I ran a more systematic experiment but had similar results. I’m at a bit of a disadvantage since I no longer have a new and measured inductor nor a calibrated meter to compare to. I do have a box full of old inductors and 2 other (uncalibrated) meters.
Summary: Four different inductors were tested. The 2 meters tend to agree with each other and the DE-5000 is reading significantly different. Changing leads, test frequency, or series/parallel on the DE-5000 does not significantly change readings.
Equipment:
Der EE DE-5000 LCR meter with modified TL-21 leads and no leads. Open/Short lead calibration before each reading.
Sinometer VC6243+
Leader LCR-740 LCR Bridge
Subject A: Air-core inductor
DE-5000, TL-21 modified leads, 1kHz
0.791mH (Ls)
0.793mH (Lp)
DE-5000, no leads
0.793mH (Ls)
0.795mH (Lp)
VC6243+ ~2.25kHz (not adjustable)
0.962mH
LCR-740 1Khz
0.962mH
Subject B: Ferrite core
DE-5000 No Leads
1.3mH (Ls, 100Hz)
1.28mH (Ls, 1kHz)
1.16mH (Ls, 10kHz)
VCG6243+
1.52mH
LCR-740
1.51mH
Subject C: Air core
DE-5000
0.448mH (Ls)
VC6243+
0.543mH
LCR-740
0.544mH
Subject D: 14AWG foil inductor marked 0.47
DE-5000
0.394mH
VC6243+
0.467mH
LCR-740
0.480mH