« Reply #1 on: August 28, 2024, 06:48:12 pm »
The "Pulse current measurement procedure" is explained in the manual (this version of it, anyway) around page 69.
The following users thanked this post: w.v.s.
« Reply #2 on: August 28, 2024, 08:02:01 pm »
Thank you! They call it "Pulse current digitization" in the manual. I've missed that somehow earlier this day.