Hi,
High voltage test...
We test our products (power supplies) as follows.
Input to ground, output to ground, input to output if necessary, if there is galvanic isolation between the two.
Whereby ground means the housing(earthing point).
As a rule, we test this with DC voltage, which has the advantage that any capacitors against ground (EMC!) are charged once.*
If we were to test with AC, there would be a unwanted constant current flow.
The inputs/outputs are connected together to protect internal components in the event of a breakdown.
The polarity does not play a role in DC tests in the constellation, but generally minus is connected to ground, while plus is connected to the inputs or outputs.
However, we once had a case where a diode was connected in series in the input as reverse polarity protection.
In this case, minus was connected to the input, plus to ground.
Now there was a breakdown against ground and since the reverse voltage of the diode was lower than the test voltage, it broke due to the breakdown, which would not have happened if plus had been connected to the input.
Since then, plus has always been connected to the input and minus to ground.
What to do in the event of a high-voltage breakdown?
Possible causes:
- Test voltage accidentally set too high
- Tripping current set too low for DC, keyword charging of EMC capacitors against ground
- Insulation of installed cables damaged
- Insulation generally too low between the potentials
- Metal chip residue present in the housing
- Wire residues present in the housing/between the potentials due to incorrect insulation of cables.
- Humidity too high during test (creepage distance: 1mm/1000V)
- Rare: Defective components in the system itself
Repetitions after a failed high voltage test are best carried out with reduced test voltage, as long as this does not fall below the minimum requirement.
Reason:
The components experience a stress phase as a result of a high voltage test, which they should not be subjected to too often (shortening of service life).