Guyzz
I have a HP 3478A that's 80ppm off (high) when feeding it 10v from my Geller reference.
I trust my 10v reference, as it is spot on. Whwn measured with my Keithley 2015 (Cal'ed May-2011)
I would like to calibrate "just" the 10v range , is that possible.
Any hints would be nice ....
I have the Servicemanual , and have read the Calibration section.
But i can't see if it's possible to just calibrate the 10v range , or if i have to make all calibrations.
I don't have equipment to calibrate AC or DC above 10v.
TIA
/Bingo
disclaimer: I have never calibrated an HP 3478A, but need to investigate, then calibrate my meter with the standards I can muster up, so I am interested in the process.
after doing the performance tests 4-21 (to log results for a plan of attack)
First of all, the HP 3478A has 5 ranges for DC voltage, 30mV, 300mV, 3V, 30V, 300V
0 and max for each of those ranges, so your 10V can't apply (except for the 30V range, 1/3 full scale)
from reading parts of the manual, you can calibrate any one of those listed ranges individually, and you have to calibrate DC voltage before DC current, then AC voltage, the machine calculates AC current from the DC voltage, DC current and AC voltage values. Ohms is calibrated either 2 wire or 4 wire on the 30ohm, 300ohm, 3Kohm, 30Kohm, 300Kohm, 3Mohm, 30Mohm ranges (entering zero and max of range for each range pass, full scale or 1/3 scale possible, so using either 30ohm, 300ohm, 3Kohm, 30Kohm, 300Kohm, 3Mohm, 30Mohm standard resistors or 10ohm, 100ohm, 1Kohm, 10Kohm, 100Kohm, 1Mohm, 10Mohm standard resistors).
calibrating DC voltageyou would do section 4-60 from a to b
c. Set the 3478A to the single trigger mode by pressing the SGL/TRIG button. Take the 3478A out of Autorange by pressing the downrange button.
c-1 set the 3478A to the required range, either one of 30mV, 300mV, 3V, 30V, 300V.d. Install a short across the 3478A's HI and LO INPUT terminals
(this sets the low value of the range)e. Press the blue shift button and then the LOCAL (CAL) button.
f. Press the SGL/TRIG button once. When the display shown CAL FINISHED, remove the short form the input.
(zero for the range entered)g. Connect the DC Volts Standard output terminals to the 3478A INPUT terminals as shown in Figure 4-2.
Set the DC Volts Standard for the same output as selected on the HP 3478A (30mV, 300mV, 3V, 30V, 300V).h. Press the blue shift button and then the LOCAL (CAL) button
i. If the actual output value of the DC Volts Standard is known, use the 3478A uprange or downrange buttons to set the 3478A display to the same value. Then press the SGL/TRIG button. If the actual output is not known or if the Standard has a high enough accuracy, just press the SGL/TRIG button.
(max of the range has been entered).
i-1 (I assume CAL FINISHED is shown on the display)repeat from c-1 for all required ranges
in red is my interpretation, go ahead, take my head offit's my interpretation of the text I read
please feel free to correct my interpretation