I had the opportunity to test the Bandwidth on the following Digital Storage Oscilloscopes:
Siglent SDS1104X-E (100MHz 1GSa/s 4 channels)
Siglent SDS1104X-E software modded to SDS1204X-E (200MHz 1GSa/s 4 channels) (Indicated as SDS1104X-E* in the table)
Siglent SDS1202X-E (200MHz 4 channels) (As tested by TAUTECH)
Keysight DSOX1102G (EDUX1002G uphacked to DSOX1102G) (2GSa/s 2+1 channels)
GWInstek GDS1054B (software license hacked to 300MHz) (1GSa/s 4 channels)
Micsig TO1104 (1GSa/s 4 channels)
Agilent 54622D (200MSa/s 2+16 channels) NOTE: Frequency over 200MHz is indicated as ">200MHz" but can see a very well defined sinewave up to 600MHz (Added 5-May-2019)
HP 54645D (2+16 channels) (Added 5-May-2019)
I used an Agilent signal generator to output a 0dBm (632 mVpp) signal.
Agilent Signal Generator --> N to SMA connector --> SMA - SMA cable --> SMA to BNC --> BNC 50ohm passthrough --> DSO
FrequencySDS1104X-ESDS1104X-E*SDS1202X-EDSOX1102GGDS1054BTO110454622D54645D
10 MHz645 mVpp652 mVpp640 mVpp640 mVpp630 mVpp644 mVpp641 mVpp643 mVpp
100 MHz500 mVpp596 mVpp608 mVpp625 mVpp485 mVpp516 mVpp497 mVpp546 mVpp
200 MHz70 mVpp500 mVpp508 mVpp215 mVpp224 mVpp257 mVpp250 mVpp287 mVpp
300 MHz---348 mVpp352 mVpp28 mVpp47 mVpp49 mVpp106 mVpp72 mVpp
350 MHz---186 mVpp240 mVpp---23 mVpp---72 mVpp---
400 MHz---125 mVpp180 mVpp---9 mVpp---53 mVpp---
500 MHz---40 mVpp------------26 mVpp---
600 MHz------------------10 mVpp---
EDIT 5-MAY-2019: Added SDS1104X-E software modded to 200MHz (SDS1104X-E*), Agilent 54622D and HP 54645D
EDIT 7-MAY-2019: Added SDS1202X-E (TAUTECH)
« Last Edit: May 07, 2019, 12:23:49 pm by TK »
The following users thanked this post: thm_w, plurn
Offline hgjdwx
Re: DSO Bandwidth test SDS1104X-E DSOX1102G TO1104 GDS1054B
« Reply #1 on: May 05, 2019, 05:31:22 am »
The worst must be SDS1104X-E. The best?
Offline rf-loop
« Reply #2 on: May 05, 2019, 08:17:04 am »
Siglent SDS1104X-E (1GSa/s 4 channels)
FrequencySDS1104X-EDSOX1102GGDS1054BTO1104
10 MHz645 mVpp640 mVpp630 mVpp644 mVpp
100 MHz500 mVpp625 mVpp485 mVpp516 mVpp
200 MHz70 mVpp215 mVpp224 mVpp257 mVpp
300 MHz---28 mVpp47 mVpp49 mVpp
350 MHz------23 mVpp---
400 MHz------9 mVpp---
Because there was also modified Good Will 1054B compared to unmodified others so here is modified SDS1104X-E --> SDS1204X-E
SDS1104X-E (sw-mod to 1204X-E)
Generator Hewlett-Packard 8642B  (also checked with HP power meter that levels are well inside specs or much better)
Generator output: N-BNC adapter Suhner.
Cable: M17/084 (RG223) 150cm + Suhner BNC's
Scope input: Terminator Tektronix 011-0049-00
Generator output level set value 0.0dBm
Scope 100mV/div.
10 MHz  636mV
100MHz 561mV
200MHz 494mV
300MHz 342mV
350MHz 256mV
400MHz 171mV
here also member @lzyengineer measurements:SDS1104X-E original + mod
https://www.eevblog.com/forum/testgear/siglent-sds1204x-e-released-for-domestic-markets-in-china/msg2326518/#msg2326518
And in other place can find my measurements also with both: 1104X-E and after mod.
Example in this thread after this message it start https://www.eevblog.com/forum/testgear/siglent-sds1204x-e-released-for-domestic-markets-in-china/msg1624558/#msg1624558
But is it so that more is better? Not always.
