If you enter the *IDN? command via SCPI, you will get the SN and FW as well as a hardware version.
It would certainly be interesting to observe that.
I checked this with the "Test Controller" software.
Results with software "Test Controller":
*IDN?
;; ZC,ET4410 ,1072240xxxx ,V6.00.2405.039 V1.12.2103.008
In this case, the version V1.12.2103.008 must be the HW version of the device.
It seems, that "Test Controller" changes the Sequence.
When I use the Software "Messinstrumente" by Peter Dreiseibner
the order of the values was correct as in the SCPI document.
The query returns <Manufacturer>,<Model>,<Firmware>,<Hardware>,<SN>
Results with software "Messinstrumente"
Length of data: 66 bytes
ZC,ET4410 ,V6.00.2405.039,V1.12.2103.008,1072240xxxx