I have the 6071A which is basically the 6070A with a frequency doubler.
If you are inclined to fix things, I strongly suggest you first download and look through the (multi-volume) manual set to see if you want to take the plunge. It is not a simple case of removing the top and/or bottom cover and gaining access to 90% of the parts, as it often is with test equipment. The 6070/1A requires a lot of careful disassembly, especially in the front panel area (where, of course, one of the tantalums in mine recently exploded).
Also, the calibration procedure is long and complicated and if any repairs effect certain calibration constants stored in EPROM, you are basically out of luck since Fluke says in the manual that one should return the unit to Fluke for reprogramming.
As far as the unit goes, it is an OK signal generator. I much prefer, however, the Fluke 6080A/B or the lesser 6060A/B if you can find one of those cheap.