All channels on 500MSa/s. SDS1104X-E have good DSP filter for reject aliasing!
~-36dB @ fNyquist
So, for avoid alias effects with measurements example with SDS1204X-E user need really know least some  basic fundamentals. - Mirror have also its backside.
But here bit more fair, both modified. Siglent for 200MHz and GoodWill for 300MHz
                       SDS1104X-E mod           GDS1054B mod
                     responds SDS1204X-Ewith 300MHz license
Frequency      @rf-loop   @TK
10 MHz636 mVpp630 mVpp
100 MHz561 mVpp485 mVpp
200 MHz494 mVpp224 mVpp
300 MHz342 mVpp47 mVpp
350 MHz256 mVpp23 mVpp
400 MHz171 mVpp9 mVpp
« Last Edit: May 05, 2019, 09:01:37 am by rf-loop »
The following users thanked this post: plurn
Online iMo
  • Country: vc
« Reply #3 on: May 05, 2019, 09:31:16 am »
A picture update..
« Last Edit: May 05, 2019, 10:00:19 am by imo »
« Reply #4 on: May 05, 2019, 09:57:35 am »
A picture..
For SDS1204X-E..........a very accurate picture.  ;)
From my own tests ~230 MHz @ -3dB.
Online iMo
  • Country: vc
« Reply #5 on: May 05, 2019, 10:08:31 am »
Better the 1104X-E 200M should follow the shape of the 1104X-E 100M (or DSOX) above, but with the corner shifted to 200M, imho..
« Reply #6 on: May 05, 2019, 10:17:22 am »
I can do -3dB points for real SDS1104X-E and SDS1204X-E if you want........I have to do my pre-dispatch checks so no real problem to use my 510 MHz HP RF gen or SDG3021X SDG3032X.
Online iMo
  • Country: vc
« Reply #7 on: May 05, 2019, 10:18:51 am »
Make it at all above freqs, I will update the picture then..
Offline rf-loop
« Reply #8 on: May 05, 2019, 10:31:57 am »
A picture..
And previously (my previously made quality control's)  SDS1104X-E's  out from box:
individual units I have measured have been between
~>120MHz   ---  ~<130MHz -3dB vs 1MHz
So this previous table is bit odd because TK did not measured any scope real -3dB point
Jump from 100 directly to 200MHz and then  between 300 and 400 also 350.
Proportionally jump from 100 to 200  is like jump from 300 to 600 or 200 to 400 so it is more than odd, specially because he test 50 - 100MHz nameplate scopes.  Peoples are some times ranting biased sellers but is it so that some times owners are even more biased or what to think. But yes, nice test.
« Last Edit: May 05, 2019, 10:41:50 am by rf-loop »
« Reply #9 on: May 05, 2019, 10:57:22 am »
A picture..
Agree 1000%
-3dB point is what matters however can be useful to know how high we can get stable performance albeit with significant attenuation. You saw me push SDS2202X-E to 520 MHz in another thread to make a point and I know the SDS1kX-E cannot be pushed that high.
Back to BW measurements..........
« Reply #10 on: May 05, 2019, 11:11:27 am »
I agree that the correct way is to measure -3dB but I saw several members asking for the max frequency that the scope (modded or original) can measure, so I tested to get to that value as soon as possible.
I will mod my sds1104x-e to sds1204x-e and edit my original results
Offline nctnico
« Reply #11 on: May 05, 2019, 11:17:56 am »
Still the graph in the 3rd reply post gives a good indication. Let's not get over obsessed with the precise -3dB point.
« Last Edit: May 05, 2019, 11:20:22 am by nctnico »
« Reply #12 on: May 05, 2019, 11:47:34 am »
Scopes tested straight from box. Tests = warm up.  ::)
SDS1104X-E
SDS1204X-E
Source SDG3021X (modded to 3.2 GHz model)
BNC cheap 1 GHz Siglent cable
Termination 50 ohm 1x Tek
Level 0dBm
Amplitude can be taken from OSD p-p measurements.
SDS1104X-E ....much as expected.
