We decided to make another transformer out of the last CM Filter we have, and removed the windings then twisted the pair together as before. This time we cut the twisted pair in the middle and created two twisted pair windings on the core. The two windings were arranged as primary in series and secondary in parallel. This should yield a Vo/Vi transfer of 1/2 or -6dBv, and we measured -6.85dBv.
For sanity checks on all 3 transformers used as Injection Bode types, we setup a simple circuit for Closed Loop Bode evaluation. The circuit is a simple Op-Amp (LM358) non-inverting gain of 21 with the feedback resistor of 20K and the shunt to ground resistor of 1K. The + Op-Amp input was grounded thru a 1K resistor and the transformers were connected from the 20K and 1K junction to the Op-Amp - input. The Op-Amp was power by +-10V.
A Siglent SDG2042X AWG was used as the input (we could have used the scope built-in AWG). This was on a LAN with the Siglent SDS2104X Plus scope. Amplitude was set to 0.05Vpp so as not to saturate the small trasnfomers, and confirmed by monitoring the transformer output waveforms.
Thanks to tautech for helping us in the past to get familiar with LAN use (and earlier the SDS2104X Plus), as we didn't know "Diddley Squat" about LANs. BTW Dave used this old redneck engineering term "Diddley Squat" meaning "not much", in his latest How a LCR Meter Works video! Guess this term found it's way to the Land Down-Under
![Grin ;D](https://www.eevblog.com/forum/Smileys/default/grin.gif)
Here's what the latest transformer rendition looks like with the series/parallel windings.
The Bode plots are shown for the mentioned Op-Amp circuit above, measuring the Open-Loop Gain and Phase of a Close-Loop-System. The plots are for the 1st & 2nd transformers with the different cores, and the last is with the split winding transformer. Please note the indicated Gain and Phase Margin indicated by P1 and P2, you see the frequencies at Markers X1 and X2.
Not sure what's going on around 100~200Hz on the left, likely some interaction with the transformer, this should be a smooth plot coming away from 100Hz.
Note the use of the 10 bit mode in the scope, big help with the large dynamic range involved, and with the dynamic channel input scaling the scope performs in Bode Mode, extending the DR even further.
Anyway, this simple, cheap Isolation Transformer seems to work quite well with the SDS2104X Plus and the built in Bode Plot capability. Very powerful feature being able to do Open-Loop measurements on Closed Loop systems without any additional equipment (you can use the scope built-in AWG) other than a DIY low cost Isolation Transformer
![ThumbsUp :-+](https://www.eevblog.com/forum/Smileys/default/icon_smile_thumbsup.gif)
Best,