Hi Chris,
When I bought mine it was also at its limits of tolerance. I didn't like that and calibrated it. I documented my procedure that I found by trial and error. Later on, I found some details for the calibration somewhere on the web but I forgot to note where. Anyway, this is how I calibrated my meter. And if I am not exactly forgetting, almost the same procedure applies to 107.
If you have a signal generator, and a multimeter you can trust, you can do a transfer calibration. The good thing with this meter is that any function can be aligned without disturbing others.
Open up 101 and push a side of the sticker from behind to carefully remove it. This way, it will remain unmarred and intact. If you don't care about that then just remove the sticker.
Turn on the meter by rotating the selector switch to the function you want to calibrate then short the S6 pads under the sticker and keep them shorted until a beep is heard and high voltage flash sign appears on the LCD.
Apply the input value for the function and press Hold to save the new value in NVRAM. Then either turn off the meter to get out of the calibration mode or move the selector switch to a new function.
Following are the input values that this meter requires for alignment.
For VDC, it requires 4.500V.
For VAC, it requires 4.500V at 50Hz.
For Capacitance, it requires 100uF capacitor.
I tried to align the Hz function but it didn't accept anything.
Also, I couldn't find anything in my notes about mVAC and Ohms functions. I guess these also not need to be aligned.
I hope it'll help you realign your meter.