Images 27, 28, 29(200 MHz) and 30.(300 MHz)
SDS1204X-E..............pushed.  >:D (Frequency counter becomes unstable @ 400+ MHz)
Images 31, 32, 33, 34, 36(350), 35(400), PNG450, PNG500, PNG550 and PNG600  :o
« Last Edit: May 05, 2019, 12:27:45 pm by tautech »
Offline rf-loop
« Reply #13 on: May 05, 2019, 11:48:08 am »
Yes. Only that it is quite common practice to measure -3dB point.
But as you said "Let's not get over obsessed"
But because typically 100MHz scopes have BW limit slope between 100 and 200MHz it is bit rough only measure 100 and 200MHz point (and after then draw straight line between these points as @imo)
Offline rf-loop
« Reply #14 on: May 05, 2019, 11:56:50 am »
SDS1104X-E
SDS1204X-E
Level 0dBm
Images 27, 28, 29 and 30.
What happen in image 29?
I expect roughly around  -6dB
Some thing is now wrong.
Oh... this trigger freq counter.. true signal is 200M.. ;)
« Last Edit: May 05, 2019, 12:00:41 pm by rf-loop »
Offline David Hess
  • DavidH
« Reply #15 on: May 05, 2019, 12:28:52 pm »
With manufacturers like Rigol producing DSOs with unspecified and variable full power bandwidth, these tests are close to useless without specifying the input sensitivity.  In the past this was not a consideration because the full power bandwidth was always higher than the bandwidth at any input sensitivity which was a result of operating the input stages over a much smaller signal range.
« Reply #16 on: May 05, 2019, 12:31:05 pm »
Specified in the OP and plainly visible in every screenshot.
Offline rf-loop
« Reply #17 on: May 05, 2019, 12:52:33 pm »
Please explain.
If I measure freq response using (example) 5, 10, 20, 50, 100, 200, 500mV, etc/div and so on and show all these results. Or if I measure freq response for 100mV/div and show result. What is then wrong there. Can you say this response is not true with used setup.    Do you mean all this is bullshit.
What is wrong. If it is wrong, please give explanation and instructions how to measure.
Or is it just only wrong because these are made in China.
I have repaired and calibrated several Tektronix old and also made frequency response measurements as told in Tek service manuals. Are these also useless.
Measured frequency response is just as result is in told situation. Or do you claim this result is wrong. If it is wrong please come here and show how it need measure so that also you are satisfied.
Offline David Hess
  • DavidH
« Reply #18 on: May 05, 2019, 12:55:51 pm »
Signal level is not the input sensitivity and there are no screenshots associated with TK's post.
Offline nctnico
« Reply #19 on: May 05, 2019, 12:58:05 pm »
Please explain.
David has a good point. On the lower sensitivity settings there usually is a pre-amplifier which also affects the frequency response.  And the attenuators may behave differently at various frequencies as well. So if you want to do a really accurate frequency response graph then you'd need to do this using multiple v/div settings. But then again this is not a very useful exercise because near the bandwidth most (non-sinusoidal) signals will be distorted anyway.
« Last Edit: May 05, 2019, 01:01:32 pm by nctnico »
Online iMo
  • Country: vc
« Reply #20 on: May 05, 2019, 01:08:13 pm »
A picture update..
The following users thanked this post: TK
Offline rf-loop
« Reply #21 on: May 05, 2019, 01:12:25 pm »
Please explain.
Yes, of course.
But still, if measure 100mV/div frequency respose using ok setup it IS 100mV/div frequency response in used setup and not bullshit. Bllshit is if after then claim it tell something what is not measured. Predictors and clairvoyants are a separate issue.
But, of course, and who even think other, it is NOT whole scope frequency response over whole signal level range with all sensitivity settings.  If it is measured using one V/div range with one signal input level and explained setup it is response just with it and nothing else. This is true what ever we test. Result tell only how it is with used setup and nothing else.
« Reply #22 on: May 05, 2019, 01:15:57 pm »
I updated the original test to include the SDS1104X-E software hacked to 200MHz, so far it is the 2nd best on my test
Offline rf-loop
« Reply #23 on: May 05, 2019, 01:16:09 pm »
A picture update..
Wrong, there is err...afaik. ;)
Look tautech 1104X-E result. Where from you get 150MHz data point?
This around -18dB point is 200MHz!
« Last Edit: May 05, 2019, 01:22:06 pm by rf-loop »
« Reply #24 on: May 05, 2019, 01:16:31 pm »
A picture update..
A picture is worth a thousand words (or numbers in this case)